The Cove Business Standards Discussion Forums
Monitor the Elsmar Forum
Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology

Single Post View
 
Old 13th February 2018, 02:40 PM
mlee97's Avatar
mlee97

 
 
Total Posts: 32
Re: Producing IATF 16949 compliant calibration certificates

Quote:
In Reply to Parent Post by Golfman25 View Post

You are going to have a real problem here. In general outside calibration labs must be 17025 certified -- so most customer are looking for the cert with that designation to file away and "show the auditor" when it comes up. Simple.

..

Your "out" is in the second part of 7.1.5.3.2 which says: "Calibration services may be performed by the equipment manufacturer when a qualified laboratory is not available for a given piece of equipment. In such cases the organization shall ensure that the requirements listed in Section 7.1.5.3.1 have been met." 7.1.5.3.1 is the requirements for internal labs -- scope and such. You'll need to figure out how to get yourself into that exception. Good luck.
The outside calibrator is the manufacturer of the test unit and they are certified to the 17025 lab standard. Their software actually produces page two and three of the certificate package, but page one, the actual cert, needs to be dressed up to Name the calibration lab and place the serials and cal data as I see it.

But other than that we should be OK, right?

Last edited by mlee97; 13th February 2018 at 02:42 PM. Reason: skipped word