The Cove Business Standards Discussion Forums
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Flammability Ratings for Internal Battery & Hand-Held ME Equipment

Monitor the Elsmar Forum
Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
battery (electric batteries), flammability, medical equipment (general)
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 29th November 2013, 06:21 AM
bennylee

 
 
Total Posts: 23
Please Help! Flammability Ratings for Internal Battery & Hand-Held ME Equipment

Hello, All,

Our company currently concern on the flammability ratings for EU & USA (including Canada)

From IEC 60601-1 section 11.3 Constructional requirements for fire ENCLOSURES, we found that
A) Enclosure ( for TRANSPORTABLE ME) need V-2 or above
B) PCB need V-2 or above
C) Internal wire need V-1 or above
As our device is small, internally battery & hand-held ME equipment, it will not have any risk to cause fire when in use.
1) may we know if the above requirements for fire ENCLOSURES must apply?
2) Any different requirement for EU & USA ?

Thanks in advance!

Sponsored Links
  Post Number #2  
Old 29th November 2013, 06:09 PM
J0anne

 
 
Total Posts: 439
Re: Flammability Ratings for Internal Battery & Hand-Held ME Equipment

1- Analyse this in your risk management file and either meet this requirement or justify why you are not meeting this requirement

2 - 60601 certification is acceptable for marketing medical devices in the US
Thanks to J0anne for your informative Post and/or Attachment!
  Post Number #3  
Old 29th November 2013, 11:36 PM
bennylee

 
 
Total Posts: 23
Re: Flammability Ratings for Internal Battery & Hand-Held ME Equipment

Hello Joanne, Thanks for your information !

Actually, flammability ratings for some of enclosure is 94HB, which is lower standard than V-2, hence, we will need to do some risk analysis for this point.

As our test is applied to en60601-1 3rd version, in the test report from test firm, already shown enclosure is 94HB, i think it will be fine for EN market (with related risk analysis)

For or USA, I search some information from test firm, it seems that flammability ratings must follow ul, it will be more tough than EN standard, I still not very clear about this, may I know if anyone can share more information about this?

Thanks in advance !
  Post Number #4  
Old 2nd December 2013, 11:46 PM
bennylee

 
 
Total Posts: 23
Re: Flammability Ratings for Internal Battery & Hand-Held ME Equipment

I have checked information (it posted on February 1, 2004) from mddionline.com

It stated that "If the fire enclosure is sourced by circuits limited to less than 15 W, flammability requirements are not required." , may I know if this sentence is still effective for medical device.
  Post Number #5  
Old 3rd December 2013, 05:48 AM
Pads38

 
 
Total Posts: 771
Re: Flammability Ratings for Internal Battery & Hand-Held ME Equipment

Fire enclosures are, generally, only required where there is a significant likelihood of fire due to the presence of a source of ignition and a significant source of fuel.

Most materials used in devices do not represent a significant source of fuel unless they are used in an oxygen rich atmosphere.

Ignition sources are those that are capable of releasing more than 900 Joules.

The 15W exemption applies to the single fault conditions of clause 13.1.2. So, if you can show that, during a fault condition, your circuit cannot provide more than 15w of power then you do not need to consider clause 13.1.2.

References:
Clause 11.3 (and rationale)
Clause 13.1.2 (and rationale)
Interpretation Sheet 2, I-SH 02, IEC January 2009.
(I would have posted a link to that but the IEC web site is not working properly at the moment).

Edit:
Link to the Interpretation Sheet:
http://webstore.iec.ch/webstore/webs...6!opendocument

Last edited by Pads38; 3rd December 2013 at 09:13 AM. Reason: Webstore working again
Thanks to Pads38 for your informative Post and/or Attachment!
  Post Number #6  
Old 8th December 2013, 08:01 PM
Peter Selvey

 
 
Total Posts: 907
Re: Flammability Ratings for Internal Battery & Hand-Held ME Equipment

Just a note that the 1st amendment to IEC 60601-1:2005 has lifted the power limit to 100VA (in normal and fault condition) for parts under 60Vdc, provided that the PCB is FV-1 and the "wire insulation of types PVC, TFE, PTFE, FEP, polychloroprene or polybromide".

That will be a big help as there are a lot of secondary circuits and battery operated devices which are >15W but less than 100VA. PCBs with FV-1 rating and allowable wiring types are readily available.
Thank You to Peter Selvey for your informative Post and/or Attachment!
  Post Number #7  
Old 1st June 2018, 04:36 PM
John Alleman

 
 
Total Posts: 2
Re: Flammability Ratings for Internal Battery & Hand-Held ME Equipment

Any idea why wire insulation of types PVC, TFE, PTFE, FEP, polychloroprene or polybromide were specified?
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Requirements for hand-held laser welding equipment EMengineer Other ISO and International Standards and European Regulations 0 11th July 2018 08:27 AM
Kanomax hand held Laser Particle Counter Calibration joemar Calibration Frequency (Interval) 6 5th September 2017 10:44 AM
Hand-held Digital Tension Meter Calibration Dongzkie General Measurement Device and Calibration Topics 1 14th November 2014 06:41 AM
Should a low cost hand held refractometer be in a calibration program? mecramur Calibration Frequency (Interval) 8 30th March 2011 10:36 AM
Testing Agencies - Gas Meters and Hand Held Analyzers Copperfin General Measurement Device and Calibration Topics 2 14th September 2009 01:32 PM



The time now is 11:07 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.



Misc. Internal Links


NOTE: This forum uses "Cookies"