How "clean" should the implant be before gamma?
Focusing only on sterilization and bioburden, and not associated issues such as pyrogenicity or gamma dosage materials effects:
Conceptually, there is no requirement for a particular level of pre-sterilization cleanliness. The requirement instead is that you validate your sterilization process for a particular "normal worst case" amount and type of contamination, then never exceed that amount and type of contamination during production operations.
Thus, the point for you is to figure out what range of cleanliness-level you can achieve
on an absolutely consistent basis, then conduct your sterilization validation with samples that are at the worst case edge of that range of contamination levels.