Hello Elsmar,
We are developing smart beds for children with Autism and other special needs. We have determined based on similar products that we're a Class 1 510K exempt product. We're trying to determine what standards are applicable to our bed and specifically if IEC 60601 is.
Our beds have calming sensory inputs (lights, speakers, vibration) as well as a camera for remote monitoring by parents/caregivers. These are the only electrical components. In the future, we may have actuating lifts to raise the head/feet but not yet. All electrical devices are previously UL certified as standalone products.
I understand that the definition of 60601 is:
We aren't transferring energy to or from the patient or detecting energy.
I am unsure if we qualify as having "applied parts". Sub-clause 3.8 states that an APPLIED PART is “part of ME EQUIPMENT that in NORMAL USE necessarily comes into physical contact with the PATIENT for ME EQUIPMENT or an ME SYSTEM to perform its function.
Patients within the bed do not necessarily need to come into contact with the electronics for them to function. They could touch the face of the camera/light though. Currently, we do provide the option for the patient to click buttons on the light/speaker to turn them on/off but we could restrict access to this or specify in instructions that only parents/caregivers, not the patient, would be the ones that should access these.
Any help is appreciated.
Thank you,
Caleb
We are developing smart beds for children with Autism and other special needs. We have determined based on similar products that we're a Class 1 510K exempt product. We're trying to determine what standards are applicable to our bed and specifically if IEC 60601 is.
Our beds have calming sensory inputs (lights, speakers, vibration) as well as a camera for remote monitoring by parents/caregivers. These are the only electrical components. In the future, we may have actuating lifts to raise the head/feet but not yet. All electrical devices are previously UL certified as standalone products.
I understand that the definition of 60601 is:
3 * Terminology and definitions
For the purposes of this document, the following terms and definitions apply.
(…)
3.63
* MEDICAL ELECTRICAL EQUIPMENT
ME EQUIPMENT
electrical equipment having an APPLIED PART or
transferring energy to or from the PATIENT or
detecting such energy transfer to or from the PATIENT
and which is:
a) provided with not more than one connection to a particular SUPPLY MAINS; and
b) intended by its MANUFACTURER to be used:
1) in the diagnosis, treatment, or monitoring of a PATIENT; or
2) for compensation or alleviation of disease, injury or disability
We aren't transferring energy to or from the patient or detecting energy.
I am unsure if we qualify as having "applied parts". Sub-clause 3.8 states that an APPLIED PART is “part of ME EQUIPMENT that in NORMAL USE necessarily comes into physical contact with the PATIENT for ME EQUIPMENT or an ME SYSTEM to perform its function.
Patients within the bed do not necessarily need to come into contact with the electronics for them to function. They could touch the face of the camera/light though. Currently, we do provide the option for the patient to click buttons on the light/speaker to turn them on/off but we could restrict access to this or specify in instructions that only parents/caregivers, not the patient, would be the ones that should access these.
Any help is appreciated.
Thank you,
Caleb