The Cove Business Standards Discussion Forums
UDI Registration - Class II Medical Device Software
UL - Underwriters Laboratories - Health Sciences
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
UDI Registration - Class II Medical Device Software
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

UDI Registration - Class II Medical Device Software


Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
class ii medical device, fda (food and drug administration), gudid (fda global udi database), medical device software, software medical devices
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 21st November 2016, 08:24 PM
Peter So

 
 
Total Posts: 5
Please Help! UDI Registration - Class II Medical Device Software

Hi everyone,

My company makes Class II medical device softwares. I was wondering if we have to register our pre-existing Class II softwares in GUDID? We currently do not have our class II medical device softwares listed in the GUDID account. It is my understanding that devices manufactured and labeled prior to compliance date is exempt from the UDI requirement up to 3 years.


Following excerpt is taken from Unique Device Identifier System: Frequently Asked Questions, Vol. 1 issued on August 20,2014.

A.11. Question: Are devices manufactured and labeled prior to the compliance date subject to UDI requirements?

UDI requirements apply to devices placed into commercial distribution after the compliance date that applies to the device. A finished device manufactured and labeled prior to its compliance date and held in inventory is excepted from UDI requirements for three years after the compliance date. This exception applies to both products held in inventory by the labeler and those consigned to a hospital or other potential purchaser and held in inventory, but not yet purchased, by the potential purchaser. This exception applies to all UDI requirements.


Our plan is to register our softwares for upcoming software update that will include UDI in user manual, "about" dialog in the software, etc. But since our GUDID account does not currently list our class II softwares, are we being non-compliant to 21 CFR 830 UDI?

Any input are greatly appreciated. Thanks!

Sponsored Links
  Post Number #2  
Old 22nd November 2016, 04:38 PM
QA-Man

 
 
Total Posts: 135
Re: UDI Registration - Class II Medical Device Software

Very interesting question! Assuming the software is stand-alone, the question you're really with is "what is the manufacturing date"?

If the software is distributed in a packaged form, I would assume the manufacturing date is the date an individual unit is released for sale.

If the software is not packaged (downloaded from Internet), I would assume the manufacturing date was the last time a change was approved and the revised product was available.

Regardless, I would base my decision on the manufacturing date in the DHR; it is a record you must have and it must include a manufacturing date.
Thank You to QA-Man for your informative Post and/or Attachment!
  Post Number #3  
Old 22nd November 2016, 05:21 PM
Deepthi36

 
 
Total Posts: 5
Re: UDI Registration - Class II Medical Device Software

Hi there,

Existing Class II devices are NOT EXEMPT from the UDI requirement.

The exception only applies to LABELING of Class II devices that were labeled before the compliance date. In such case, i.e. inventory that was not labeled prior to the compliance date, you don't have to go and rework all the parts to have UDI-compliant labels.

Similarly any Class II product that is already in the field does not need to be retrofitted with UDI-compliant labels.

However you are expected to do everything else - i.e. create a UDI, register with the GUDID etc. for all Class II (and Class III devices).

Hope this helps. Let me know if you have any further questions.
Deepthi
  Post Number #4  
Old 22nd November 2016, 05:55 PM
QA-Man

 
 
Total Posts: 135
Re: UDI Registration - Class II Medical Device Software

Quote:
In Reply to Parent Post by Deepthi36 View Post

However you are expected to do everything else - i.e. create a UDI, register with the GUDID etc. for all Class II (and Class III devices).
That's not true.

If he doesn't have a product required to bear a UDI then he doesn't have to submit data to the GUDID database. That means if everything was manufactured before September 24th and then nothing since, no GUDID entry is required because no products are required to bear a UDI. See 21CFR830.300(a).
Thanks to QA-Man for your informative Post and/or Attachment!
  Post Number #5  
Old 22nd November 2016, 07:46 PM
Peter So

 
 
Total Posts: 5
Re: UDI Registration - Class II Medical Device Software

Quote:
In Reply to Parent Post by Deepthi36 View Post

Hi there,

Existing Class II devices are NOT EXEMPT from the UDI requirement.

The exception only applies to LABELING of Class II devices that were labeled before the compliance date. In such case, i.e. inventory that was not labeled prior to the compliance date, you don't have to go and rework all the parts to have UDI-compliant labels.

Similarly any Class II product that is already in the field does not need to be retrofitted with UDI-compliant labels.

However you are expected to do everything else - i.e. create a UDI, register with the GUDID etc. for all Class II (and Class III devices).

Hope this helps. Let me know if you have any further questions.
Deepthi
Perhaps you are thinking of this in terms of physical items. I feel that for software, the manufacture date is the date of the software release (except when you produce it in packaged form). When people download the software from the website, the software is not being manufactured and given to the customer.

Also, I do not see the point in registering the UDI for a device that does have UDI in the labeling. Let's say FDA wants to recall Product X. Announcement was made that Product X with UDI:xxxxxxxxxx should be returned to the manufacturer. UDI does not help identify the device at fault because device it self or labeling do not have UDI.
  Post Number #6  
Old 24th November 2016, 02:56 PM
Deepthi36

 
 
Total Posts: 5
Re: UDI Registration - Class II Medical Device Software

Quote:
In Reply to Parent Post by QA-Man View Post

That's not true.

