Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 29th September 2004, 01:58 AM
A C Diggs A C Diggs is offline
Inactive Registered Visitor

Registration Date: Sep 2004
Location: Kankakee, IL
 
Posts: 1
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 21
Karma: 10
A C Diggs has less than 100 Karma points so far.
Please Help! Calibration Interval (Frequency) Adjustment using Hilliard type 2 algorithm

We have a new manager for our pharmaceutical calibration lab, and he has asked me for help deciphering our SOP on cal interval adjustments. Currently, the SOP states that we use the Hilliard type 2 calibration interval assignment algorithm. I have run some senerios through the formula and cannot come up with a situation that would qualify to extend the interval (not even zero OOTs worked). Perhaps, I'm not applying the formula correctly.
I work in a fair size shop with 3000 instruments to cal (approx 400 per month). We do not have a validated program to crunch these numbers for interval adjustment according to the Hilliard method.
We have kicked around the idea of rewriting the SOP and using a simple look-up table based on the previous 4 calibrations, but then we ran into the question of how do we test this table for its ability to meet our relaibility target?
Reply With Quote

Sponsored Links
  #2  
Old 19th December 2004, 12:31 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,859
Thanks Given to Others: 1,895
Thanked 1,568 Times in 1,020 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11569
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Please Help!

Can any stats experts help out on this one?

I know it's dated, but I'm trying to catch up on some zero reply threads.
Reply With Quote
Sponsored Links

  #3  
Old 20th December 2004, 10:36 AM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 53
 
Posts: 1,621
Thanks Given to Others: 1
Thanked 310 Times in 224 Posts
Karma Power: 186
Karma: 6526
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default

I confess to ignorance of the specific algorithm (until I read up on it). Graeme might be the best source for a good answer here.

Hershal
Reply With Quote
  #4  
Old 21st December 2004, 05:14 AM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,859
Thanks Given to Others: 1,895
Thanked 1,568 Times in 1,020 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11569
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

I tried a search for "Hilliard type 2 algorithm" (with and without the quotes) on just about every search engine and haven't found anything. I did find references to Cahn-Hilliard type equations but they are beyond me.
Reply With Quote
  #5  
Old 21st December 2004, 10:21 AM
rmccallum rmccallum is offline
Inactive Registered Visitor

Registration Date: Dec 2004
Location: Waterloo, Ontario, Canada
 
Posts: 6
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 21
Karma: 10
rmccallum has less than 100 Karma points so far.
Default Software

If you are willing to spend a few $$ there is software out there that will optimize your cal intervals for you. I would have to assume it is statistically significant. Correct me if I am wrong.
Reply With Quote
  #6  
Old 21st December 2004, 01:04 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,859
Thanks Given to Others: 1,895
Thanked 1,568 Times in 1,020 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11569
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

I've never seen any company go so far as to use something like a Hilliard type 2 algorithm (whatever that is). It's typically factors such as use frequency and environment, and results of previous calibrations.

Let's see - If we're going to analyze data to set calibration intervals, I'd look at 'as received' vs. 'as returned' if adjusted.

Anyone do data analysis and have a formula, or do most of you folks just look at the calibration results and 'gut feel' it?

This thread may be of interest: http://Elsmar.com/Forums/showthread.php?t=997
Reply With Quote
  #7  
Old 21st December 2004, 02:10 PM
rmccallum rmccallum is offline
Inactive Registered Visitor

Registration Date: Dec 2004
Location: Waterloo, Ontario, Canada
 
Posts: 6
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 21
Karma: 10
rmccallum has less than 100 Karma points so far.
Default Agreed

Currently we don't statistically analize to establish calibration frequency. We use the "gut feeling" method. Curious if this is acceptable for a ISO 17025 accredited automotive lab. The standard does call out uncertainty analysis. I just assumed this would be applied to calibration optimization. I did find the calculations on a google search to optimize calibration using fitted modelling and reliability but I don't think I have the time or willpower to program an excel spread sheet so I found a company that supplies the software to perform this function.

I am just starting out in pursuit of the accreditation. I have a pretty good foundation but a long ways to go. You will most likely see me hanging around over the next while. One question that may sound stupid but can I reference ISO 17025 requirements such as internal auditing or corrective action to our ISO 9000: 2001 system or do I need to rewrite all these?
Reply With Quote
  #8  
Old 21st December 2004, 04:45 PM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 53
 
Posts: 1,621
Thanks Given to Others: 1
Thanked 310 Times in 224 Posts
Karma Power: 186
Karma: 6526
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default

If we get away from the algorithm, then I recommend bouncing over to NCSLI RP (Recommended Practices) 1.

http://www.ncsli.org

It is well respected, established, and sound.

Some of the calibration software in the newer versions include cycle adjustment. Check with Cybermetrics and Blue Mountain.

Hershal
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks

Tags
adjustment, calibration frequency, calibration interval, hilliard type 2 algorithm


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Calibration of a frequency measurement device/frequency source Ruebenn ISO 17025 - Calibration, Measurement Gages and Test Laboratories 2 25th March 2009 01:48 AM
Adjustment of Calibration Intervals (Calibration Frequency) Jerry Eldred Calibration Frequency (Interval) 14 19th February 2009 12:00 PM
Safeguarded from Adjustment - Definition - Calibration - 7.6 d) r29728 Definitions, Acronyms, Abbreviations and Interpretations 9 15th November 2008 10:17 PM
Recommendations on calibration interval (frequency) ISO/IEC 17025:2005 para 5.10.4.4 joecalibratio Calibration Frequency (Interval) 5 23rd October 2005 11:14 PM
In-house calibration - Resource for calibration, adjustment of calipers, indicators janah214 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 2 10th July 2005 11:15 PM



The time now is 07:49 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts