Hello,
Can anyone provide guidance on how to determine and document production equivalence prior to V&V. We are approaching our design freeze with a prototype ME assembly. This assembly uses small scale materials/methods (casting) to keep costs low until preliminary clinical trials and pre-testing have de-risked the device. Eventually we will proceed into a production run with injection molded parts but ideally we would start the pivotal clinical trials and V&V with the prototype before putting the money into molds. The injection molded design will be equivalent in form and function to the prototype but it will use different materials (with better mechanical specifications than the cast per TDS).
This seems like a reasonable plan when referencing the flow charts in FDA's guidance: Deciding When to Submit a 510(k) for a Change to an Existing Device, however it will likely take significant documentation and justification. My main concern is that we would get to the other side of clinicals/conformance testing to international standards and all the results would be invalidated becuase of the material change. I fully expect we would be required to repeat testing that would be effected by the material change but I am hoping we can isolate those tests instead of repeating clinicals or all of 60601 general/collateral/particular.
Please let me know your thoughts/experiences with production equivalence justifications.
Thanks!!
Can anyone provide guidance on how to determine and document production equivalence prior to V&V. We are approaching our design freeze with a prototype ME assembly. This assembly uses small scale materials/methods (casting) to keep costs low until preliminary clinical trials and pre-testing have de-risked the device. Eventually we will proceed into a production run with injection molded parts but ideally we would start the pivotal clinical trials and V&V with the prototype before putting the money into molds. The injection molded design will be equivalent in form and function to the prototype but it will use different materials (with better mechanical specifications than the cast per TDS).
This seems like a reasonable plan when referencing the flow charts in FDA's guidance: Deciding When to Submit a 510(k) for a Change to an Existing Device, however it will likely take significant documentation and justification. My main concern is that we would get to the other side of clinicals/conformance testing to international standards and all the results would be invalidated becuase of the material change. I fully expect we would be required to repeat testing that would be effected by the material change but I am hoping we can isolate those tests instead of repeating clinicals or all of 60601 general/collateral/particular.
Please let me know your thoughts/experiences with production equivalence justifications.
Thanks!!