The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 21st April 1999, 12:20 AM
George Trybulski George Trybulski is offline
Involved in Discussions

Registration Date: Jan 1997
Location: Rochester, NY
 
Posts: 20
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 53
Karma: 10
George Trybulski has less than 100 Karma points so far.
Question Gage Pin Dilemma - Is there an "easier" way to put our gage pins in our system?

Our company is ISO registered, however I want to know if there is an "easier"way to put our gage pins in the system. Example: we name our gage sets "GP-1", "GP-2 and so on. Each pin in the set is listed on a calibration report as 1-250,1-251 in set GP-1. Have I confused you so far, let me continue. The pins larger than .150 have the set number etched on the end "1". So these pins can be traced to the individual sets. HOWEVER the pins less then .150 are not marked because of size constraints. We have 4 common size pin sets and the pins less then .150 can possibly get mixed between sets. In addition we have "and get this" 5300 gage pins in our calibration system. Is this overkill on our part ? This gage pin dilemma has been a pain in my butt for years. Any comments would be appreciated other then get rid of all the pins ! Thanks.
Reply With Quote

Sponsored Links
  #2  
Old 21st April 1999, 02:04 AM
Marc's Avatar
Marc Marc is online now
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,567 Times in 1,019 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11564
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

I have seen this addressed with paint drops and I once saw a company use dyes. But - Use 'common sense' here. Each pin gage does not have to be identified on the actual pin. There is an allowance for 'due care' with respect to reality.

For example, if I come into your lab and it is neat and orderly and you show me 4 labeled (identified and cal sticker on box) boxes each of which contain identical sets of 50 pin gages, I may ask you:

1. Why the different sets - what do you use each set for?

2. How do you keep them apart?

If you told me you use 'due care', and the area is neat, orderly and the users are 'aware', there's not much I can say. There is reality I have to deal with. The stated requirement (show me) in QS/ISO9000 which dictates that each individual gage be identified is 4.11.2 d "...identify....with a suitable indicator or approved identification record to show the calibration status. Typically things like pin gages are identified by the box they are in. The key is the verbiage "or approved identification record. It simply does not say: "Unconditionally, each item must be individually marked or labeled." While I realize that the QS folks decided should and shall mean virtually the same thing, I have not seen them compromise the word or - yet - which implies a choice or option.

An example of 'common sense' is where they talk about protecting measurement equipment against 'adjustment'. 4.11.2 i "...safeguard ....from adjustments which would invalidate the calibration setting." Well, the question becomes just how far do you go? And to take it to an extreme, can you protect against deliberate 'sabatageOr is that even the intent? I take the inadvertent view - if normal use could cause 'inadvertent adjustment', protect against it. The point here is that most equipment is by design protected against 'inadvertent' calibration. Let's face it, if you buy a calipers it is pretty well sealed against someone accidentally changing it unless they are really trying to do so - like on purpose. Next comes the "well, seal it" which brings us back to 'Can you really protect against sabatage?" Heck, if I want to change the setting and it is sealed, I can get a seal if I really want to sabatage the instrument. So - do I place a guard by the instrument 24 hours a day to ensure no one changes it? What if the 'sabateur' buys off the guard?

Just be able to show you use 'due care' in keeping your pin gage sets 'together'. A big part of doing that is a clean, neat, orderly laboratory. Consider whether there is a significant difference in use of the sets and be ready to talk to that specifically with regard to calibration cycle length. If the auditor comes in and the place is not kept up and they notice inattention, sloppiness or such, they will drill you because the appearance will be that you do not use 'due care' in general operations.

How can you ensure that one or more will never become mixed between boxes? Well, you really can't. You should be able to show that, for example, if three boxes are open there is a positive awareness of what pin is out of what box. This is really just good methodology. If you have two boxes open and each has the same pin out, it should be because two different users are using them. If they're at the same bench together, I would want to boxes separated if they are on the bench and I would want to ask each user which box their pin belongs in. The best word I can think of to use here is awareness on the part of the user. Put them away properly after use - don't leave lying around. That sort of thing. Good practices. Cleanliness, order, neatness, awareness.

Quote:
In addition we have "and get this" 5300 gage pins in our calibration system.
Sounds like a lot of pins to me! How many 'sets' do you have? How many duplicates? Could some be retired? Are they all sets 'working' or are some for cal?

