The Cove Business Standards Discussion Forums More Free Files Forum Discussion Thread Post Attachments Listing Cove Discussion Forums Main Page
Saudi Arabia and Dubai - Class III Medical Device Regulations?
UL - Underwriters Laboratories - Health Sciences
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Saudi Arabia and Dubai - Class III Medical Device Regulations?
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Saudi Arabia and Dubai - Class III Medical Device Regulations?


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
class iii medical device, dubai, regulations and regulatory requirements, saudi arabia, ce mark and ce marking, medical devices
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 5.00 average. Display Modes
  Post Number #1  
Old 26th July 2010, 11:45 AM
tehuff

 
 
Total Posts: 56
Please Help! Saudi Arabia and Dubai - Class III Medical Device Regulations?

Hi all,
I have a request to ship our product to Saudi Arabia and Dubai. We currently have CE mark (class III product) but not FDA approval. Does anyone know the medical device regulations in either of these geographies? Are there import requirements that may also apply?

I found this info on the Saudi Network website under the Importation of Medicines and Pharmaceuticals section
"...imports of pharmaceuticals and medicinal products require a certificate issued by the Exporting country Department of Health and Human Services stating that the medicines are actually used by the public in the Exporting country under the same trade name and formula and must include the name of each product, the formula, and the date and number of the permit to manufacture them if one is required. "

Does anyone know if this also applies to Medical Devices?

Thanks!

Last edited by tehuff; 26th July 2010 at 11:51 AM.

Sponsored Links
  Post Number #2  
Old 26th July 2010, 01:32 PM
MIREGMGR

 
 
Total Posts: 3,685
Re: Saudi Arabia and Dubai - Class III Medical Device Regulations?

Quote:
In Reply to Parent Post by tehuff View Post

I have a request to ship our product to Saudi Arabia and Dubai. We currently have CE mark (class III product) but not FDA approval.

I found this info on the Saudi Network website under the Importation of Medicines and Pharmaceuticals section
"...imports of pharmaceuticals and medicinal products require a certificate issued by the Exporting country Department of Health and Human Services stating that the medicines are actually used by the public in the Exporting country under the same trade name and formula and must include the name of each product, the formula, and the date and number of the permit to manufacture them if one is required. "

Does anyone know if this also applies to Medical Devices?
My experience to date has been that the info at www.the-saudi.net is reliable, even though unofficial, in regard to Saudi Arabia.

Yes, the requirement for a COFG/CFS does apply to medical devices. That is in fact the only requirement for qualification for import: that the medical device must be legal for sale and use in its country of origin, as certified by that nation's (not the EC's) medical device regulatory authority.

Dubai is part of UAE. My understanding is that UAE judges import applications on a case by case basis, but generally following the same approach as the Saudis, i.e. if it's legal to market and use in its country of origin, it's accepted.

I didn't do a complete search here, but one past thread you could reference is http://elsmar.com/Forums/showthread.php?t=25421.
Sponsored Links

  Post Number #3  
Old 26th July 2010, 02:19 PM
bio_subbu's Avatar
bio_subbu

 
 
Total Posts: 617
Re: Saudi Arabia and Dubai - Class III Medical Device Regulations?

Quote:
In Reply to Parent Post by tehuff View Post

Hi all,
I have a request to ship our product to Saudi Arabia and Dubai. We currently have CE mark (class III product) but not FDA approval. Does anyone know the medical device regulations in either of these geographies? Are there import requirements that may also apply?

I found this info on the Saudi Network website under the Importation of Medicines and Pharmaceuticals section
"...imports of pharmaceuticals and medicinal products require a certificate issued by the Exporting country Department of Health and Human Services stating that the medicines are actually used by the public in the Exporting country under the same trade name and formula and must include the name of each product, the formula, and the date and number of the permit to manufacture them if one is required. "

Does anyone know if this also applies to Medical Devices?

Thanks!
As regards to KSA Import requirement, the Saudi Food & Drug Authority (SFDA) – Medical Devices Sector (MDS) has adopted Medical Devices Interim Regulation (MDIR), issued by Saudi Food and Drug Authority Board of Directors Decree number 1-8-1429 and Dated 27 December 2008.

The Interim applies to medical devices that fall under the definition of medical device as per Chapter One (Refer Article One of MDIR).

The manufactures, authorised representatives, importers and distributors intend to place a medical device on Saudi market, shall satisfy the following:

1. When the manufacturer is located outside the KSA, he shall appoint an authorised representative to act on his behalf (Refer chapter Four - Article Eleven, of MDIR).

2. Manufacturers established within the KSA, authorised representatives, importers and distributors of medical devices shall:

a) Register the establishments with the SFDA – Medical Devices National Registry (MDNR) by using this link http://Mdnr.sfda.gov.sa

b) List the medical devices with the SFDA – MDNR (Refer chapter Four - Article Ten, of MDIR).

3. Local manufacturers involved in distribution activities, as well as importers, distributors and authorised representatives involved in importation or distribution activities, shall apply for an establishment license. (Refer chapter Five - Article Fifteen, of MDIR) .

