Calibration Discrimination - Acceptance criteria is not having the accuracy required?


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links



Courtesy Quick Links

Links that Cove visitors will find useful in your quest for knowledge:


International Standards Bodies - World Wide Standards Bodies

ASQ - American Society for Quality

International Standards Organization - ISO Standards and Information

Atul's
Quality Forum Online

Howard's
International Quality Services

Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum

Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology

Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 29th March 2004, 03:40 PM
Andrews

 
 
Total Posts: 312
Question Calibration Discrimination - Acceptance criteria is not having the accuracy required?

A temperature indicator is used to check temperature of a bath that has to be maintained at 35.2 deg celsius to 39.4 deg celsius.This temp. indicator is being calibrated and the acceptance criteria is fixed as +/- 1 deg celsius.Isn't this a non-conformity as the acceptance criteria is not having the accuracy required by the test?Shouldn't the acceptance criteria be +/-0.1 deg celsius?

Comments pls.

Andrews

Sponsored Links
  Post Number #2  
Old 29th March 2004, 03:52 PM
The Taz!'s Avatar
The Taz!

 
 
Total Posts: 499
Depends . . . you must have the discrimination to read to 0.1 or 0.05 to see the variation in the process. IMHO, I would like to see the instrument calibrated within the working range and at the resolution required. You are paying for the calibration thus you are the customer and make the requirements.

You might want to check around and find out what is common practice for that type of process before getting into a dispute.
Sponsored Links

  Post Number #3  
Old 29th March 2004, 04:04 PM
tomvehoski

 
 
Total Posts: 944
There are several questions to ask:

How critical is 35.2 to 39.4? What happens if it is 39.8? Or 42?

If nothing happens, why is the spec. so tight and why are you bothering to calibrate the gage. If it may create bad material, slow down the process, etc - it is probably best to have it calibrated. If the building explodes, it should definately be calibrated. You need to make the call based on the risk and cost/benefit analysis.

What type of device is being used? RTD, Thermocouple, thermometer, etc?

For a range that narrow you may need to look at calibrating both the sensor (RTD/TC) and the readout device. Its been awhile, but I recall you can get a TC certified - for big $$$. They are not very accurate devices. Even if you have resolution to .1 degree, you need to make sure the device is still accurate enough.

Hope this helps,

Tom
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > Common Quality Assurance Processes and Tools > General Metrology, Measurement Device, Calibration and Test Laboratories > General Measurement Device and Calibration Topics

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
How to define Calibration Acceptance Criteria xixide General Measurement Device and Calibration Topics 3 10th October 2014 01:31 AM
Acceptance Criteria for Calibration by an External Calibration Laboratory ram4302 General Measurement Device and Calibration Topics 5 21st August 2010 10:19 AM
Torque Calibration Accuracy Criteria Douglas E. Purdy General Measurement Device and Calibration Topics 7 10th August 2007 03:45 PM
Calibration 4.11 2c - Defining acceptance criteria for all equipment Wendell Goodson ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3 27th June 2000 09:07 AM