ISO/TS 16949 Compliant Calibration Method - Laboratory Calibration Levels

Monitor the Elsmar Forum
Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
iso 17025 - testing and calibration laboratories, ts 16949 - automotive quality system standard, 7.6.3.2 - external and commercial and independent labs, calibration (metrology) laboratories, calibration (general topics), a2la (american association for laboratory accreditation)
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 23rd August 2007, 11:34 AM
Icy Mountain's Avatar
Icy Mountain

 
 
Total Posts: 902
Please Help! ISO/TS 16949 Compliant Calibration Method - Laboratory Calibration Levels

I have a quandary that I have never really considered until now. I am requesting quotes for suppling calibration services for my quality system. Only calibration labs that can demonstrate a scope that is accredited by A2LA to ISO/IEC 17025 for all my calibration requirements are being considered. One of the laboratories has quoted three levels of calibration:
Quote:
In Reply to Parent Post by Joe's Calibration Shack

Standard Calibration - is traceable to national standards, reports whether the unit is in or out of tolerance, and reports data only on out of tolerance points. This is accepted throughout industry and meets the needs of many customers.

Calibration with before and after data - This calibration is performed with full data regardless if the instrument was out of tolerance or not. This is a more expensive calibration as the recording of data takes longer. Most FDA and regulated type industries require this level of calibration where product safety is a high concern.

A2LA accredited calibration with full data and measurement uncertainty - This calibration is performed with full data and will also include the uncertainty of the measurements. This calibration is accredited to ISO/IEC 17025. This is a more expensive calibration as the recording of data and the calculations of uncertainty takes longer. The electronic components industry is often required to have this level of calibration.
I believe that the Standard Calibration will allow me to remain compliant with TS16949 as Section 7.6.2 requires ISO/TS 16949 requires calibration records that contain:
"any out-of-specification readings as received for calibration/verification,
statements of conformity to specification after calibration/verification,
an assessment of the impact of out-of-specification condition."
Section 7.6.3.2 requires "the laboratory shall be accredited to ISO/IEC 17025 or national equivalent."

I am I correct in my belief or have I missed a memo, guidance, or implied language somewhere?

Sponsored Links
  Post Number #2  
Old 23rd August 2007, 11:45 AM
vandenbar

 
 
Total Posts: 23
Re: ISO/TS 16949 Compliant Calibration Method

Icy,

If you are using an external lab for your calibration (which from your post is the intention) then you will need option 3 form Joe's Calibration shack. The external lab needs to be accredited and the test/ calibration that they are doing for you needs to be on their scope of accreditation as well. The only exceptions are if the customer accepts a non accredited lab or if you can not find an accredited lab you can use teh equipment manufacturer.

That's my interpretation anyway.
  Post Number #3  
Old 23rd August 2007, 11:47 AM
Stijloor's Avatar
Stijloor

 
 
Total Posts: 15,339
Re: ISO/TS 16949 Compliant Calibration Method

Quote:
In Reply to Parent Post by Icy Mountain View Post

I have a quandary that I have never really considered until now. I am requesting quotes for suppling calibration services for my quality system. Only calibration labs that can demonstrate a scope that is accredited by A2LA to ISO/IEC 17025 for all my calibration requirements are being considered. One of the laboratories has quoted three levels of calibration:I believe that the Standard Calibration will allow me to remain compliant with TS16949 as Section 7.6.2 requires ISO/TS 16949 requires calibration records that contain:
"any out-of-specification readings as received for calibration/verification,
statements of conformity to specification after calibration/verification,
an assessment of the impact of out-of-specification condition."
Section 7.6.3.2 requires "the laboratory shall be accredited to ISO/IEC 17025 or national equivalent."

I am I correct in my belief or have I missed a memo, guidance, or implied language somewhere?
Hello Icy,

Quote: "an assessment of the impact of out-of-specification condition."

I don't think that Joe's Calibration Shack can provide you with that information.....

Stijloor.
  Post Number #4  
Old 23rd August 2007, 11:59 AM
AndyN's Avatar
AndyN

 
 
Total Posts: 9,029
Caution Re: ISO/TS 16949 Compliant Calibration Method

Icy:
To run an effective calibration system, you will need before and after data. That will help you decide when the equipment will need to be calibrated again. Furthermore, for you to be able to conduct any internal calibration or to perform an MSA study you will need the uncertainty statements from the cal. lab.