Hopefully some of the other kind folks will help out here!

[This message has been edited by Marc Smith (edited 04-21-99).]
Reply With Quote
Sponsored Links

  #3  
Old 21st April 1999, 10:13 AM
barb butrym barb butrym is offline
Courtesy Access

Registration Date: Dec 1998
Location: South Central Massachusetts
 
Posts: 788
Thanks Given to Others: 0
Thanked 15 Times in 12 Posts
Karma Power: 63
Karma: 151
barb butrym is appreciated, and has over 100 Karma points.barb butrym is appreciated, and has over 100 Karma points.
Default

doesn't look like you need help, Marc....you are right on.....the appearance of the lab is what gives the auditor a comfort level for the control. What ever you do, in regard to the marking, needs to make sense...but due care is my choice
Reply With Quote
  #4  
Old 21st April 1999, 03:36 PM
Batman
Unregistered Guest

 
Posts: n/a
Default

Agree. Keep it simple and effective. We have a few sets of gage pins, - .050 up to .750" - each box is marked as to the calib date. We only identify each set as a number, not each individual pin. We also [usually] have a few missing, so the gage guy puts a dated list of "missing at last calibration" inside each box so if one or more pins are later returned, the user will know that those are not calibrated, and can get them done; those then come off the list.

The only issue I see here is if it is routine to lend out pins to anyone. Probably need a "checkout" list so the correct pin has a chance of getting back into the correct box. I don't think this is a big issue unless you could mix "minus" pins with "plus" pins, if you have that sort of inventory.
Reply With Quote
  #5  
Old 21st April 1999, 10:48 PM
George Trybulski George Trybulski is offline
Involved in Discussions

Registration Date: Jan 1997
Location: Rochester, NY
 
Posts: 20
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 53
Karma: 10
George Trybulski has less than 100 Karma points so far.
Default

Thank you very much for the speedy reply. Actually my thoughts were on the same line but just wanted other peoples imput. We do issue pins for CNC machine floor use and have "conformance sheets" that list the attribute critera that the pins check against. We use the data from these sheets for formulating final inspection acceptance, so the gage pins can be said to be used for final acceptance, so calibration is a must. My last question is this: If we state that our "tolerance" for our gage pins is +/-.0001
can we just report that the "set" is within specification without having a seperate record for each pin. When we issue the pins they are in a peg board block with the gage set and size labeled by each pin. That way when they are returned, they go back in their respective sets. WORKS FOR ME ! Have a great life ! Thanks again !
Reply With Quote
  #6  
Old 2nd May 1999, 01:56 AM
Marc's Avatar
Marc Marc is online now
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,567 Times in 1,019 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11564
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

Subject: pins
Date: Fri, 23 Apr 1999 18:17:27 EDT
From: GFT1005@aol.com
To: cheech@qs9000.com


Thank you very much for the speedy reply. Actually my thoughts were on the same line but just wanted other peoples imput. We do issue pins for CNC machine floor use and have "conformance sheets" that list the attribute critera that the pins check against. We use the data from these sheets for formulating final inspection acceptance, so the gage pins can be said to be used for final acceptance, so calibration is a must. My last question is this: If we state that our "tolerance" for our gage pins is +/-.0001 can we just report that the "set" is within specification without having a seperate record for each pin. When we issue the pins they are in a peg board block with the gage set and size labeled by each pin. That way when they are returned, they go back in their respective sets. WORKS FOR ME ! Have a great life ! Thanks again !
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Gage pin calibration procedure for accurately calibrating gage pins Coleman Donnelly ISO 17025 - Calibration, Measurement Gages and Test Laboratories 15 5th December 2008 02:38 PM
Gage Block Measurement - Home made gage blocks at 1" intervals Coleman Donnelly ISO/TS 16949 - International Automotive Quality Systems Standard 7 31st March 2008 12:24 PM
Pin Gage - Gage R&R Studies - One Pin or Entire Set? Adam Paris Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2 8th December 2007 11:50 AM
"Product" and "Process" columns in CP - what to put there? brutas FMEA and Control Plans 28 24th August 2007 10:51 AM
Gage Block Search - Looking for gage blocks that are 12, 24 and 40" gpainter ISO 17025 - Calibration, Measurement Gages and Test Laboratories 5 10th December 2001 11:26 PM



The time now is 09:39 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts