Starting up a new Calibration Laboratory
UL - Underwriters Laboratories - Health Sciences
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Starting up a new Calibration Laboratory
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Starting up a new Calibration Laboratory


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
calibration (metrology) laboratories
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 5.00 average. Display Modes
  Post Number #1  
Old 20th May 2009, 03:40 PM
SGCalLab

 
 
Total Posts: 21
Please Help! Starting up a new Calibration Laboratory

Hello everyone,

I just got hired on to a company and have been tasked with creating a calibration lab from the ground up. They have not had a need for a real cal lab before, but some customers and other regulations are pushing them to create one and start using it. It will need to be FDA compliant.

From what I know right now (only my third day) Iíll be calibrating lotís of temperature indicators, thermocouples (J/K), RTD, pressure gauges (air/liquid), vacuum gauges, RPM (converting Hz to RPM), compound pressure sensors, etcÖ

Iíve been tasked to buy equipment to calibrate with. Mindful, I have to keep it in reason.

Some of the equipment I was looking at are:
  • Process Calibrators (Fluke 743b or 725)
  • Temperature Calibrators (Fluke 724)
  • Pressure (Fluke 718)
  • Loop Calibrator (Fluke 707)
  • Power supply
  • RPM Calibrator
  • Field (portable) Metrology Well (for thermocouples)
These donít have to be all Fluke. I am familiar with Fluke more than some others like Omega for instance, but I donít have to stick with Fluke if others are better.

Looking for suggestions on what equipment to buy.

Thanks!!

Sponsored Links
  Post Number #2  
Old 20th May 2009, 03:57 PM
Coury Ferguson's Avatar
Coury Ferguson

 
 
Total Posts: 4,557
Re: Starting up a new Cal Lab

Quote:
In Reply to Parent Post by SGCalLab View Post

Hello everyone,

I just got hired on to a company and have been tasked with creating a calibration lab from the ground up. They have not had a need for a real cal lab before, but some customers and other regulations are pushing them to create one and start using it. It will need to be FDA compliant.

From what I know right now (only my third day) I’ll be calibrating lot’s of temperature indicators, thermocouples (J/K), RTD, pressure gauges (air/liquid), vacuum gauges, RPM (converting Hz to RPM), compound pressure sensors, etc…

I’ve been tasked to buy equipment to calibrate with. Mindful, I have to keep it in reason.

Some of the equipment I was looking at are:
  • Process Calibrators (Fluke 743b or 725)
  • Temperature Calibrators (Fluke 724)
  • Pressure (Fluke 718)
  • Loop Calibrator (Fluke 707)
  • Power supply
  • RPM Calibrator
  • Field (portable) Metrology Well (for thermocouples)
These don’t have to be all Fluke. I am familiar with Fluke more than some others like Omega for instance, but I don’t have to stick with Fluke if others are better.

Looking for suggestions on what equipment to buy.

Thanks!!

Here is a Manual that I found while looking for something similar. It is the NIST's and it is a template. I didn't see any "Copyright" statement on the Manual. I have split it into two parts, because the file size exceeds the upload limit.

Maybe it may help you.
Attached Files: 1. Scan for viruses before opening, 2. Please report any 'bad' files by Reporting this post, 3. Use at your Own Risk.
File Type: pdf Pages from ISO17025 Manual-Pages 1-55.pdf (2.80 MB, 1856 views)
File Type: pdf Pages from ISO17025 Manual-Pages 56-110.pdf (1.85 MB, 1313 views)
Thank You to Coury Ferguson for your informative Post and/or Attachment!
Sponsored Links

  Post Number #3  
Old 20th May 2009, 04:02 PM
Jeff Frost

 
 
Total Posts: 646
Re: Starting up a new Calibration Laboratory

Quote:
In Reply to Parent Post by SGCalLab View Post

Hello everyone,

I just got hired on to a company and have been tasked with creating a calibration lab from the ground up. They have not had a need for a real cal lab before, but some customers and other regulations are pushing them to create one and start using it. It will need to be FDA compliant.

From what I know right now (only my third day) I’ll be calibrating lot’s of temperature indicators, thermocouples (J/K), RTD, pressure gauges (air/liquid), vacuum gauges, RPM (converting Hz to RPM), compound pressure sensors, etc…

I’ve been tasked to buy equipment to calibrate with. Mindful, I have to keep it in reason.

Some of the equipment I was looking at are:
  • Process Calibrators (Fluke 743b or 725)
  • Temperature Calibrators (Fluke 724)
  • Pressure (Fluke 718)
  • Loop Calibrator (Fluke 707)
  • Power supply
  • RPM Calibrator
  • Field (portable) Metrology Well (for thermocouples)
These don’t have to be all Fluke. I am familiar with Fluke more than some others like Omega for instance, but I don’t have to stick with Fluke if others are better.

Looking for suggestions on what equipment to buy.

Thanks!!
What is your background in calibration?

