Request Calibration Info

QAguy

Involved In Discussions
We have a 3ft long steel ruler which is 17025 calibrated. It gets very low use, so we set the calibration interval to 5 years. At the end of the 5 years we buy a new replacement ruler with a calibration certificate, and retire the old one.
We have a measuring device that we calibrated last year, however, we have only used this measuring device 5 times during the year, and the calibration cost is expensive. How can we extend the calibration period for this measuring device until next year in order to remain in compliance. Really appreciated for your reply.
 

Tagin

Trusted Information Resource
We have a measuring device that we calibrated last year, however, we have only used this measuring device 5 times during the year, and the calibration cost is expensive. How can we extend the calibration period for this measuring device until next year in order to remain in compliance. Really appreciated for your reply.

Compliance with what? ISO 9001 or some other standard?
 

Ninja

Looking for Reality
Trusted Information Resource
FWIW, we used the same (type of) device...mechanical meter for measuring feet for sale of plastic film on a core. (daily use)
We had ours calibrated every year for a while...and when it passed with no issues every time, we lengthened the cal period to 2yrs...then to 5yrs.

We measured hundreds of feet...I would not have trusted verification on a 3ft scale. I did not watch how innocal verified it, so I can't suggest much toward in-house calibration other than a tape measure and consideration of what accuracy is desired.
 

Tagin

Trusted Information Resource
We have a measuring device that we calibrated last year, however, we have only used this measuring device 5 times during the year, and the calibration cost is expensive. How can we extend the calibration period for this measuring device until next year in order to remain in compliance. Really appreciated for your reply.

Compliance to ISO 9001-2015.

ISO 9001 does not specify any required interval. So what you need to comply with is 7.1.5.1b ("[the resources provided] are maintained to ensure their continuing fitness for their purpose."), and use the risk-based thinking of 9001.

Typically, you might use a measurement device's interval as recommended by the manufacturer. However, if you can reasonably justify extending that interval, due to low usage or other reasons, then there is no restriction to that. It is a risk to extend an interval (imagine having a device calibrated only every 2 years and finding out that the product produced over the last 2 years needs to be recalled when you find out the measuring device was out-of-spec), so you might add a risk register entry for that interval change and then monitor effectiveness over the next 2 years (?) to see if there was any problems resulting from extending the interval. You can also cite the history of calibrations on that device to show that the device always came back from outside calibration with "AS FOUND" in spec, indicating that it historically never needed adjusting in the past.
 

QAguy

Involved In Discussions
ISO 9001 does not specify any required interval. So what you need to comply with is 7.1.5.1b ("[the resources provided] are maintained to ensure their continuing fitness for their purpose."), and use the risk-based thinking of 9001.

Typically, you might use a measurement device's interval as recommended by the manufacturer. However, if you can reasonably justify extending that interval, due to low usage or other reasons, then there is no restriction to that. It is a risk to extend an interval (imagine having a device calibrated only every 2 years and finding out that the product produced over the last 2 years needs to be recalled when you find out the measuring device was out-of-spec), so you might add a risk register entry for that interval change and then monitor effectiveness over the next 2 years (?) to see if there was any problems resulting from extending the interval. You can also cite the history of calibrations on that device to show that the device always came back from outside calibration with "AS FOUND" in spec, indicating that it historically never needed adjusting in the past.
Thank you! I fully agree. Great advice.
 