As indicated it may depend on 'singe use' or 'reusable'.
Under Annex IX of MDD 93/42/EEC I believe a needle holder would be covered under rule 6 i.e. "All surgically invasive devices intended for transient use are in Class IIa unless the are:"
And here is the point:
"....... unless they are:
- reusable surgical instruments, in which case they are in Class I".
As a matter of interest look at section 3.10 of the attachment (i.e. needle holder reference) that comes with this supplier's CE Certificate.
We have surgical tips which are connected to an electrosurgical generator which will remove body fluids and cut, coagulate tissue which we have classified per MDD Rule 11 and as Class 2b device as we feel it is under the "potentially hazardous" category. Not sure if this is fitting for this device, any input would be appreciated.
Asking about a medical device's classification without sharing the detailed intended use (e.g. IFU / user manual) is like asking someone to drive a car blindfolded.
thanks for you guidence
My issue was as the needle is invasive but the holder is not I was a liitle con fussed as would it be classed as an accesories therefore 2a?
thanks again
thanks for you guidence
My issue was as the needle is invasive but the holder is not I was a liitle con fussed as would it be classed as an accesories therefore 2a?
thanks again
Late again Mr. Keene ... remember this - the classification is always based on the intended purpose. The intended use of the holder is different to that of the needle. Classify them both seperately.