Michelle,
For implantable devices, usually the 510(k) submission is for the system, which includes the implant, class II device and the accessories, which might be class I or II depends on the classification, and depends on it are subject on the 510(k) submission or otherwise exempt from 510(k) submission. In your product description, including the 510(k) summary, all the parts of the system should be mentioned, while the completed list of sizes, diameters etc. should be only provided for the devices subject to the 510(k) clearance.
Moreover, as you can see in the recent FDA guidance -
Refuse to Accept Policy for 510(k)shttp://www.fda.gov/downloads/Medica...onandGuidance/GuidanceDocuments/UCM315014.pdf
para. 13 states: If device is intended to be marketed with multiple components, accessories, and/or as part of a system,
Select “N/A” if the device is not intended to be marketed with multiple components, accessories, and/or as part of a system.
a. Submission includes a
list of all components and accessories to be marketed with the subject device.
b. Submission includes a description (as detailed in item 11.a. and b. and 12 above) of each component or accessory.
Select “N/A” if the component(s)/accessory(ies) has been previously cleared, or is exempt, and the proposed
indications for use are consistent with the cleared indications
As for the representative drawing for the components:
Submission contains representative engineering drawing(s), schematics,illustrations and/or figures of the device that are clear, legible, labeled, and include dimensions.
In lieu of drawings, schematics, etc. of each device to be marketed, ”representative” drawings, etc. may be provided, where
“representative” is intended to mean that the drawings, etc. provided capture the differences in design, size, and other important
characteristics of the various models, sizes, or versions of the device(s) to be marketed.
Select “N/A” if the submitter provided a rationale for why the submission does not contain engineering drawings, schematics, etc. (e.g., device is a reagent and figures are not pertinent to describe the device).
As you can see, if the full list of the components is not provided, this might lead to the refuse to accept 510(k) decision, so I would say that it is rather critical to provide as full list of the components subject to 510(k) clearance as possible.
Good Luck,
Yana