Hi All,
We are surgical instrument manufacturer and would like to ensure our instrument is meeting EMC Directive. Our instrument is work together with electrosurgical unit(ESU) which is not our product. In order to compliance to The EMC Requirement , do we need to perform IEC 60601-1-2 EMC Test sufficient or we need to perform both IEC 60601-1-2 EMC Test and IEC 60601-2-2 Annex BB ? On top of that, I was adviced that when performing IEC 60601-1-2 EMC Test , the sytem should be in idle mode and the ESU unit should not generate energy, is it Correct ? Thanks for your value input.
James
We are surgical instrument manufacturer and would like to ensure our instrument is meeting EMC Directive. Our instrument is work together with electrosurgical unit(ESU) which is not our product. In order to compliance to The EMC Requirement , do we need to perform IEC 60601-1-2 EMC Test sufficient or we need to perform both IEC 60601-1-2 EMC Test and IEC 60601-2-2 Annex BB ? On top of that, I was adviced that when performing IEC 60601-1-2 EMC Test , the sytem should be in idle mode and the ESU unit should not generate energy, is it Correct ? Thanks for your value input.
James