The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 2 votes, 4.50 average. Display Modes
  #1  
Old 13th October 1998, 03:45 PM
SCOTT SNYDER
Unregistered Guest

 
Posts: n/a
Default CAL DUE DATE

THIS IS A 2 PART QUESTION.
1) I am have a dissagrement with my boss over my cal due date. My computer software gives a due date as mm/dd/yy, however I would like to state in the procedure that the gauge is not past due until the last day of the month. Everyone outside of my plant says this is how they do it. He insists we would not pass our qs audit with the procedure written this way. He claims that a gauge due on the 1st of the month would still be overdue by 30 days?
2) How can I control or justify changing calibration frequency's ? Is there specific statistical controls or test to do this without great time investments in reserching history files? I am refering now to mechanical gauges(ie. calipers,mic's,etc.)
Reply With Quote

Sponsored Links
  #2  
Old 13th October 1998, 04:26 PM
Leslie Garon
Unregistered Guest

 
Posts: n/a
Default

Scott,

I have answers to your questions. So will others.

As long as your procedure says that the day is unimportant and that calibration is due sometime within the month listed and only considered late when that month is passed (wording better than this is highly recommended LOL) than QS will accept it as long as you do it that way and you can prove it is appropriate.

If your program is set for mm/dd/yy why not try to always use the same /dd/ for all instruments. or see if you can change the format to mm/yy.

This is allowed by QS. Stating that the day listed is a target but it is not considered late until after the last day of the month works.

As for justifying the change, look at your operations:
Is it critical that certian equipment be calibrated on a precision (exact) time schedule or only that the schedule must be accurate (on or about). Is it cost effective? Do you have the resources? Is it efficient? Is there waste as a result? Is the time schedule practical? does it promote quality or make the cost of quality too high? and lastly is it really necessary to calibrate at the frequency currently implemented? For this last question, use history data and mfg'r recommendation as a guide.

I hope this was some help.
Reply With Quote
Sponsored Links

  #3  
Old 13th October 1998, 09:18 PM
Don Winton's Avatar
Don Winton Don Winton is offline
dWizard

Registration Date: Nov 1998
Location: Tullahoma, TN
Age: 53
 
Posts: 475
Thanks Given to Others: 0
Thanked 12 Times in 9 Posts
Karma Power: 63
Karma: 488
Don Winton is appreciated, and has over 400 Karma points.Don Winton is appreciated, and has over 400 Karma points.Don Winton is appreciated, and has over 400 Karma points.Don Winton is appreciated, and has over 400 Karma points.Don Winton is appreciated, and has over 400 Karma points.
Default

Scott,

I agree with Leslie. I would suggest the mm/yy (or yy/mm) format. As far as justifying the changes, I am sure your document control procedure allows for this kind of flexibility. You are not changing the frequency, just the details. The yy/mm format that I used in a MIL-I-45208 environment was fine with the GSA auditors (surely QS assessors are not as bad as the GSA guys). I would drop the /dd/ from the date format, audit your documentation to ensure that it is up to date, and carry on.

I am not a QS specialist, but I know that this method would work for an ISO style of assessment. Justify and document; you should be fine.

Best Regards,
Don

Reply With Quote
  #4  
Old 14th October 1998, 05:02 PM
Jennifer
Unregistered Guest

 
Posts: n/a
Default

Scott,
1) We state in our procedure that the calibration is due sometime during the month noted as the due "date". So far, it has passed 2 surveillance audits and 3 different auditors with no problems.
2)As far as changing the calibration frequency, we base it on historical data (we have checked this gage once a month and have never found a problem - we can change it to every 6 months.) We do not mess with frequencies on items that measure critical parameters in our system (such as chemistries and temperatures).
I hope this helps.
Jennifer
Reply With Quote
  #5  
Old 16th October 1998, 01:05 AM
Scott Knutson
Unregistered Guest

 
Posts: n/a
Default

Scott -
We do the same thing as Jennifer's company, including using historical data to change cal frequencies, and we are QS certified.
Reply With Quote
  #6  
Old 10th September 1999, 05:14 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,852
Thanks Given to Others: 1,892
Thanked 1,563 Times in 1,016 Posts
Blog Entries: 4
Karma Power: 604
Karma: 11544
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Lightbulb

I know we had a calibration frequency debate going in a recent thread, but the search engine didn't find it - so I'm putting this here. If anyone knows where the other cal frequency thread is, let me know and I'll link these.

