The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 17th February 1999, 11:09 PM
Dawn Dawn is offline
E-Mails Invalid or Rejected

Registration Date: Sep 1998
Location: St. Marys, PA
 
Posts: 302
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 40
Dawn has less than 100 Karma points so far.
Please Help! Dial Indicator (Analogue) Calibration Work Instructions

Can anyone help me out on work instructions for how to calibrate dial indicators? I have an idea, but I'm not sure its all I need. Thanks!
Reply With Quote

Sponsored Links
  #2  
Old 5th March 1999, 09:34 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,567 Times in 1,019 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11564
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

I'm not going to promise, but I might. I'll look tomorrow.
Reply With Quote
Sponsored Links

  #3  
Old 6th March 1999, 10:49 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,567 Times in 1,019 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11564
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

Best I can do - maybe give you some ideas.

--------snippo----------

1. Material Required
1.1. Cleaning solution
1.2. Supermic (XXX)
1.3. Magnetic base
1.4. Clean cloth
1.5. Calibration void if seal damaged sticker
2. Instructions
2.1. Calibration shall be performed in the Q. A. Laboratory.
2.2. Inspect and record as found condition prior to calibration. Record results on worksheet, W-11019A.
2.3. Clean indicator thoroughly using cleaning solution and clean cloth.
2.4. Visually inspect contact tip for wear, also check contact tip and screws in the back. Make sure they are tight.
2.5. Place indicator stem in the hole of bracket fixture of supermic.
2.6. Next push in on Indicator slightly until dial hand comes in contact with anvil of supermic and indicator reads zero. Tighten snug by turning screw on side of bracket.
2.7. Zero the supermic by pushing recall button and enter 0.0000, then push reset button twice.
2.8. Turn wheel on supermic inward until dial hand of Indicator is at one revolution.
2.9. Read the results from supermic and record on worksheet, W-11019A.
2.10. Continue to turn supermic wheel inward to second revolution, repeat 2.8, then to two and a half revolutions and record results on worksheet, W-11019A.
2.11. Next turn supermic wheel backward two and a half revolution to zero immediately after the inward movement, without resetting the indicator dial. (The maximum difference is known as "hysteresis").
2.12. Record as left results on worksheet, W-11019A.
2.13. Enter all results in computerized gage control post calibration file.
2.14. Retain worksheet, W-11019A, one day to ensure computerized file backup.

3. To Calibrate a Maximum Indicator Do The following:
3.1. Repeat section 2.1 through 2.6.
3.2. Turn wheel on supermic inward and to max. readout - record error worksheet, W-11019A.
3.3. Next turn supermic wheel backward to zero reading without resetting.
3.4. Record error results on worksheet, W-11019A.
3.5. If return to zero isn't at zero, push zero button on front of Indicator and turn supermic wheel outward and repeat 3.1 through 3.5 and record results on work- sheet, W-11019A. (If necessary to adjust, turn adjustment screw on back of indicator).
3.6. Enter results in computerized gage control post calibration file.
3.7. Retain worksheet, W-11019A, one day to ensure computerized file backup.
3.8. Calibration void if seal is damaged ; sticker must be on adjustment screw on back of indicator.
4. To Check Bestest Interapid Type Indicators, Do The Following:
4.1. Calibration shall be performed in the Q. A. Laboratory.
4.2. Set up Pratt and Whitney supermic with a magnetic base and special clamp to hold indicator.
4.3. Place Indicator in clamp on magnetic base and tighten. Bring Indicator tip in contact to anvil of supermic until reading is observed.
4.4. Rotate setting wheel of supermic clockwise until zero reading is on the Indicator.
4.5. Zero supermic by pushing recall button and enter 0.000, then push reset button twice.
4.6. Rotate setting wheel of supermic clockwise throughout travel of the Indicator.
4.7. Take reading at various increments and record results on worksheet, W-11019A (See computerized file for increments of particular Indicators).
4.8. Next turn supermic wheel counter clockwise back to zero reading on Indicator and record results on worksheet, W-11019A.
4.9. Continue counter clockwise (with supermic and Indicator both reading zero) through out travel of the Indicator, repeat 4.7, then counter clockwise to zero, and record results on worksheet, W-11019A.
4.10. Record results in computerized gage control post calibration file.
4.11. Retain worksheet, W-11019A, for one day to ensure computerized file backup.
5. Tolerance
5.1. Dial indicator shall be acceptable when within + least graduation to 2 1/2 revolutions.
5.2. Maximum indicator shall be acceptable when within + one according to graduation at full range.
6. References
6.1. W-11019A Calibration Results for Indicators Worksheet


[This message has been edited by Marc Smith (edited 03-06-99).]
Reply With Quote
  #4  
Old 6th March 1999, 10:57 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,567 Times in 1,019 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11564
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

See:

http://Elsmar.com/ubb/Forum4/HTML/000049.html
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Digital Dial Indicator Recommendation SGCalLab ISO 17025 - Calibration, Measurement Gages and Test Laboratories 4 29th September 2009 05:27 PM
How to Calibrate a Digimatic Dial Indicator? jhoniegudel ISO 17025 - Calibration, Measurement Gages and Test Laboratories 5 25th June 2007 07:04 AM
Difference between written calibration procedures and calibration work instructions? Timmers ISO 17025 - Calibration, Measurement Gages and Test Laboratories 8 5th June 2007 08:39 AM
Calibration Wording for a Dial Indicator WI QC@s2yachts.com ISO 17025 - Calibration, Measurement Gages and Test Laboratories 1 14th June 2001 01:09 AM
How to Calibrate Dial Indicator brossbach ISO 17025 - Calibration, Measurement Gages and Test Laboratories 5 10th May 2001 12:25 AM



The time now is 08:07 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts