The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Measurement Uncertainty (MU)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 4.00 average. Display Modes
  #1  
Old 21st October 2003, 08:31 AM
garcas garcas is offline
Inactive Registered Visitor

Registration Date: Oct 2003
Location: spain
Age: 36
 
Posts: 4
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 25
Karma: 10
garcas has less than 100 Karma points so far.
Cry Baby the GUM against the logic (?)

Hi,

I am currently developing some models for evaluating the uncertainty of measurement in the calibration of acoustical instruments (sound level meters, SLM) conforming to the Guide of Expression of the Uncertainty (GUM).

Although we are talking about acoustical instruments, it is actually an electrical calibration. The calibration is made of several tests, some more complicated than others. In the simplest ones, we apply a stimulus (eg. a sinusoidal signal of certain amplitude and frequency) with a function generator to the SLM and then we check the instrument’s (SLM) response. From an uncertainty point of view, this kind of test does not present any inconvenient to me: I have identified all sources of uncertainty (signal generator’s accuracy, resolution and calibration information; SLM’s resolution; etc) and combined them to find the expanded uncertainty.

The problem arises with the “other” type of tests. In many cases, the tests are what we could call “differential” tests, which means that in those tests we don’t check for the instrument response to a single stimulus, but we check for the difference in the indication of the instrument to two consecutive (and possibly different) stimulus.

One clear example of this kind of test is the “differential linearity level test”. In this test we apply an stimulus S1 and register the SLM response R1. Then we apply a second stimulus S2 whose amplitude is 1 dB higher, and we register the SLM response R2. Then we have to check that the indication of the SLM has incremented in 1 dB (R2-R1=1 dB).

From a logical point of view, I would expect that in this kind of test the uncertainty should be lower than in the “simple tests”, because the systematic errors should be cancelled. But from a GUM point of view, it doesn’t seem to be true: I would have to combine the same sources of uncertainty twice (signal generator’s accuracy, resolution and calibration information; SLM’s resolution; etc), for the first measurement and for the second one. And this combination will be done in squares, so there will be no cancellation. What am I missing? Should I suppose that the two readings are completely correlated (in this case a subtraction term will appear)?

The thing is that in the documentation of a Bruel and Kjaer calibration system similar to mine, I read that the signal generator didn’t have to be calibrated because all measurements where “differential” (this is not 100% true, but almost). Does this mean, as I suspected, that systematic errors should cancel and thus errors from the generator would not contribute to the total measurement uncertainty?

Thanks for any help. I am quite desperate with this stuff.
Reply With Quote

Sponsored Links
  #2  
Old 22nd October 2003, 12:58 AM
Marc's Avatar
Marc Marc is online now
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,564 Times in 1,017 Posts
Blog Entries: 4
Karma Power: 604
Karma: 11549
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

Bump. Any takers on this one? Sounds interesting but it's beyond me.
__________________
A Search is a terrible thing to waste!
One Test is Worth 1000 Expert Opinions - The plural of anecdote is not data.
We can't solve problems by using the same kind of thinking we used when we created them. - Unknown
Reply With Quote
Sponsored Links

  #3  
Old 22nd October 2003, 02:18 PM
Ryan Wilde's Avatar
Ryan Wilde Ryan Wilde is offline
Inactive Registered Visitor

Registration Date: Feb 2001
Location: Mineola, NY, USA
Age: 42
 
Posts: 187
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 41
Karma: 146
Ryan Wilde is appreciated, and has over 100 Karma points.Ryan Wilde is appreciated, and has over 100 Karma points.
Default

Garcas,

Technically, the GUM is correct, except there are a few things that you can do to lower your uncertainty.

1. A 1 dB step from a modern function generator does not generally alter the attenuator setting (a major source for level uncertainty), unless it requires a range switch, and usually it only involves the leveling circuit.

2. The B&K procedure sounds as if it was monitoring the signal generator output with something along the lines of a AC Voltmeter, hence source accuracy is moot. That is actually the way that I would go. If you can get your hands on something with 6-1/2 digits or more with a good frequency range, you're golden. Since you will once again be in the same Voltmeter range, the baseline uncertainty cancels, and only the "PPM of Reading" would be correlated. Several voltmeters are actually specified separately for ratio - which is two measurements in a single range.

