Is there a checklist which shows the deliverables required when a new design is from New Product Development to New Product Introduction ?
Hi Milas,Thank you everyone, I have quick follow-up question.
If we take a "typical" medical device development cycle to include the following phases (proof of concept - Engineering Validation - Design Validation - Production validation and commercial build.
During which of the above phases, does the following take place:
Bench test/functional test?
Animal testing?
Clinical Testing?
Regulatory approval?
Any other key milestones?
Thanks
Please can you elaborate on the sottware specfications which need to be transferred, is it just the software version or is there more to it than that ?Design transfer activities really depend on your internal processes, whether you are using a contract manufacturer, and other factors. Design transfer doesn't have to be a one-time thing. This list may help though:
Documentation Transferred to Production
Device Specifications
Software Specifications
Subassembly Specifications
Components, Raw Materials Specifications
Packaging Specifications
Labeling Specifications
Environment Specifications
Bill of Materials (BOM)
History Record (HR) Forms/Travelers
Manufacturing Work Instructions
QC Test Instructions and Forms
Tooling/Equipment Specifications
Specifications for Manufacturing Material
Design Transfer Tasks
Process Risk Analysis Performed and Documented
Equipment Installed and Qualified
Suppliers Evaluated and Approved
Product Monitoring Implemented
Process Monitoring Implemented
Supplier Monitoring Implemented
Completion, Approval, Release of DMR
Transfer Conclusion Documented
It is a requirements for producing the software for use in the medical device. It is everything that defines what it is and how it gets integrated to each system. The specific documentation can vary by company.Please can you elaborate on the software specifications which need to be transferred, is it just the software version or is there more to it than that ?