Medical Device Embedded Software Classification

Monitor the Elsmar Forum
Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
class iii medical device, embedded software, iec 62304 - medical device software life cycle processes, medical devices
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 10th March 2011, 06:38 PM
sl344

 
 
Total Posts: 28
Please Help! Medical Device Embedded Software Classification

Hello,

We have this piece of firmware that is embedded inside the medical device. the device itself is classified as class III. Based on the software hazard analysis, this firmware could potentially cause minor pain (and/or bruising) on patient skin. Based on IEC 62304, class A software has to be 'no injury or damage to health is possible'. Given that, what class do you think this firmware should be? is minor pain(bruising) considered an injury??

thanks a lot!

Sponsored Links
  Post Number #2  
Old 10th March 2011, 06:52 PM
QA_RA_Lady

 
 
Total Posts: n/a
Re: Software Classification

Are you looking for the FDA classification?
  Post Number #3  
Old 10th March 2011, 07:44 PM
sl344

 
 
Total Posts: 28
Please Help! Re: Medical Device Embedded Software Classification

thanks. I'm looking for software classification per IEC 62304. It could be class A, B or C.

the bottom line question is: How to determine software classification?

1) based on severity level prior to mitigation?
2) Based on analysis of level of concern?

thanks

Last edited by Stijloor; 10th March 2011 at 07:51 PM. Reason: Combined two posts for clarity.
  Post Number #4  
Old 11th March 2011, 05:52 AM
sagai

 
 
Total Posts: 1,133
Re: Medical Device Embedded Software Classification

Quote:
The MANUFACTURER shall assign to each SOFTWARE SYSTEM a software safety class (A, B, or C) according to the possible effects on the patient, operator, or other people resulting from a HAZARD to which the SOFTWARE SYSTEM can contribute.

The software safety classes shall initially be assigned based on severity as follows:
Class A: No injury or damage to health is possible
Class B: Non-SERIOUS INJURY is possible
Class C: Death or SERIOUS INJURY is possible

If the HAZARD could arise from a failure of the SOFTWARE SYSTEM to behave as specified, the probability of such failure shall be assumed to be 100 percent.

If the RISK of death or SERIOUS INJURY arising from a software failure is subsequently reduced to an acceptable level (as defined by ISO 14971) by a hardware RISK CONTROL measure, either by reducing the consequences of the failure or by reducing the probability of death or SERIOUS INJURY arising from that failure, the software safety classification may be reduced from C to B; and if the RISK of non-SERIOUS INJURY arising from a software failure is similarly reduced to an acceptable level by a hardware RISK CONTROL measure, the software safety classification may be reduced from B to A.
So,
1) based on severity level prior to mitigation, BUT
you can reduce in case it was mitigated by reducing to the acceptable level of the Risk by severity reduction.

regards
Sz.

Last edited by sagai; 11th March 2011 at 05:54 AM. Reason: clarification added
  Post Number #5  
Old 11th March 2011, 11:40 AM
glork98's Avatar
glork98

 
 
Total Posts: 84
Re: Medical Device Embedded Software Classification

Sagai has the pertinent section with highlights.

Unless you've got total mitigation outside of the software, you won't get away from the B classification.
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Software Medical Device Classification nellooo Medical Information Technology, Medical Software and Health Informatics 5 4th October 2017 08:33 AM
Classification of the Reseller of Stand Alone Medical Device Software Kuhalit ISO 13485:2016 - Medical Device Quality Management Systems 3 20th October 2014 04:25 PM
Device History Record - Post Shipment - Class II Medical Device w/ Embedded Software Watchwait 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6 10th July 2011 10:52 AM
Safety Classification in Medical Device Standalone Software theinonen IEC 62304 - Medical Device Software Life Cycle Processes 26 10th May 2011 09:29 AM
Software embedded in a Medical Device DuttonT IEC 62304 - Medical Device Software Life Cycle Processes 2 22nd April 2010 01:02 PM