Hope this helps a bit,

Ryan
Reply With Quote
  #4  
Old 23rd October 2003, 12:24 PM
garcas garcas is offline
Inactive Registered Visitor

Registration Date: Oct 2003
Location: spain
Age: 36
 
Posts: 4
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 25
Karma: 10
garcas has less than 100 Karma points so far.
Default

Hi Ryan,

First of all, thanks for your help. It's the first time in weeks that I get some.

I agree with you that a 1 dB step usually does not alter the attenuator settings, so I guess in those cases linearity is the most important source of uncertainty in the generator. But the problem is that manufacturers usually give a single accuracy figure in their specifications, so when computing the uncertainty due to the generator I can not distinguish between the “attenuator uncertainty” and the “uncertainty due to linearity”.

I think it would be easier to work with an example. Let’s say my calibration consists of the following steps:
0. Connect the signal generator’s output to the sound level meter (SLM) electrical input.
1. Apply a sinusoidal signal of 94 dB to the SLM. The microphone sensitivity is 50 mV/Pa, so the generator should be configured to produce a sinusoidal signal of 50 mV.
2. Annotate the SLM’s indication (let’s say the indication is 94.1).
3. Apply a sinusoidal signal 10 dB higher: which means configuring the generator with a 158.5 mV signal.
4. Annotate the SLM’s indication (let’s say the indication is 104.5).
5. Check that the SLM’s response has incremented in 10 dB, with a tolerance of +/- 0.4 dB.

Thus, my variable is the difference in indication (104.5-94.1=10.4). So I would have to check that 10.4 +/- U lies within +/-0.4dB of the expected difference (10 dB). Then I have to calculate the uncertainty of measurement U. Here is my doubt and this is what I was trying to ask in my last post: do I have to take into account the accuracy specifications of the generator when calculating the uncertainty of this measurement?

My logic says no, or more accurate “not completely”. What I mean is that: the offset error of the signal generator would be cancelled between the two readings; but linearity and gain errors would not. The problem is that I only have a single accuracy specification figure for the signal generator: its accuracy. So what do I do? Do I use it or not? I think that if I apply the GUM I have to add the generator accuracy in squares, so there would be no cancellation. I only see a GUM case where there would be some short of cancellation: if we say that both measurements are correlated with a positive correlation coefficient.
Reply With Quote
  #5  
Old 24th October 2003, 03:55 PM
Ryan Wilde's Avatar
Ryan Wilde Ryan Wilde is offline
Inactive Registered Visitor

Registration Date: Feb 2001
Location: Mineola, NY, USA
Age: 42
 
Posts: 187
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 41
Karma: 146
Ryan Wilde is appreciated, and has over 100 Karma points.Ryan Wilde is appreciated, and has over 100 Karma points.
Default

Yes, you most definitely have to take your generator level accuracy into account. Actually, if you change the generator the 10 dB you cited in your example, the only amount of change that I can guarantee that you didn't have is exactly 10 dB. You will have to determine what the actual change is, or take the level uncertainty at face value, which is generally going to be an unreasonably large figure for your application.

You are correct that the positive correlation coefficient would be appropriate. The problem is that the coefficient is an unknown, because the function generator isn't spec'd for it. The only way to know would be to monitor it with a ac voltmeter to find out, in which case, you are probably better off using the uncertainty of the ac voltmeter (minus the noise floor spec, which DOES cancel out directly). If you monitor, you don't need to uncertainty of the generator, because you are monitoring it.

In short, unless you are monitoring the generator (HP 3458A, 34401A, or 3478A; Fluke/Wavetek/Datron 1281, etc. all have fairly low uncertainties with good frequency range) you have to either use the level uncertainty twice, perform tests on your generator to prove its gain and linearity error, or make an educated "guess" - which I wouldn't suggest.

