Can we do a Partial Gamma Validation on a Similar Product (Syringes)
UL - Underwriters Laboratories - Health Sciences
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Can we do a Partial Gamma Validation on a Similar Product (Syringes)
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Can we do a Partial Gamma Validation on a Similar Product (Syringes)


Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
bioburden, dose mapping, dose verification, gamma irradiation sterilization, iso 11137 - medical device radiation sterilization, medical devices (disposable), sterility and sterilization techniques, sterilization validation, syringes
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 31st March 2014, 11:22 PM
Nash27

 
 
Total Posts: 63
Please Help! Can we do a Partial Gamma Validation on a Similar Product (Syringes)

Dear All,

Due to the delays in the regulatory approval, our medical device (syringe) now have a manufacturing gap of 1 year. Last year we performed a full gamma validation (VDmax25) that included the dose mapping.

We manufacture other products at our clean room and regularly monitor environmental monitoring data and bio-burden determination of few products (they are not kind of products though).

Considering there are no major changes in the design and packaging can we conduct partial gamma validation? (as below)

? Bioburden determination on 10 samples
? Dose verification (10 sample with ref to table 9 ISO 11137-2)
? Sterility test (accept if no more than one positive test obtained)
? Prepare partial validation report (summarizing details)

Just want to know am I on right track or we need to perform a full validation along with dose mapping?


Thanks,

Nash

Sponsored Links
  Post Number #2  
Old 1st April 2014, 03:19 AM
pkost

 
 
Total Posts: 824
Re: Can we do a Partial Gamma Validation on a Similar Product (Syringes)

I don't believe this is a valid approach, you should validate the new products from scratch

Its not clear to me what the existing valdiation is for, you suggest that you manufacture other sterile products but say they are not the same. You then go on to say that there are no major changes to design or packaging

The standard gives a whole list of things that would require revalidation, I can't remember them off the top of my head but you should review that list.

It is possible to validate a worst case scenario but you have to be able to be able to state your assumptions and justify your decision making.
Thanks to pkost for your informative Post and/or Attachment!
  Post Number #3  
Old 1st April 2014, 04:28 AM
Nash27

 
 
Total Posts: 63
Re: Can we do a Partial Gamma Validation on a Similar Product (Syringes)

My apologies, i wasnt clear with my question. The device is in question is the same, I mean we validated our Syringe last year and because we havent received approval for the same, there was delay in manufacturing of one year. Now we have device approved and ready to go to the market. Although, there is gap of one year, we would like to ensure that the Syringe achieves SAL of 10-6

As we have already validated Bioburden method and sterility we only required to perform bio-burden determination and sterility testing (no need to validate product for bio-burden and sterility).

As per my information if there is no any change related to the device including packaging we can prove partial validation. We will opt this approach as product will be intermittent (that is there will be gap of more than 3 months). So, we have to do partial validation for every production run untill sufficient history established. Dose mapping we will need not do as the sterilizer has not changed their gamma source. When the sterilser provider will change the gamma source we will perform dose mapping again.

By doing the partial validation we might able to save some money.
  Post Number #4  
Old 1st April 2014, 06:48 AM
pkost

 
 
Total Posts: 824
Re: Can we do a Partial Gamma Validation on a Similar Product (Syringes)

I hope someone else with more experience in sterilisation can offer there advise here -

I have a vague recollection of the standard and what you suggest in this case seems to be valid, but there has been a significant change that you have to consider....time!

sorry I can't be any more helpful
Thanks to pkost for your informative Post and/or Attachment!
  Post Number #5  
Old 2nd April 2014, 05:47 PM
Nash27

 
 
Total Posts: 63
Please Help! Re: Can we do a Partial Gamma Validation on a Similar Product (Syringes)

I have reviewed ISO 11137 and confirmed with our sterilisation facilitator, I guess we are on right track.

If anyone are having some other ideas please share
  Post Number #6  
Old 4th April 2014, 11:11 AM
iamtroll's Avatar
iamtroll

 
 
Total Posts: 81
Re: Can we do a Partial Gamma Validation on a Similar Product (Syringes)

I agree that you are on the right track. Although my experience has been in ETO sterilization, my advise would be to consult with your sterilization provider since they see all the ins and outs of what is allowed and what is prohibited.
Thanks to iamtroll for your informative Post and/or Attachment!
  Post Number #7  
Old 7th April 2014, 08:31 PM
Nash27

 
 
Total Posts: 63
Re: Can we do a Partial Gamma Validation on a Similar Product (Syringes)

Yes, I confirmed with the sterilisation provider and yes the path we have selected is acceptable as long as I document all rationale. Thanks for your help.
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Gamma Irradiation Process Validation Help alonFAI Other Medical Device Related Standards 2 18th January 2017 05:59 PM
Gamma Validation vs. Quarterly Dose Audit Differences Nash27 ISO 13485:2016 - Medical Device Quality Management Systems 3 29th January 2013 06:20 PM
Gamma Validation - What should a Third Party Irradiation/Gamma Company Supply Andre EU Medical Device Regulations 2 16th November 2010 07:15 AM
Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements mknight ISO 13485:2016 - Medical Device Quality Management Systems 6 23rd July 2008 02:13 AM
Gamma Sterilization Validation Documents - 13485:2003 section 7.5.1.2.1 scwik2006 ISO 13485:2016 - Medical Device Quality Management Systems 5 3rd May 2006 03:34 PM



The time now is 08:45 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.



Misc. Internal Links


NOTE: This forum uses "Cookies"
The Elsmar Cove web site and all content is owned by Rabbithash Corp. as of 1 MAY 2018.
Unlike Facebook and other properties such as Linkedin, We don't buy, sell, or "share" visitor information. Never have, Never will...
GDPR Compliant since 1996, keeping Marc's vision alive!