-----snippo-----

Date: Thu, 9 Sep 1999 10:33:25 -0700
From: Bruce Mayfield
To: Greg Gogates
Subject: Survey Results

To learn more about how other companies, organizations and our customers handled calibration intervals I set up a Focus group to discuss two questions

These were:

1. Does your company use the recommended interval on the Calibration Certificate to determine your calibration interval?

2. Does this change cause you any problem with your subcontractors not providing a recommended calibration interval in compliance to the requirement?

Here are the results I obtained through the sample of respondents I received:

Demographically here is how the group broke down:

Government Internal Service Company
Labs Company Labs Testing/ Calibration Labs

13.7% 48.3% 37.9%


The focus group contained organizations from the USA and other countries:


USA Other

78.6% 21.4%

Question #1. 27.5% responded Yes they used the recommended intervals.
They also stated that they used them (about 50% of the
respondents) as the introductory interval and then adjusted
the interval to meet their internal requirements. About 50%
of those responding Yes, represented facilities out side of
the USA. The balance, 72.4%, developed their own
internal calibration intervals and did not use the
recommended intervals from other companies/organizations.


Question #2. 17.2% responded Yes that it would cause a problem. In
reviewing the respondents 50% (8.5%) of those identifying
Yes were out side of the USA. About 50% represented the
USA. Again the balance of the respondents 82.7% stated
that it would not make any difference at all.

In summary, if I take all of the comments and numbers as a whole, in only a minority of cases do individual organizations need or use externally derived recommended intervals. Most of those that do, use them as a starting basis for determining their internal derived interval. Those that do use recommended intervals, have not developed the means to properly determine intervals. Those individuals responding from countries that are not as industrialized as the USA, I can understand why there would be problems with this concept. For those organizations in the USA deriving intervals solely based on someone's else's concept of intervals are not in step with concepts outlined in the various guides and standards.

I also want to thank all of those who responded and to Greg Gogates for his patience in forwarding many of the e-mails over to me.

Bruce Mayfield
Telogy
Reply With Quote
  #7  
Old 20th September 2001, 10:24 PM
Yasupawoman
Unregistered Guest

 
Posts: n/a
Question Frequency of Gage R&R

Help!

I am rather new to the area of Gage R&R, Does QS9000 via MSA set a guideline on how often something needs to have Gage R&R studies done? I have read through each....maybe I missed it...

Or is this based on usage or %R&R values

Thanks for your help!
Reply With Quote
  #8  
Old 21st September 2001, 08:05 AM
Audrey
Unregistered Guest

 
Posts: n/a
Default

I am not aware of QS requiring specific calibration period.
It is your decision based on how often you use the gauge, how critical is the feature measured.....
What could help as well would to do a wear sudy on the gauges based on at least 3 previous calibration results.My company has just been certified to the TS16949 standard, and the auditor was quite picky on the MSA studies, how they are performed and why.
Hoping this will help a bit, good luck!!!
Audrey
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
How to get the Calibration Due Date? Calibration Certificates harinandu2003 Calibration Frequency (Interval) 11 3rd June 2009 03:16 AM
Done date & due date on calibration stickers mhirand20 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 12 3rd December 2008 01:14 AM
Calibration grace period after due date GreatManBehinD Calibration Frequency (Interval) 5 31st October 2008 01:24 PM
Extension of Measurement Device Calibration Due Date. digitalfemale AS9100 Aerospace Standard and Requirements 12 24th January 2008 01:56 PM
Late calibration (Calibration Past Due Date) - What do I do now? dbulak ISO 17025 - Calibration, Measurement Gages and Test Laboratories 32 25th May 2007 10:18 AM



The time now is 12:18 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts