Design Input/Output and Design V&V (Verification and Validation) Interpretations
UL - Underwriters Laboratories - Health Sciences
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Design Input/Output and Design V&V (Verification and Validation) Interpretations
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Design Input/Output and Design V&V (Verification and Validation) Interpretations - Page 2


Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
definitions, design and development, design inputs and outputs, design validation, design verification, fda (food and drug administration), fda requirements, requirements and specifications, medical devices
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #9  
Old 2nd January 2014, 06:45 PM
deethreepio

 
 
Total Posts: 5
Re: Design Input/Output and Design V&V (Verification and Validation) Interpretations

I am dealing with a little different situation. It's a little off subject but I was hoping to get your opinion on the following:
If design validation is proving that a representative device meets the user requirements using objective evidence, and process validation is proving that a process can reliable create that same device, should IQ/OQ activities for process validation occur prior to design validation?

Thank you

Sponsored Links
  Post Number #10  
Old 2nd January 2014, 07:30 PM
Julie O

 
 
Total Posts: 221
Let Me Help You Re: Design Input/Output and Design V&V (Verification and Validation) Interpretations

D3, I just sent you a friend request. I'm thinking this will allow us to communicate offline, which is probably a better way to help you with this question.
  Post Number #11  
Old 2nd January 2014, 07:36 PM
deethreepio

 
 
Total Posts: 5
Re: Design Input/Output and Design V&V (Verification and Validation) Interpretations

Hi Julie,

I need more posts to respond to private messages. I will work on it.

Thanks
  Post Number #12  
Old 3rd January 2014, 02:19 AM
Ronen E

 
 
Total Posts: 3,736
Re: Design Input/Output and Design V&V (Verification and Validation) Interpretations

Quote:
In Reply to Parent Post by deethreepio View Post

I am dealing with a little different situation. It's a little off subject but I was hoping to get your opinion on the following:
If design validation is proving that a representative device meets the user requirements using objective evidence, and process validation is proving that a process can reliable create that same device, should IQ/OQ activities for process validation occur prior to design validation?

Thank you
Hello and welcome to the Cove

Design Validation and Process Validation are related but not directly / rigidly connected. For Design Validation it is important that you use devices built on the commercial production line (full scale or pilot), or at least closely representative of such devices (strong documented rationale must be in place), and that are proven to be within the finalised device specification for that point in time. Whether or not the processes were validated prior to that is usually a secondary question. Pay attention though that for critical processes where the outcome can't be practically verified (sterilization for instance), you'd have no choice but to validate the process first.

Cheers,
Ronen.
Thanks to Ronen E for your informative Post and/or Attachment!
  Post Number #13  
Old 3rd January 2014, 10:50 AM
deethreepio

 
 
Total Posts: 5
Re: Design Input/Output and Design V&V (Verification and Validation) Interpretations

Thanks for the response. If your statement holds true for sterilization, then it would also be necessary to perform full process validation on other processes that cannot be practically verified (injection molding and welds) prior to the design validation, correct?
  Post Number #14  
Old 3rd January 2014, 03:46 PM
dblnkl

 
 
Total Posts: 37
Re: Design Input/Output and Design V&V (Verification and Validation) Interpretations

Here is a MUCH MUCH abbreviated example of the concepts addressed by the primary questioner relating to design controls. In practice, for such a complex product below, there may be many branches and levels of design inputs and specifications from one set of user requirements.


Sample Product: Steerable ablation catheter


User requirement: (testable during product validation/clinical trials)
Physician must be able use the catheter to deliver an ablation to the medial wall of the left atrium via venous access and transeptal puncture.
(One of many potential user requirements for this product.)


Design Input: (testable during verification/animal studies)
Catheter must be long enough to reach heart from femoral venous entry and have length and flexibility sufficient to move freely within all chambers of the heart.
Distal segment curvature must be controlled by operator.
Distal catheter segment much be able to enter right atrium via IVC and cross septum via septal puncture into left atrium (steerable distal segment must have length no less than x).
Curvature of distal segment must permit catheter to be able to bend catheter tip back onto catheter body with a radius no greater than x, while remaining in the same plane.
Etc.

Design Specifications: (testable during verification/bench studies)
Reliability = x operations without failure
Catheter body material, diameter, thickness, length = w, x, y, z
Catheter distal segment material, diameter, thickness, length = w, x, y, z
Pull wire material, size, length = x, y, z
Pull wire attachment = x
Material bonding = x
Handle = x
Etc.
  Post Number #15  
Old 3rd January 2014, 05:45 PM
Ronen E

 
 
Total Posts: 3,736
Re: Design Input/Output and Design V&V (Verification and Validation) Interpretations

Quote:
In Reply to Parent Post by deethreepio View Post

Thanks for the response. If your statement holds true for sterilization, then it would also be necessary to perform full process validation on other processes that cannot be practically verified (injection molding and welds) prior to the design validation, correct?
It might be so. The answer depends on the criticality of the output you can't verify (assuming it is the case), to the ability of the device as a whole to satisfy user/patient needs. WRT sterility it's obvious. WRT welding, it depends on the weld's role. WRT injection molding, some outcomes are hard / impossible to practically verify, but others - like dimensions - should be easy; so again, it comes down to the part's role.
Thanks to Ronen E for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Design Input vs. Design Output - What constitutes Design Input? tantan Design and Development of Products and Processes 11 2nd January 2014 08:11 PM
Design Input & Output in Stages - Overkill? mr.mike ISO 13485:2016 - Medical Device Quality Management Systems 16 13th May 2011 01:08 AM
Are some Design Input elements part of the Design Output? sagai 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2 12th October 2010 01:26 AM
Is Prototyping Design Output or Design Verification? dyeysi Design and Development of Products and Processes 12 1st April 2008 01:05 PM