The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 10013 - Quality Management System (QMS) Manuals


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 13th October 2003, 07:54 PM
QA Tech QA Tech is offline
Inactive Registered Visitor

Registration Date: Oct 2003
Location: Minnesota
 
Posts: 2
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 25
Karma: 10
QA Tech has less than 100 Karma points so far.
Default ISO 13485- Advisory Notices/Regulations/Quality manual

We are an electronics manufacturing company that is ISO 9001:2000 and adding ISO 13485:2003 by January of 2005. ISO 14969 explains advisory notices to an extent, however it would be nice to see some examples. Anyone have a model for advisory notice process/procedure?

ISO 13485 constantly referrs to national or regional regulations and typically this is FDA QSR, but are there other regulating bodies for medical devices?

Any tips/suggestions on how to incorperate two quality systems (ISO 9001:2000 and ISO 13485:2003) in one quality manual, quality policy, and one set of procedures?

Thanks!
Reply With Quote

Sponsored Links
  #2  
Old 13th October 2003, 08:20 PM
Jimmy Olson Jimmy Olson is offline
E-Mails Invalid or Rejected

Registration Date: Jul 2002
Location: WA
 
Posts: 331
Thanks Given to Others: 0
Thanked 5 Times in 5 Posts
Karma Power: 0
Karma: 226
Jimmy Olson is appreciated, and has over 200 Karma points.Jimmy Olson is appreciated, and has over 200 Karma points.Jimmy Olson is appreciated, and has over 200 Karma points.
Default

Hello and welcome to the cove

I am in the same situation as you. We are an electronic manufacturing company that is looking at adding 13485:2003 (so we can compare notes ) I have a copy of a manual based on the 1996 version of 13485 that is helpful, but I have to clean it up and remove names and references first. I also have some other material that I've gathered that compares the two that I can share, I am on my way out the door, so I will sort through all my stuff tomorrow (just thought I would say hi before taking off).

There are also other people here that have a very good understanding of the medical device field and are very helpful.
Reply With Quote
Sponsored Links

  #3  
Old 15th October 2003, 09:56 AM
Vincnet's Avatar
Vincnet Vincnet is offline
$ Contributor

Registration Date: Jul 2002
Location: France
 
Posts: 54
Thanks Given to Others: 3
Thanked 1 Time in 1 Post
Karma Power: 31
Karma: 20
Vincnet has less than 100 Karma points so far.
Default

Quote:
Originally Posted by QA Tech

ISO 13485 constantly referrs to national or regional regulations and typically this is FDA QSR, but are there other regulating bodies for medical devices?
Hi QA Tech,

As ISO is an Internationnal standard, regional regulations refers to the EU Directives in Eurpore 93/42/EC and 98/79/EC, regulated by the so called Competent Authorities, and to the Japanese regulation or any other regulations in the world.

Vincnet
Reply With Quote
  #4  
Old 15th October 2003, 01:08 PM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Forum Moderator

Registration Date: Jun 2002
Location: Lawn Guyland
Age: 59
 
Posts: 3,101
Thanks Given to Others: 48
Thanked 390 Times in 272 Posts
Karma Power: 192
Karma: 4968
Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Default

Quote:
Originally Posted by QA Tech

Any tips/suggestions on how to incorperate two quality systems (ISO 9001:2000 and ISO 13485:2003) in one quality manual, quality policy, and one set of procedures?

Thanks!
QATech:

I wrote my manual using 9k2k, then I made a comparison clause by clause against 13485:2003(FDIS). Where they differed I modified it approriately to incorporate the additional requirements of 13485 and still meet the customer satisfaction requirement of 9001. I haven't been assessed yet to 13485:2003, but I have been assesed and registered to 13485:1996 and CMDCAS. I haven't had any problems.

BTW, you will need more than 6 documented procedures.
__________________
Al

Last edited by Al Rosen; 15th October 2003 at 05:08 PM.
Reply With Quote
  #5  
Old 11th November 2003, 09:08 AM
Cathy Cathy is offline
Inactive Registered Visitor

Registration Date: Jun 2003
Location: Northern Ireland
 
Posts: 62
Thanks Given to Others: 0
Thanked 8 Times in 3 Posts
Karma Power: 31
Karma: 176
Cathy is appreciated, and has over 100 Karma points.Cathy is appreciated, and has over 100 Karma points.
Default

Hi QA Tech - don't know if you have found an anwser to your advisory notice question. Here's my input

Advisory notices can be issued from the country's competent authorities or by the manufacturer to the competent authority.

