Author Topic:   Design input
Monica Lewis
Forum Contributor

Posts: 19
From:Sweden
Registered: Oct 98

posted 13 January 1999 03:55 AM     Click Here to See the Profile for Monica Lewis   Click Here to Email Monica Lewis     Edit/Delete Message   Reply w/Quote
According to 820.30(c). The procedures shall include a mechanism for addressing incomplete, ambiguous, or conflicting requirements.

If one would write in the SOP that incomplete,ambiguous or conflicting requirements are resolved with those responsible for imposing these requirements would this satisfy an FDA inspector

IP: Logged

Don Winton
Forum Contributor

Posts: 498
From:Tullahoma, TN
Registered:

posted 13 January 1999 10:56 AM     Click Here to See the Profile for Don Winton   Click Here to Email Don Winton     Edit/Delete Message   Reply w/Quote
Monica,

quote:
If one would write in the SOP that incomplete, ambiguous or conflicting requirements are resolved with those responsible for imposing these requirements would this satisfy an FDA inspector

I would think so, but you should also consider the following in drafting your SOP.

From the FDA's Small Entity Compliance Guide:

quote:
The QS regulation requires that the input procedures shall address incomplete, ambiguous or conflicting requirements. Every reasonable effort should be made to collect all of the requirements from which designers can generate detailed design specifications that are clear, correct and complete.

Using the above, the FDA included the following guideline and question (among others) for FDA inspectors to follow in regards to ¤ 820.30(c).

Question 3. Review the process for resolving incomplete, ambiguous, or conflicting requirements. For the design reviewed, determine whether any incomplete, ambiguous, or conflicting requirements were not resolved per the manufacturer's procedures.

See http://www.fda.gov/ora/inspect_ref/qsreg/dcrpgd.html for additional information.

From these two references, be certain that your SOP includes record maintenance methodology when resolving incomplete, ambiguous or conflicting requirements.

For consideration:

Incomplete, ambiguous or conflicting requirements are resolved with parties responsible for imposing these requirements and records are maintained in accordance with ¤820.30(j).

Regards,
Don

IP: Logged

All times are Eastern Standard Time (USA)

next newest topic | next oldest topic

Administrative Options: Close Topic | Archive/Move | Delete Topic
Post New Topic  Post A Reply
Hop to:

Contact Us | The Elsmar Cove Home Page

Your Input Into These Forums Is Appreciated! Thanks!


UBB 5.45c

Main Site Search
Y'All Come Back Now, Ya Hear?
Powered by FreeBSD!Made With A Mac!Powered by Apache!