ISO 9001 does not specify any required interval. So what you need to comply with is 7.1.5.1b ("[the resources provided] are maintained to ensure their continuing fitness for their purpose."), and use the risk-based thinking of 9001.
Typically, you might use a measurement device's interval as recommended by the manufacturer. However, if you can reasonably justify extending that interval, due to low usage or other reasons, then there is no restriction to that. It is a risk to extend an interval (imagine having a device calibrated only every 2 years and finding out that the product produced over the last 2 years needs to be recalled when you find out the measuring device was out-of-spec), so you might add a risk register entry for that interval change and then monitor effectiveness over the next 2 years (?) to see if there was any problems resulting from extending the interval. You can also cite the history of calibrations on that device to show that the device always came back from outside calibration with "AS FOUND" in spec, indicating that it historically never needed adjusting in the past.