4. Medical devices may be placed on the market and/or put into service only if they comply with the applicable provisions of this Interim Regulation, as signified by the SFDA issuing the manufacturer with a written marketing authorisation (Refer Chapter Two – Article Four, of MDIR).

Regards
S. Subramaniam
Thank You to bio_subbu for your informative Post and/or Attachment!
  Post Number #4  
Old 26th July 2010, 03:23 PM
tehuff

 
 
Total Posts: 56
Re: Saudi Arabia and Dubai - Class III Medical Device Regulations?

Thanks for the excellent info! As I was reviewing it, I noticed in the MDIR that approval in EU, US, Canada, Japan, or Australia is also required.
  Post Number #5  
Old 27th July 2010, 03:39 AM
bio_subbu's Avatar
bio_subbu

 
 
Total Posts: 617
Re: Saudi Arabia and Dubai - Class III Medical Device Regulations?

In addition to my above post regarding KSA product import requirement, and I would like add some more comments to your question toward Dubai product registration, Dubai is part of UAE and medical Devices manufacturers must be registered with UAE Ministry of Health before you market your devices.

Companies wishing to export medical devices to UAE must have a local UAE REPRESENTATIVE or DISTRIBUTOR who holds Medical Store licensed from Ministry of Health. Imported medical Devices are subject to local registration requirements.

All the manufacturers of Class I, Class II a, Class II b, Class III and Active Implantable Devices has to be registered with UAE Ministry of Health.

The UAE local distributor must submit a registration application to MOH Drug Control Department.

A registration number will be given when an application is approved. This number will be valid for a 5 years period unless significant changes are made to the approved application data. The Director General can cancel the registration number, if any of the following takes place:

a. Based on the request of the applicants;

b. Based on evaluation or monitoring, following products may fail to meet the criteria because:

(1) they are not safe or are harmful to health,

(2) the quality is substandard,

(3) they differ from the approved label; this includes using a brand name which is originally owned by another rightful party.

(4) If submitted documents prove to be untrue

The Director will notify the registration holder in writing of the cancellation. Applications are considered complete if application forms and enclosures are filled up correctly according to the instructions. Application forms must be signed by the head of the company, the director or technical representative as indicated in the agency agreement. Indicate name of place, date, complete name of responsible person and company's seal. The exoneration process: 8-12 weeks after submission of complete documents.

For more details, please refer following guidance documents in existing thread “UAE IVD Registration (Professional Use Only, Non-A, Non-B of EU Regulation)” post # 5.

Registration of Manufacturer of Medical Device
Medical device manufacturers and products Registration flow chart
UAE Medical Devices Registration Guideline and receipt for submission of registration dossier Medical Device Product.

Regards
S. Subramaniam
Thanks to bio_subbu for your informative Post and/or Attachment!
  Post Number #6  
Old 28th July 2010, 10:19 AM
tehuff

 
 
Total Posts: 56
Re: Saudi Arabia and Dubai - Class III Medical Device Regulations?

Quote:
In Reply to Parent Post by bio_subbu View Post

Companies wishing to export medical devices to UAE must have a local UAE REPRESENTATIVE or DISTRIBUTOR who holds Medical Store licensed from Ministry of Health. Imported medical Devices are subject to local registration requirements.

The UAE local distributor must submit a registration application to MOH Drug Control Department.
Can a US manufacturer submit the registration application to MOH? I didn't see in the guidance documents that a local distributor is required (or is it a situation where it is "highly recommended" in order to successfully register)?
  Post Number #7  
Old 29th November 2012, 10:56 AM
drewsky1

 
 
Total Posts: 122
Re: Saudi Arabia and Dubai - Class III Medical Device Regulations?

Does anyone know if Dubai requires the CE mark for registration?

Thank you
  Post Number #8  
Old 29th November 2012, 12:10 PM
tehuff

 
 
Total Posts: 56
Re: Saudi Arabia and Dubai - Class III Medical Device Regulations?

Dubai is in the United Arab Emirates (UAE). CE mark is not required for UAE, but a simplified registration process is available for devices that have received approval from a recognized regulatory agency such as those in Europe, Canada, US, Australia or Japan.
Thanks to tehuff for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Is it possible to have more than one authorized representative in Saudi Arabia? ManTheDan88 Other Medical Device Regulations World-Wide 6 4th October 2017 03:56 PM
Saudi Arabia - Temperature Storage Requirements RA Princess Other Medical Device Regulations World-Wide 3 13th February 2017 04:11 AM
Saudi Arabia - Renewal of Registration kbryde Other Medical Device Regulations World-Wide 1 24th June 2014 05:48 AM
Does Saudi Arabia mandate that a CE Mark for a Medical Device is Required? Bunny Other Medical Device Regulations World-Wide 10 16th August 2012 02:21 PM
FTC "Made In U.S.A" Labeling Rules vs. Saudi Arabia's "Country of Origin" Regulations tosborn Other Medical Device Regulations World-Wide 6 19th November 2010 08:38 PM



The time now is 03:40 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


 
 
 


NOTE: This forum uses "Cookies"