Does anyone calibrate "Z" class ring gauges internally using a CMM ????
What other equipment should be used for calibrating rings.
"Z" class tolerance is +/- 0.0025mm (I believe)
My understanding is that you should be able to measure to within a tenth of the tolerance allowed for the item being measured. so take your tolerance on the rings divide it by ten and see if your CMM is at least that accurate. It has been awhile but I believe this is the case. Can anyone validate what I am saying or is the lid off of my white out again?
Typically the 10:1 ratio is geared towards gage tolerance vs. part tolerance. For measuring gages, a 4:1 ratio between working gage and master is more the standard. Of course, if you can manage 10:1, that's even better.
We have a CMM that is accurate to only +/-.0051mm, so it would not be appropriate at all for measuring Z rings. (or any class, for that matter). For cylindrical rings, we use a universal measuring machine with some ID jaws that is accurate to about +/-.00025mm.
I would think that a simple bench mic would do the job. With only 15 rings calibrated every 5 years, I wouldn't invest too much $$ into it, unless you can find other items to calibrate with it as well (plain plugs, thread plugs, indicators). But if you just have the 15 rings calibrated every 5 years, I don't see why you wouldn't go to an accredited outside lab.
Keep in mind, too, that the .0025 is the Class Z tolerance only between .254mm and 20.95mm. The entire chart can be found
here [.pdf]