Ryan
Reply With Quote
  #6  
Old 27th October 2003, 04:34 AM
garcas garcas is offline
Inactive Registered Visitor

Registration Date: Oct 2003
Location: spain
Age: 36
 
Posts: 4
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 25
Karma: 10
garcas has less than 100 Karma points so far.
Default

Hi again,

That was really good, thank you. I think you are helping me a lot to find my way

The only way I think I could evaluate the correlation coefficient, would be by repeating several times this measurement and estimating it from the experimental results. Because the coefficient would depend on both the generator and the sound level meter (SLM) under test, isn’t it?

This would lower my uncertainty, but I would have to repeat each measurement several times, which means more calibration time (and I cannot afford that) .

So what about estimating the correlation coefficient from previous calibration data? Does it make sense? I have an historical of all calibrations I have perform until now, so maybe I could do an “off-line” estimation of the correlation coefficient for each generator-SLM pair and then use this to calculate the uncertainty.

This way I wouldn’t have to repeat each measurement several times, unless I would I have to calibrate a SLM that I haven’t calibrate yet. What do you think about this approach?

Thanks again Ryan… (and anyone else on this?)
Guillermo
Reply With Quote
  #7  
Old 12th November 2003, 06:54 PM
Ryan Wilde's Avatar
Ryan Wilde Ryan Wilde is offline
Inactive Registered Visitor

Registration Date: Feb 2001
Location: Mineola, NY, USA
Age: 42
 
Posts: 187
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 41
Karma: 146
Ryan Wilde is appreciated, and has over 100 Karma points.Ryan Wilde is appreciated, and has over 100 Karma points.
Default

Guillermo,

Using past history will give you reproduceability (lack of variation over time), but will not show bias (is 10 dB actually 10 dB?).

For example, if my generator is actually supplying an 11 dB change, and my SLM is reading this as 10 dB, I might unwittingly call it in-tolerance. Next year, if my generator supplies 11.02 dB change, and my SLM reads 10.02 dB, I have fantastic reproduceability, but I am reproducing a false number.

Without knowing exactly what the generator is supplying, my data means very little, since the generator is not specified for that type of measurement. Hence, I monitor it with a piece of equipment that IS specified for that type of measurement, and transfer the accuracy to my supply, and therefore to my SLM.

Does that make sense?

Ryan
Reply With Quote
  #8  
Old 13th November 2003, 07:59 AM
garcas garcas is offline
Inactive Registered Visitor

Registration Date: Oct 2003
Location: spain
Age: 36
 
Posts: 4
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 25
Karma: 10
garcas has less than 100 Karma points so far.
Default

Hi Brian,

I agree with you, but I think I am already having this in mind as I calibrate my generator once a year and this information is also included in my uncertainty budget.

The problem is about the correlation coefficient between 2 successive measurements. I wonder if I could estimate it from previous data or should I estimate it from repeated measurements.

Another problem I have is about how to include the systematic effect’s correction uncertainties according to GUM clause 4.1.4 when I have a non linear uncertainty model. I have already defined a model for evaluating the uncertainty of measurement in a simple test where I measure a certain quantity X. Then, I have to develop an uncertainty model for a second test where we measure a quantity Y = X1/X2, where X1 and X2 are the result of two measurements of the quantity X. According to the GUM (4.1.4 and H2 example) there are two approximations for this case (determination of the Y=X1/X2). I think I should use the second approximation in example H2, but I do not know how to include the systematic effect’s correction uncertainties in this approximation (the GUM says it should be included, but it does not say how to do it).

Guillermo
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Measurement Uncertainty (MU)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Logic riddle abelardo Brain Teasers and Puzzles 4 6th June 2009 01:27 PM
Final draft Supplement 1 to the GUM stefanhg Measurement Uncertainty (MU) 3 7th August 2008 07:20 PM
GPS Gum Balls - High Speed Police Chases Cari Spears General News and Business Articles 4 8th February 2006 03:38 PM
A2LA Contact and Guide 25 Comment - NOTE: Guide 25 is long since dead and buried Marc ISO 17025 - Calibration, Measurement Gages and Test Laboratories 4 18th October 2004 02:29 PM
How to determine Type-B uncertainty?What distribution?With Calibration beginner guide nick1980 Measurement Uncertainty (MU) 11 5th October 2004 11:44 AM



The time now is 08:37 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts