Does IEC 61010-1 apply for Surgical Microscopes?

L

leroybrown

Hi there. My company is supplying teaching labs with surgical microscopes only intended for use in the teaching labs. These microscopes will only be used for surgical technique practice using prosthetics - no contact with real patients. Would 61010-1 apply here whereas 60601-1 normally applies to the microscopes that are used in the doctor's offices/exam rooms? I am also curious as to what other regulations might apply, i.e. LVD directive.
All help is greatly appreciated!
Thanks! :bigwave:
 

Peter Selvey

Leader
Super Moderator
For Europe it depends on the stated intended purpose, so if it is clearly stated, not intended for use on patients, then the MDD (and hence EN/IEC 60601-1) would not apply.

The LVD and EMC are expected to apply (assuming there is mains powered, normal electronics), and probably EN/IEC61010-1, etc you would need to check the scope statement in the standard to be sure.
 

Marc

Fully vaccinated are you?
Leader
Peter, We all appreciate your helping with this type of question.

Thank you,

Marc
 