The Cove Business Standards Discussion Forums More Free Files Forum Discussion Thread Post Attachments Listing Cove Discussion Forums Main Page
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
UL - Underwriters Laboratories - Health Sciences
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
iec 60601 - medical electrical equipment, ingress protection, medical device electronics and electrical equipment
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 6th March 2013, 01:55 PM
petruchio

 
 
Total Posts: 26
Please Help! IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirements

Hi

Sub-clause 8.3.1 of this standard states that MEE (Medical Electrical Equipment) shall provide protection to at least IP21 but sub-clause 7.2 seems to allow a get-out here by permitting marking the product with the umbrella/raindrops symbol if it fails to meet the water ingress requirements.

Does this get-out only apply to products with a carrying case or is it generally applicable?

petruchio

Sponsored Links
  Post Number #2  
Old 7th March 2013, 02:22 AM
Stijloor's Avatar
Stijloor

 
 
Total Posts: 15,219
Re: IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirement

A Quick Bump!

Can someone help?

Thank you very much!!
Thanks to Stijloor for your informative Post and/or Attachment!
Sponsored Links

  Post Number #3  
Old 7th March 2013, 05:24 AM
Pads38

 
 
Total Posts: 700
Re: IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirement

My reading of that would suggest that the minimum requirement is IP21, but that may be provided by the enclosure and a carrying case in combination, but only when the device is intended to only be used with the case.

The rationale for 8.3.1 in Annex A is useful.

The requirement for IP21 is, I guess, driven by the higher likelihood of liquid spills and general misuse. There is additional requirements in clause 5 for Accessible Parts (smaller finger used to judge what an accessible part is).
Thanks to Pads38 for your informative Post and/or Attachment!
  Post Number #4  
Old 7th March 2013, 06:06 AM
petruchio

 
 
Total Posts: 26
Re: IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirement

Thanks for this. The standard is appallingly worded, the get-out statement in sub-clause 7.2 being an isolated paragraph which makes no reference - implicitly or explicitly - to anything other than the enclosure.

My company has been making light therapy equipment for many years and this standard renders the non-LED products obsolete at a stroke. However, the sub-syndromal variant - Winter Blues - can continue to use these products so long as they are not represented as medical devices. It's an absurd situation. For once I'm thankful of ISO14971 which can apply common sense here.

petruchio
  Post Number #5  
Old 8th March 2013, 09:08 AM
Peter Selvey

 
 
Total Posts: 860
Re: IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirement

Yes, the text of 7.2 conflicts with 8.3.1, it's a mistake in the standard that will need to be rectified. There are a number of other cases of outright and provable mistakes in IEC standards (for example, the alarm standard IEC 60601-1-8, and all of the ECG standards like 601-2-25) so it's not completely uncharted territory.

Further, there is a reasonable argument to say that having a minimum requirement of IPX1 is excessive as a general requirement, taking into account the number and nature of electrical devices typically used in the home. That is, if it's required for medical, it should be required for a lot of other electrical devices, and there really is not the market data to support this.

It's also worth to note that a "true" IPX1 rating is not just a matter of wiping the water off and doing a dielectric strength test. Rather, IEC 60529 says the water " ... shall not ... interfere with the correct operation of the equipment ... ".

There is expected to be a range of medical equipment where the water would not cause any problem for basic safety (shock/fire) but could easily interfere normal operation such as sensors or getting into battery compartment or secondary electronics. Even though basic safety is OK, it fails IPX1.

One manufacturer of home use blood pressure meters found that IEC 80601-2-30, which clearly refers to home use equipment, does not specify that IPX1 or better is required. They pointed out that a particular standard (-2-30) overrides the collateral (-1-11), and concluded that the IPX1 test was not applicable. It's a kind of fudge, but gets to a reasonable result.

As always, modern regulation does not require compliance with standards. However, you do need to carefully document and justify any cases where you do not comply with a standard, even if it appears, as in this case, that the standard in fault.
Thank You to Peter Selvey for your informative Post and/or Attachment!
  Post Number #6  
Old 6th May 2014, 12:18 AM
flyingfrisbee

 
 
Total Posts: 17
Re: IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirement

To piggyback on this thread,

I was wondering if testing is required to declare an IP rating. If so, what kind of in-house or external testing would be required for say IP53?

Take care,

Friz
  Post Number #7  
Old 6th May 2014, 02:57 AM
Marcelo Antunes's Avatar
Marcelo Antunes

 
 
Total Posts: 3,099
Re: IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirement

Quote:
I was wondering if testing is required to declare an IP rating. If so, what kind of in-house or external testing would be required for say IP53?
You can declare without testing, but if you need to prove it, you need to test it. Tests for IP ratings are defined in IEC 60529.

Some general info can be found for example here: http://www.specfast.com.au/media/dow...ystem_ip53.pdf and here http://www.google.com.br/url?sa=t&rc...,d.cWc&cad=rja
Thanks to Marcelo Antunes for your informative Post and/or Attachment!
  Post Number #8  
Old 7th May 2014, 12:40 AM
flyingfrisbee

 
 
Total Posts: 17
Re: IEC 60601-1-11 - Medical Electrical Equipment IP (Ingress Protection) Requirement

In what situation would you need to prove an IP rating?

Would CE marking a device be one of those cases?

How about for infant phototherapy equipment located under the patient?

IEC 60601-2-50
"If INFANT PHOTOTHERAPY EQUIPMENT is located under the PATIENT it shall at least comply with
IPX3 specified in IEC 60529."
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Amendment 1 to 60601-1, 3rd Edition Medical Electrical Equipment Standard Pads38 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4 27th August 2012 09:35 AM
Marking (Labeling) Requirements on Medical Equipment per IEC 60601-1 jayanthi IEC 60601 - Medical Electrical Equipment Safety Standards Series 5 6th March 2012 12:39 AM
IEC 60601 Medical electrical equipment - Send declaration and guidance to customers? mwa IEC 60601 - Medical Electrical Equipment Safety Standards Series 1 10th July 2009 12:09 PM
Medical Electrical Equipment / System Peripherals - IEC 60601-1-3 freewind IEC 60601 - Medical Electrical Equipment Safety Standards Series 16 29th September 2008 11:47 AM
IEC60601-1 3rd edition +ve vote - Medical Electrical Equipment Safety Requirements celeo IEC 60601 - Medical Electrical Equipment Safety Standards Series 6 6th October 2006 08:39 PM



The time now is 11:22 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


 
 
 


NOTE: This forum uses "Cookies"