The Cove Business Standards Discussion Forums More Free Files Forum Discussion Thread Post Attachments Listing Cove Discussion Forums Main Page
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
UL - Underwriters Laboratories - Health Sciences
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
percent of tolerance, measurement uncertainty (mu)
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 5.00 average. Display Modes
  Post Number #1  
Old 19th May 2008, 01:42 PM
Charles Wathen

 
 
Total Posts: 129
Please Help! Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance

Hi everyone,
I have a question regarding a practice that our Ireland facility has been doing with measurement uncertainty.

For example, they send out an instrument for calibration. It comes back stating that the tolerance was within, but it was right at the edge. The vendor also quotes a measurement uncertainty associated with the measurement. Our Ireland facility is taking the measurement uncertainty and adding that to the tolerance and then saying its out of tolerance.

I told them that the purpose of the measurement uncertainty is just that - you are not sure of the exact true value. You would need to perform some type of risk associated with the measurement to detemine if it would impact the process.

Thoughts?

Sponsored Links
  Post Number #2  
Old 19th May 2008, 02:44 PM
Jerry Eldred's Avatar
Jerry Eldred

 
 
Total Posts: 530
Re: Measurement uncertainty & tolerance

I won't take any side on the issue, but I can see some possible logic as I'll explain in an example below.

To keep things simple, I'll use "PERCENT OF TOLERANCE" (i.e.: 100% of tolerance is the tolerance limit, etc.).

EXAMPLE: If As Found condition on an instrument was near the tolerance limit (let's say 95% of tolerance), and the measurement uncertainty was (what worked out to be) 10% of that tolerance; that means the As Found (Received) condition could have ranged from 85% to 105% of Tolerance. I hope I'm making sense (and not wanting to ramble too long).

Therefore, an instrument that came back with Received condition In-Tolerance (with the above example) could be POSSIBLY construed as Out-Of-Tolerance. I've thought about this myself in some instances over the years.

However, I would also agree that In-Tolerance is In-Tolerance. I would (personally) not call the item above Out-of-Tolerance because per measurand on the certificate, it was not. But as in my example, if there was a need to be prudent, there may be justification for risk assessment in such borderline cases.
Thank You to Jerry Eldred for your informative Post and/or Attachment!
Sponsored Links

  Post Number #3  
Old 19th May 2008, 09:08 PM
jfgunn

 
 
Total Posts: 88
Re: Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance

Your basic question is as follows: What decision rule will you use when taking measurement of uncertainty into account when making pass fail statements?

As an ISO 17025 accredited calibration laboratory (accredited by L-A-B which is part of the ILAC MRA), we have to have a set of decision rules that we use if we make pass/fail statements.

There are two documents that I find helpful is describing how uncertainty should be taken into account.

ASME "B89.7.3.1 - 2001 Guidelines for Decision Rules: Considering Measurement Uncertainty Determining Conformance to Specifications". This document costs $30 from www.asme.org

ILAC "G8:1996 Guidelines on Assessment and Reporting of Compliance with Specification" this is available for free at
http://www.ilac.org/documents/g8_1996.pdf

Note that though the ILAC document is free, I find the examples in the ASME to be a little more descriptive.

Note that there are also facilities that use the opposite approach. They expand the acceptance window by a value equal to the uncertainty. Either approach is reall fine as long as the quality sytem defines what is to be done. It is all a matter of risk.
Thank You to jfgunn for your informative Post and/or Attachment!
  Post Number #4  
Old 20th May 2008, 09:35 AM
merrick65's Avatar
merrick65

 
 
Total Posts: 44
Re: Measurement uncertainty & tolerance

I will give an example similar to what Jerry posted. I sent a set of gage blocks out for calibration and was reviewing the certificate for errors when I noticed one of the gage blocks was out of tolerance. When I called the calibration source to ask why it wasn't flagged/marked they replied that due to their uncertainty they could not say the block failed and would not mark it as being out of tolerance. The calibration source is accredited and also the manufacturer of the blocks.

I realize this is basically the opposite of the above posted question but thought it was interesting because the same rational is being used in both cases.
Thanks to merrick65 for your informative Post and/or Attachment!
  Post Number #5  
Old 20th May 2008, 10:17 AM
CalRich's Avatar
CalRich

 
 
Total Posts: 126
I Say... Re: Measurement uncertainty & tolerance

Quote:
In Reply to Parent Post by merrick65 View Post

I will give an example similar to what Jerry posted. I sent a set of gage blocks out for calibration and was reviewing the certificate for errors when I noticed one of the gage blocks was out of tolerance. When I called the calibration source to ask why it wasn't flagged/marked they replied that due to their uncertainty they could not say the block failed and would not mark it as being out of tolerance. The calibration source is accredited and also the manufacturer of the blocks.

