Author Topic:   Verification Activities
Mark Smith
Forum Contributor

Posts: 32
From:
Registered: May 99

posted 11 April 2000 11:12 AM     Click Here to See the Profile for Mark Smith     Edit/Delete Message   Reply w/Quote
My company is in the process of moving to a new location and expecting to resume manufacturing in 2 months time. We are a medical device manufacturer that is currently developing devices and manufacturing for clinical evaluation only!!. We have several designs that we have performed verification testing on however, clinical evaluations intended for DESIGN VALIDATION are ongoing and are scheduled for at least 1 more year.
Since validation FOLLOWS successful verification, the manufacturing process as well as equipment used to manufacture the devices will not be validated until we are certain of the final design (AFTER we move to our new location and prior to distributing any approved product)
When we did verification testing at our current location, we required a number of pass / fail tests on 60 devices and allowed zero failures as evidence of 95% confidence/ 95%Reliability for the design.

I am aware that processes and equipment must be re-validated as the result of a relocation but do I need to repeat the design verification testing of these devices?
If so,
Will I need to repeat the sample size of 60 devices?
If not,
Can I significantly reduce the number of units to something around 10 devices?

IP: Logged

Don Winton
Forum Contributor

Posts: 498
From:Tullahoma, TN
Registered:

posted 13 April 2000 09:10 AM     Click Here to See the Profile for Don Winton   Click Here to Email Don Winton     Edit/Delete Message   Reply w/Quote
I am aware that processes and equipment must be re-validated as the result of a relocation but do I need to repeat the design verification testing of these devices?

I would think that the design verification would not have to be if you can show that the process re-validation was satisfactory. Meanwhile, check out the links below for additional info:

http://www.fda.gov/cdrh/comp/designgd.html

http://www.fda.gov/cdrh/comp/designgd.pdf

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfTopic/topicindex/topindx.cfm?alpha=d

IP: Logged

All times are Eastern Standard Time (USA)

next newest topic | next oldest topic

Administrative Options: Close Topic | Archive/Move | Delete Topic
Post New Topic  Post A Reply
Hop to:

Contact Us | The Elsmar Cove Home Page

Your Input Into These Forums Is Appreciated! Thanks!


UBB 5.45c

Main Site Search
Y'All Come Back Now, Ya Hear?
Powered by FreeBSD!Made With A Mac!Powered by Apache!