Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
UL - Underwriters Laboratories - Health Sciences
Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
class i medical device, field service, recalls (general), software (general topics), software medical devices
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 15th March 2017, 11:25 AM
blah01

 
 
Total Posts: 34
Question Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?

We produce a standalone software considered as a Class I medical device under this MDR. During our last HC inspection, the inspector told us that all field software upgrades should be considered a ‘recall’ based on point ‘b’ of the definition which reads as follows:
(b) may fail to conform to any claim made by the manufacturer or importer relating to its effectiveness, benefits, performance characteristics or safety;

Like any software product, we occasionally have field upgrades to improve functionality or slight anomalies detected over time. However, being a Class I device, these do not pose any risks to users. We therefore find it to be an extreme interpretation that any field upgrades have to be reported per the Recall requirements (sections 63-65).

I’m hoping to get feedback regarding interpretation of the Recall definition and requirements in the context of software upgrades.

Note that I browsed through several threads in the ‘Canada Medical Device Regulations’ forum already and couldn’t find anything about Recall. If there is such a thread I’d appreciate a link to it.

Thanks in advance.

Sponsored Links
  Post Number #2  
Old 16th March 2017, 07:12 AM
shimonv

 
 
Total Posts: 255
Re: Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?

Hi blah01,
Let's zoom out for a second. This is what the HC guidance says:

================\\
"Recall" is defined in the MDR as follows:

A "recall" in respect of a medical device that has been sold, means any action taken by the manufacturer, importer or distributor of the device to recall or correct the device, or to notify its owners and users of its defectiveness or potential defectiveness, after becoming aware that the device:

(a) may be hazardous to health
(b) may fail to conform to any claim made by the manufacturer or importer relating to its effectiveness, benefits, performance characteristics or safety
(c) may not meet the requirements of the Food and Drugs Act or these regulations

A recall may include:

- removing a medical device from the market
- instructing customers to stop using a medical device and destroying remaining units in stock
- doing an on-site correction of a medical device
- advising users of a device about a problem or potential problem
- supplying different labelling (which may include updates to instructions or manuals)"
=================\\.

The regulator doesn't want to be bothered by every little change that you do. If this was the case then they wouldn't be able to cope with the amount of recall notifications.

The examples in the guidance document shows that that recall is about changes with significant impact on safety or effectiveness.

The inspector was latching onto point "b". Please not that the word "claim". It's about the major features of the device that fulfill its intended use. It's not about small enhancements or minor fixes.
Of course, there is some grey area between enhancement, a fix with no impact on safety and effectiveness and a fix which comes in response to safety or effectiveness issue.
The way you justify your decision is through your documentation and risk analysis.

-Shimon

Last edited by shimonv; 16th March 2017 at 10:31 AM.
Thank You to shimonv for your informative Post and/or Attachment!
  Post Number #3  
Old 16th March 2017, 09:22 PM
mihzago

 
 
Total Posts: 189
Re: Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?

I concur with Shimonv's suggestion.
Maybe what the inspector was saying is that you should consider recall requirements whenever you make software changes.
What I do to address this, when I document the change, I include a brief paragraph with an assessment explaining why this change is not considered recall.
  Post Number #4  
Old 17th March 2017, 12:08 AM
Access2hc's Avatar
Access2hc

 
 
Total Posts: 26
Re: Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?

I concur with shimonv's opinions. just to focus a bit more on the justification of each bug fix or enhancement.

in the unlikely event that the software fails to meet (any part of) its intended use, and has an impact on safety/performance, perhaps a consultation with the regulator is required.

most events are due to nuisance software events, or errors/inefficiencies of the workflow that does not impact the intended function of the software, or that the user wants to give feedback about the usability of the software which again does not impact on the intended purpose or the safety/performance of the software. these are not reportable

hope it helps

Cheers,
Ee Bin
@Access2HC
  Post Number #5  
Old 17th March 2017, 02:58 PM
blah01

 
 
Total Posts: 34
Re: Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?

Thanks everyone for the useful feedback. This really helped clarify things.

All I've decided to do is make an amendment to our change review record, which already has a section to identify if a Recall is applicable or not, to now also capture the rationale why a Recall is not applicable. I am thinking of using the following as standard text going forward:
"Fixes in this release have no correlation to product claims, intended use or safety."

I'd appreciate feedback if you think I should tweak this statement, otherwise thanks again for the help.
Thanks to blah01 for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Recall - Definition of AS9100 Section 7.6 F for Recall ME Man Definitions, Acronyms, Abbreviations and Interpretations 15 7th May 2012 10:28 PM
Recall Documentation - Recall Submission vs. Recall Notification DaisyMae US Food and Drug Administration (FDA) 4 13th December 2010 05:33 PM
The Future of the Forum Software - vBulletin 4.0 - Fall 2009? Marc Forum News and General Information 1 27th August 2009 06:18 AM
Recall - Definition of Recall by the Canadian Mde dev reg, CMDR Rezzi Definitions, Acronyms, Abbreviations and Interpretations 7 2nd May 2006 02:22 PM
vBulletin Software Upgrades - 5 July 2004 Marc Old Forum News and Issues Threads Archive 1 5th July 2004 01:09 PM



The time now is 10:40 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.



Misc. Internal Links


NOTE: This forum uses "Cookies"

The Elsmar Cove is currently owned by the Marc Smith Retirement Trust and operated by Law-Nutz Wyoming©2018