It's always tempting to go for the least expensive option, but you get what you pay for and in this case it's not much with the standard option........
  Post Number #5  
Old 23rd August 2007, 12:05 PM
Icy Mountain's Avatar
Icy Mountain

 
 
Total Posts: 902
Re: ISO/TS 16949 Compliant Calibration Method

Quote:
In Reply to Parent Post by vandenbar View Post

The external lab needs to be accredited and the test/ calibration that they are doing for you needs to be on their scope of accreditation as well.
The external lab is accredited and the test/calibrations they are doing are within their scope. Do I really need all that data? The calibration method, equipment, technician, methods, etc. don't change.
  Post Number #6  
Old 23rd August 2007, 12:06 PM
Icy Mountain's Avatar
Icy Mountain

 
 
Total Posts: 902
Re: ISO/TS 16949 Compliant Calibration Method

Quote:
In Reply to Parent Post by Stijloor View Post

Quote: "an assessment of the impact of out-of-specification condition."I don't think that Joe's Calibration Shack can provide you with that information.....
Yes, that's true. However, I can make that assessment from the data provided by their Standard Calibration, yes?
  Post Number #7  
Old 23rd August 2007, 12:13 PM
Icy Mountain's Avatar
Icy Mountain

 
 
Total Posts: 902
Re: ISO/TS 16949 Compliant Calibration Method

Quote:
In Reply to Parent Post by AndyN View Post

Icy:
To run an effective calibration system, you will need before and after data. That will help you decide when the equipment will need to be calibrated again. Furthermore, for you to be able to conduct any internal calibration or to perform an MSA study you will need the uncertainty statements from the cal. lab.
We perform testing but do not provide calibration or calibration certificates so I don't need uncertainties for that. As for MSA studies, I'm in a rather unusual situation in that I have had the requirement waived by my customer for every product that falls under the TS scope. Our volumes with these products are too low to generate any meaningful data. In addition, 120 VAC outputs are highly dependent, in both voltage and current, on how long that the unit has been providing current, amount of load, etc....they're sloppy. You know, 220, 221, whatever it takes.
  Post Number #8  
Old 23rd August 2007, 12:30 PM
AndyN's Avatar
AndyN

 
 
Total Posts: 9,029
Read This! Re: ISO/TS 16949 Compliant Calibration Method

Quote:
In Reply to Parent Post by Icy Mountain View Post

We perform testing but do not provide calibration or calibration certificates so I don't need uncertainties for that. As for MSA studies, I'm in a rather unusual situation in that I have had the requirement waived by my customer for every product that falls under the TS scope. Our volumes with these products are too low to generate any meaningful data. In addition, 120 VAC outputs are highly dependent, in both voltage and current, on how long that the unit has been providing current, amount of load, etc....they're sloppy. You know, 220, 221, whatever it takes.
Well, that helps - but not much! Does your customer really know what they're doing by waiving the requirements. It might be nice of them to do that, but I've also seen customers who sign off on all kinds of FMEA's, CP's etc that were wrong too.......

I'd also suggest that since you're testing you do need the uncertainties to know what contribution they make to the testing uncertainties.......if you want to run effectively. IMHO this is the area that customers often start to get lost in their understanding, but when things go wrong they'll be all over your case.......

And I stand by the fact that to run an effective calibration system you must use the cal. data to drive the recall date. Just saying do it once or twice a year isn't effective.....
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Auditing TS 16949 Cl. 7.6.3.1 - Internal Calibration Laboratory Requirements SandyStarkie IATF 16949 - Automotive Quality Systems Standard 8 29th June 2016 02:55 PM
TS 16949 Audit Minor non conformance - Outside Calibration Laboratory iandb IATF 16949 - Automotive Quality Systems Standard 15 15th April 2014 07:37 AM
ISO 17025 External Calibration Laboratory Requirements within ISO/TS 16949 ntrprizxile ISO 17025 related Discussions 28 7th November 2011 06:17 PM
Compliant Calibration vs. Accredited 17025 Calibration - 1000ml Graduated Beaker jmbhbf36 ISO 17025 related Discussions 16 15th February 2009 02:45 AM