Flow Meter Calibration - Outside Calibration Service vs. In-House Calibration

Monitor the Elsmar Forum
Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
in-house calibration, outsourced calibration, traceability (misc), traceability of calibration (measurement) standards, calibration (general topics)
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 1st November 2011, 11:42 AM
CDISO

 
 
Total Posts: n/a
Question Flow Meter Calibration - Outside Calibration Service vs. In-House Calibration

Recently, it was suggested that we start calibrating our own Flow-meters. We have 22 of them. The suggest-er felt that it isn't difficult, would save money, etc. The suggest-ee was not comfortable with the idea and said so.

I tend to agree with the suggest-ee because 1) At our ISO audit last May, the auditor wanted to see Calibration records for a piece of equipment. I had the work order showing that a Calibration was completed by an outside service instantly available, but he stated that he needed to see something with Traceability information on it. (We had that in the paper files). 2) If we need to provide Traceability information, we would need more than a calibrated bucket and a stopwatch.

I would like some guidance as to whether in-house calibration of flow-meters is compliant with the ISO 9001 standard, without purchasing calibration lab-quality equipment.

Thanks!

Sponsored Links
  Post Number #2  
Old 1st November 2011, 04:30 PM
htquality

 
 
Total Posts: 43
Re: Flow Meter Calibration - Outside Calibration Service vs. In-House Calibration

I believe that if you approach the calibration properly, it can be done in-house with appropriate traceability available. You would need to use calibation equipment (another flow meter) that was calibrated with full traceability as well as appropriate tolerance that is higher (4:1 I believe) than the flow meters that you want to calibrate. So you might have to purchase a "better" flow meter, with traceable calibration, and use it to calibeate your other 22 meters in-house.
  Post Number #3  
Old 1st November 2011, 06:38 PM
BradM's Avatar
BradM

 
 
Total Posts: 6,031
Re: Flow Meter Calibration - Outside Calibration Service vs. In-House Calibration

Quote:
In Reply to Parent Post by CDISO View Post

Recently, it was suggested that we start calibrating our own Flow-meters. We have 22 of them. The suggest-er felt that it isn't difficult, would save money, etc. The suggest-ee was not comfortable with the idea and said so.

I tend to agree with the suggest-ee because 1) At our ISO audit last May, the auditor wanted to see Calibration records for a piece of equipment. I had the work order showing that a Calibration was completed by an outside service instantly available, but he stated that he needed to see something with Traceability information on it. (We had that in the paper files). 2) If we need to provide Traceability information, we would need more than a calibrated bucket and a stopwatch.

I would like some guidance as to whether in-house calibration of flow-meters is compliant with the ISO 9001 standard, without purchasing calibration lab-quality equipment.

Thanks!
Good question!

First, understand there are a lot of variables when it comes to flow, to name a few:
  • Flow Rate
  • Temperature
  • Pressure
  • Restrictions (if any)
  • Allowing adequate distance before and after flow rate
  • Achieving steady state
So I'm thinking of three approaches. One, if it is liquid, you might could use a method of placing a bucket on a scale, use a timer, and verify (by weight dispensed per time) if the flow is accurate.

The second, is to build a "dummy" system. You can have a dummy flow system with a flow standard, and have a spot for the U.U.T. in the flow system. Finally, you can add a spot for the U.U.T. in the current process system to verify the flow meters.

You can add up the uncertainties to the bucket/stopwatch system and see if it is sufficient. Other than that, I don't see how you could get around not purchasing a standard flowmeter. other than.... creating a dummy flow system and having a spot for three or four meters at a time. Then you could develop a comparison method with min/max deviation. However, there is no N.I.S.T. traceability here.
  Post Number #4  
Old 1st November 2011, 08:00 PM
Hershal's Avatar
Hershal

 
 
Total Posts: 2,498
Re: Flow Meter Calibration - Outside Calibration Service vs. In-House Calibration

BradM has a good answer. One thing you will notice if you peruse cal lab scopes, is there are not that many that do flow. It is a smaller and somewhat more specialized discipline within metrology and most cal labs have trouble with it.

Maybe for calipers inside cal can be done with assurance. But for some disciplines like flow, get an outside accredited lab.
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Flow Meter calibration for mixed gases cadevon Calibration and Metrology Software and Hardware 1 31st May 2015 10:58 AM
In house Volt Meter Calibration medguy123 General Measurement Device and Calibration Topics 5 18th November 2013 01:53 AM
In-House pH Meter Calibration Bigfoot General Measurement Device and Calibration Topics 13 15th April 2010 10:03 AM
Soliciting recommendations for a fast turnaround calibration house - Mass Flow Meter ScottK Manufacturing and Related Processes 1 19th December 2008 10:52 AM
Calibration Outsourcing vs. In House Calibration - Outrageous Costs of Calibration Cathy Everett General Measurement Device and Calibration Topics 28 5th September 2006 12:50 AM



The time now is 04:43 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.



Misc. Internal Links


NOTE: This forum uses "Cookies"