Re: Do Software Upgrades Fall Under the Definition of Recall per SOR/98-202?
Hi blah01,
Let's zoom out for a second. This is what the HC guidance says:
================\\
"Recall" is defined in the MDR as follows:
A "recall" in respect of a medical device that has been sold, means any action taken by the manufacturer, importer or distributor of the device to recall or correct the device, or to notify its owners and users of its defectiveness or potential defectiveness, after becoming aware that the device:
(a) may be hazardous to health
(b) may fail to conform to any claim made by the manufacturer or importer relating to its effectiveness, benefits, performance characteristics or safety
(c) may not meet the requirements of the Food and Drugs Act or these regulations
A recall may include:
- removing a medical device from the market
- instructing customers to stop using a medical device and destroying remaining units in stock
- doing an on-site correction of a medical device
- advising users of a device about a problem or potential problem
- supplying different labelling (which may include updates to instructions or manuals)"
=================\\.
The regulator doesn't want to be bothered by every little change that you do. If this was the case then they wouldn't be able to cope with the amount of recall notifications.
The examples in the guidance document shows that that recall is about changes with significant impact on safety or effectiveness.
The inspector was latching onto point "b". Please not that the word "claim". It's about the major features of the device that fulfill its intended use. It's not about small enhancements or minor fixes.
Of course, there is some grey area between enhancement, a fix with no impact on safety and effectiveness and a fix which comes in response to safety or effectiveness issue.
The way you justify your decision is through your documentation and risk analysis.
-Shimon