Hi, I am working on a medical device being certified to IEC 60601-1-2 with two different models rated 120V (USA/Canada) and 100V (Japan). We would like to claim compliance with the IEC 61000-3-2 Harmonic Content sub-standard even though we are technically outside the 220V-240V rated voltage specified by 60601-1-2 section 7.2.1 and Table 2 note (b). Our NRTL is directly applying the limits in IEC 61000-3-2 Table 1 (Class A) which we do not meet...
My question - an online search turned up a user manual for a piece of harmonics test equipment (Yokogawa) containing the following equation for converting IEC 61000-3-2 limits for different mains voltages. Does anyone know the source of this conversion equation (e.g. an IEC / JIS / ANSI standard or otherwise) or other justification for using it?
Converted limit = ( IEC 61000-3-2 limit ) * 230 / ( Rated voltage of equipment )
I can't find much info on applying IEC 61000-3-2 to 120V equipment searching this forum or online, although it seems fairly common to do this...
My question - an online search turned up a user manual for a piece of harmonics test equipment (Yokogawa) containing the following equation for converting IEC 61000-3-2 limits for different mains voltages. Does anyone know the source of this conversion equation (e.g. an IEC / JIS / ANSI standard or otherwise) or other justification for using it?
Converted limit = ( IEC 61000-3-2 limit ) * 230 / ( Rated voltage of equipment )
I can't find much info on applying IEC 61000-3-2 to 120V equipment searching this forum or online, although it seems fairly common to do this...