When you (as the manufacturer) issue an advisory notices you need to include detailed information on the following:
1. the type of incident (or potential incident)
2. whether any medical device has been involved that the manufacturer is responsible for
3. whether the incident was caused, wholly or partly, by the medical device
4. if there is any short comings in the information supplier with the product.

To see exmaples of advisory notices check out FDA website or try this link though it is for UK competent authority.
http://www.medical-devices.gov.uk/md...pen&count=9999

One thing I will say about the transition from manufacturing to manufacturing medical devices is you need to be very careful in understanding the medical device vigilance. This is built into ISO13485:2003 to some extent but consult the FDA regulations on this area.
Reply With Quote
  #6  
Old 21st December 2004, 11:12 AM
Kari's Avatar
Kari Kari is offline
Inactive Registered Visitor

Registration Date: Oct 2002
Location: Anoka, MN
 
Posts: 12
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 29
Karma: 10
Kari has less than 100 Karma points so far.
Please Help! Advisory Notices

I work for a contract manufacturer that is persuing ISO13485, and builds several different finished devices. I am still confused about Advisory Notices.
Would our customers (i.e., the corporations marketing the devices) be responsible for dealing with them?

(We have never had to deal with adverse event reporting; it is handled by our customer).
Reply With Quote
  #7  
Old 28th December 2004, 10:04 AM
db's Avatar
db db is offline
Where's the shall?

Registration Date: Jul 2001
Location: Plymouth, MI
Age: 53
 
Posts: 2,202
Thanks Given to Others: 208
Thanked 166 Times in 128 Posts
Karma Power: 129
Karma: 2579
db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.db is appreciated, and has over 1700 Karma points.
Default

Quote:
Originally Posted by Kari

I work for a contract manufacturer that is persuing ISO13485, and builds several different finished devices. I am still confused about Advisory Notices.
Would our customers (i.e., the corporations marketing the devices) be responsible for dealing with them?

(We have never had to deal with adverse event reporting; it is handled by our customer).
I would think that advisory notices would be those issued by your company, not your customer. (See Cathy's post). Advisory notices are related to the work your company does.
__________________
Dave B (the other Dave)
Reply With Quote
  #8  
Old 28th December 2004, 05:15 PM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Forum Moderator

Registration Date: Jun 2002
Location: Lawn Guyland
Age: 59
 
Posts: 3,101
Thanks Given to Others: 48
Thanked 390 Times in 272 Posts
Karma Power: 192
Karma: 4968
Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Default

Kari,

Advisory notices are not adverse event reports to the competent authority, they are notices to your customers.

I believe it is the responsibility of the entity that places the device on the market to issue the advisory notices to their customers. But, if you are a contract manufacturer and are to be registered to ISO 13485, it is necesary to address the issue of advisory notices to your customer in accordance with 7.3(d) and 8.5.1.
__________________
Al
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 10013 - Quality Management System (QMS) Manuals

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Who is responsible of Advisory notices atila73 ISO 13485 - Medical Devices - Quality Management Systems 5 7th July 2008 08:19 AM
Established procedure for Advisory Notices - ISO 13485 Chiefs491 ISO 13485 - Medical Devices - Quality Management Systems 7 25th June 2008 11:31 AM
Advisory Notices - Who is responsible for dealing with Advisory Notices? Marc ISO 13485 - Medical Devices - Quality Management Systems 7 11th September 2007 05:41 PM
Advisory Notices - What is the Requirement in ISO 13485 gfux01 ISO 13485 - Medical Devices - Quality Management Systems 7 12th September 2006 11:49 AM
Implementing ISO 13485:2003 - 8.5.1 - Advisory Notices - Recall Procedure Dan LaBelle ISO 13485 - Medical Devices - Quality Management Systems 12 26th June 2006 10:13 AM



The time now is 05:10 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts