Hello guys, new to the forum!
I'm in the process of doing a validation plan for a Tray Sealing Process (PETG and Adhesive Coated Tyvek). I'm reviewing the requirements in ISO 11607-2: 2019, section 4.4.3 indicates "All test methods used to show conformity to this document shall be validated and documented by the laboratory performing the test. NOTE Annex B of ISO 11607-1 contains a list of test methods. Publication of a method by a standards body does not make it validated in any laboratory."
We plan to do Peel Strength Test per ASTM F88/F88M−21 and Visual Inspection per ASTM F1886/F1886M−16 as acceptance criteria.
- The way I read the standard it pretty much says that you have to perform TMV, referencing the standard is not enough, is this correct? Have you had experience in which a rationale was acceptable?
- Does anyone have experience performing TMV for Visual Inspection and Peel Strength any examples you can provide in sample size and defects generated?, I think I need quite a little bit of guidance on the Peel Strength as this is a destructive test.