Before you purchase calibration equipment you will need to review process equipment to understand the equipment accuracy you will be calibrating and what accuracy the actual process requires.

Since you are working with process equipment that contain multiple elements such as display modules and thermal couples you will also need to determine total system accuracy which is not the adding together of the display unit accuracy and thermal couple accuracy.

Calibration equipment itself should have a system accuracy at a minimum 4:1 to a preferred 10:1 ratio to the unit under test with calibration certification reflecting actual results traceable to a national or international standard body.

When calibrating pressure gauges the actual calibration standard should have a 1/4 of 1% accuracy over the most accurate process gauge using nitrogen as a calibration media.

Last edited by Jeff Frost; 21st May 2009 at 10:22 AM. Reason: Rewrote last sentence to improve clarity
Thanks to Jeff Frost for your informative Post and/or Attachment!
  Post Number #4  
Old 22nd May 2009, 09:14 AM
SGCalLab

 
 
Total Posts: 21
Re: Starting up a new Calibration Laboratory

Quote:
In Reply to Parent Post by Jeff Frost View Post

What is your background in calibration?

Before you purchase calibration equipment you will need to review process equipment to understand the equipment accuracy you will be calibrating and what accuracy the actual process requires.

Since you are working with process equipment that contain multiple elements such as display modules and thermal couples you will also need to determine total system accuracy which is not the adding together of the display unit accuracy and thermal couple accuracy.

Calibration equipment itself should have a system accuracy at a minimum 4:1 to a preferred 10:1 ratio to the unit under test with calibration certification reflecting actual results traceable to a national or international standard body.

When calibrating pressure gauges the actual calibration standard should have a 1/4 of 1% accuracy over the most accurate process gauge using nitrogen as a calibration media.
I had calibration lab experience and training in the military. I'm new to this type of enviroment.

Most of what I have found out thus far, is the the tolerances on the process is between 1 - 3%.
  Post Number #5  
Old 22nd May 2009, 10:00 AM
BradM's Avatar
BradM

 
 
Total Posts: 5,956
Re: Starting up a new Calibration Laboratory

Look at my response in your other thread. Before you buy anything...

You need to make a thorough (and accurate) inventory of the equipment. The labor you will need to do the work is going to need to be sold to management. Also, you will need to evaluate vendors, audit them, etc. Thus, you need to determine what is before you.

As far as standards, look at buying the standards that are critical for your process. Or more simply put, get the most bang for your buck.

Starting off, you may want to qualify some vendors, and have them calibrate the equipment for you. It's going to take you some time to get things going. So assure the equipment is calibrated and tagged properly first.

Then, you can work on setting up your calibration system.
  Post Number #6  
Old 22nd May 2009, 05:57 PM
Hershal's Avatar
Hershal

 
 
Total Posts: 2,498
Re: Starting up a new Calibration Laboratory

Step number one is to join NCSLI www.ncsli.org and get the Recommended Practices (RPs). There is a RP specifically for how to set up a cal lab.

Make sure you control both temp and RH.

Also, get a copy of ANS/ISO/IEC 17025:2005 whether you intend to get accreditation or not as it will help develop a good management structure to maintain control and provide services. If you intend to seek accreditation don't just pick any, shop around a bit first.

If your parent organization has Government contracts, join GIDEP www.gidep.org to have access to military calibration porocedures and avoid validation issues.

Get training in the specific types of calibration you will be doing and make sure it is documented. There are NIST sessions and similar at NCSLI in San Antonio end of July.

Hope this helps.
Thank You to Hershal for your informative Post and/or Attachment!
  Post Number #7  
Old 26th May 2009, 02:34 PM
SGCalLab

 
 
Total Posts: 21
Re: Starting up a new Calibration Laboratory

Hershal, what's the difference between NCSLI and ANSI? (as for becoming members of).
  Post Number #8  
Old 27th May 2009, 10:45 PM
Hershal's Avatar
Hershal

 
 
Total Posts: 2,498
Re: Starting up a new Calibration Laboratory

NCSLI is a membership organization of Metrology professionals.....for calibration, this is the better organization.....
Thanks to Hershal for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Starting ISO 17025 Accreditation from scratch - Petroleum Laboratory mtak33 ISO 17025 related Discussions 3 23rd February 2013 01:38 PM
Starting a new Calibration and Repair Laboratory kd818 General Measurement Device and Calibration Topics 15 31st July 2011 01:18 PM
Starting an ISO 17025 Accredited Laboratory kat.k. ISO 17025 related Discussions 15 10th February 2011 01:21 AM
Starting a Calibration Laboratory in State College, PA fortyfour73 Career and Occupation Discussions 4 24th February 2007 05:41 PM
Internal Laboratory - Defining the Laboratory Scope - Inspection, Test, Calibration DJN Inspection, Prints (Drawings), Testing, Sampling and Related Topics 18 7th March 2005 12:20 PM



The time now is 02:45 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


 
 
 


NOTE: This forum uses "Cookies"