M
Medevice
There are medical devices on the market that require the use of a connected laptop that the operator touches in order to perform their function, but they do not have a laptop supplied with them, only the software to run on the laptop.
My reading of IEC 60601-1 regarding ME SYSTEMS and PATIENT ENVIRONMENT indicates to me that any laptop used would most likely be in the PATIENT ENVIRONMENT and therefore subject to the testing required for the ME EQUIPMENT.
If I am correct, how is it that a device manufacturer can claim that any laptop can be used with the medical device when it is being used on a patient?
What have I missed?
My reading of IEC 60601-1 regarding ME SYSTEMS and PATIENT ENVIRONMENT indicates to me that any laptop used would most likely be in the PATIENT ENVIRONMENT and therefore subject to the testing required for the ME EQUIPMENT.
If I am correct, how is it that a device manufacturer can claim that any laptop can be used with the medical device when it is being used on a patient?
What have I missed?