I realize this is basically the opposite of the above posted question but thought it was interesting because the same rational is being used in both cases.
That's right. The uncertainty is a bilateral value, and so in any particular example, you can say that the uncertainty brings it "in tolerance" or "out of tolerance". As jfgunn alluded to, labs have to determine their decision rule about how they use uncertainty to decide whether these results that are 'on the border' are handled. For those companies that contract a lot of accredited calibrations out, they should likewise have some kind of policy or method to handle results that are on the edge and measurement uncertainty. Essentially, the guy (or gal) who is checking in the equipment needs (and wants) some clearly defined way to know whether they can accept the calibration or not.
Thanks to CalRich for your informative Post and/or Attachment!
  Post Number #6  
Old 20th May 2008, 01:20 PM
lee01

 
 
Total Posts: 89
Re: Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance

Often get this one, though not in the Lab environment.

Have had so many discussions with Productive thinking employees who try to take on the basics of MSA and somehow use the percentage of error as extra tolerance, in order to get products out the door.

Yet in the same argument I have the inspectors saying that it should be used to reduce the design tolerance!!!

But I think to myself is it as clear cut as that? Is it more a grey area and not black and white (either add or subtract from the design tolerance). If it includes things like user/ environment etc etc, then how can it be simply add or subtract whatever the percentage from the design tolerance.

Iíll have to get hold of those standards you mentioned, like I said Iím not talking about a Lab environment but I would like to view the thought process of how you introduce the decisions


I could debate this one all day
Thanks to lee01 for your informative Post and/or Attachment!
  Post Number #7  
Old 22nd May 2008, 03:51 PM
BradM's Avatar
BradM

 
 
Total Posts: 5,950
Re: Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance

Also, if I'm not mistaken, uncertainty is reported within a confidence level. Thus, it's not a slam-dunk, 100% given the reported uncertainty is your uncertainty, but rather an estimated uncertainty.

BTW, if this instrument is so close, I believe I would be decreasing the calibration interval, suggesting a different tolerance, etc.

Saying, rather you report in/out, the more pressing issue to me is figuring out why it's doing it, and how to keep this issue from happening again.
  Post Number #8  
Old 22nd May 2008, 10:39 PM
Hershal's Avatar
Hershal

 
 
Total Posts: 2,486
Re: Measurement Uncertainty & Tolerance - Adding Uncertainty to the Tolerance

Quote:
In Reply to Parent Post by Charles Wathen View Post

Hi everyone,
I have a question regarding a practice that our Ireland facility has been doing with measurement uncertainty.

For example, they send out an instrument for calibration. It comes back stating that the tolerance was within, but it was right at the edge. The vendor also quotes a measurement uncertainty associated with the measurement. Our Ireland facility is taking the measurement uncertainty and adding that to the tolerance and then saying its out of tolerance.

I told them that the purpose of the measurement uncertainty is just that - you are not sure of the exact true value. You would need to perform some type of risk associated with the measurement to detemine if it would impact the process.

Thoughts?
In a pure sense that actually is what MU is for, but it is also a trap.....

A trap because it may suggest that you are OOT when in fact you may not be, but because of the MU, you can't prove it either way.....the other side of that coin, you can't prove it is OOT either.....it casts doubt on the result.....

The obvious easy fix is bring the tolerance closer in, and sometimes there is value in that, but it is not a universal fix.....

One other potential that is now starting to come into vogue as a result of ANSI/NCSL Z540.3-2006 is guardbanding to get a more true picture of the effect of MU and its effect on tolerance.....and I suspect it will be debated till the cows come home, but it may provide one avenue to help your Irish colleagues.....

However, remember they do not work under the American National Standard.....

They are taking the more conservative and safer route.....the best thing they can do is if their stuff can be adjusted, to have it adjusted to nominal.....not just in tolerance.....

Hope this helps.
Thank You to Hershal for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Maximum acceptable Tolerance - Uncertainty Ratio (TUR) casviq General Measurement Device and Calibration Topics 2 26th August 2013 12:38 PM
Test Uncertainty Ratio TUR (UUT Tolerance / STD Uncertainty) Question jfgunn General Measurement Device and Calibration Topics 8 15th February 2013 03:26 PM
Uncertainty of Measurement vs. Tolerance TitaniumMan425 Measurement Uncertainty (MU) 5 6th February 2013 06:18 PM
Tolerance and Uncertainty have the same Symbol - How to Calculate the Tolerance charlie79 Measurement Uncertainty (MU) 1 14th June 2012 10:58 AM
In-tolerance Probability - Level of Confidence - Uncertainty and Proficiency Marc Measurement Uncertainty (MU) 0 20th October 1999 07:08 PM



The time now is 08:48 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


 
 
 


NOTE: This forum uses "Cookies"