If he doesn't have a product required to bear a UDI then he doesn't have to submit data to the GUDID database. That means if everything was manufactured before September 24th and then nothing since, no GUDID entry is required because no products are required to bear a UDI. See 21CFR830.300(a).
Hi there,

You seem to be confused between two different things here:
1.) UDI Compliance to the Regulation 21CFR Part 830
2.) UDI in Labeling, 21 CFR 801 Subpart B

The exception you are citing is to 21CFR801 Subpart B, i.e. "Do I need to put UDI barcodes on labels for all product that I have sitting in the field and in inventory that were labeled prior to the compliance date?"

So yes, you are excepted from putting the UDI in the Label. However you are absolutely required to apply UDI to all products that are currently in distribution. Therefore, you are required to assign a UDI, publishing the data in the GUDID etc. I have worked at 2 giant medical device companies since the UDI rule came into effect and at both places - all devices in current distribution were made to comply with the UDI rule - regardless of when the they first entered the market.

However we did use the inventory exception (in 21 CFR 801 Subpart B) to justify not reworking all our labels of products in inventory.

TLDR: All Class II and Class III device must at this point have assigned UDIs and comply with all the steps of the UDI regulation including registration in the GUDID. The UDI is meaningless without the GUDID registration piece. You may use the inventory exception in 801 Part B for parts labeled prior to the UDI compliance date.

Hope this helps.

Last edited by Deepthi36; 24th November 2016 at 03:34 PM. Reason: Added a summary at the bottom, added quote by QA Man to the top to clarify my response.
Thanks to Deepthi36 for your informative Post and/or Attachment!
  Post Number #7  
Old 24th November 2016, 03:16 PM
Deepthi36

 
 
Total Posts: 5
Re: UDI Registration - Class II Medical Device Software

Quote:
In Reply to Parent Post by Peter So View Post

Perhaps you are thinking of this in terms of physical items. I feel that for software, the manufacture date is the date of the software release (except when you produce it in packaged form). When people download the software from the website, the software is not being manufactured and given to the customer.

Also, I do not see the point in registering the UDI for a device that does have UDI in the labeling. Let's say FDA wants to recall Product X. Announcement was made that Product X with UDI:xxxxxxxxxx should be returned to the manufacturer. UDI does not help identify the device at fault because device it self or labeling do not have UDI.
Peter,

Agreed with you on the manufacturing date of a particular software version. If you are selling software that is imaged at one point and every sale of that software is the exact same image - then the manufacturing date is the date of approval and release of that software image. Therefore, let's say you created an image of your Class II software before Sept 24, 2016, then you can use the inventory exception rule to not have the UDI in the about screen - until the next release that needs a UDI. Then, you have to put the UDI in the About Screen.

Your second point --> Just because the UDI is not in the label (for 3 years max by the way), doesn't mean that it cannot be traced. By entering your data into the GUDID, you are providing FDA with information such as Pro Code, GMDN, and tons of other information that is useful in case of a recall. So, let's say you don't have the UDI in your About Screen. Customer calls you and says "Hi, I have an adverse event related to your product... Software X, Version 1.1" --> you will enter this product name, version and UDI which your service team will have on file, in the Service Record, Complaint File and the MDR and you may initiate a recall. --> FDA can take your UDI and use the GUDID attributes to identify all similar devices that are affected (by using the GMDN Code, or Pro Code for example). This is one of the MANY ways in which UDI is useful even if it is not on the label, for now. It is meant to become a holistic, big-data approach to recalls and is the real push for the UDI regulation by the FDA.

Finally, reading the UDI FAQ response, it says "UDI requirements apply to devices placed into commercial distribution after the compliance
date that applies to the device. " The key phrase -- placed into commercial distribution. Are you planning to sell more of your software after Sep 24 2016? If yes, you are "placing" the device being sold "into commercial distribution" after the deadline and therefore this exception does not apply.

Not complying with the UDI requirement is a serious non-compliance and I suggest contacting the UDI Help Desk directly for advice.

All the best!

Last edited by Deepthi36; 24th November 2016 at 03:37 PM.
Thanks to Deepthi36 for your informative Post and/or Attachment!
  Post Number #8  
Old 29th November 2016, 07:59 PM
yodon

 
 
Total Posts: 1,092
Re: UDI Registration - Class II Medical Device Software

Quote:
In Reply to Parent Post by Deepthi36 View Post

Then, you have to put the UDI in the About Screen.
Very interesting discussion, thanks to all who contributed. This may be a silly line of questions but would the UDI label in the about screen be the standard barcode plus other information? Would that even be readable (with a barcode reader - I know the scanners at, say, the airport can read barcodes off mobile devices but I'm not sure about all readers and monitors)? Further, if the software was intended to run (also) on a mobile device, I guess there would be an expectation that it renders properly on the device?
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Spain Class I Medical Device Registration MD_Simy86 EU Medical Device Regulations 14 21st September 2017 05:01 PM
Registration of Class I Medical Device in UK nickie EU Medical Device Regulations 2 7th August 2014 06:14 AM
Portugal Class I Medical Device Registration of device made in Italy MD_Simy86 EU Medical Device Regulations 1 16th June 2014 06:33 AM
Medical Device Registration Process in Italy - Class I and Class II Medical Devices ISO 13485 - Medical ISO 13485:2016 - Medical Device Quality Management Systems 2 21st May 2011 03:05 AM
Medical Device Class II Registration Meghna 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2 23rd June 2010 09:24 AM



The time now is 11:58 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.



Misc. Internal Links


NOTE: This forum uses "Cookies"
The Elsmar Cove web site and all content is owned by Rabbithash Corp. as of 1 MAY 2018.
Unlike Facebook and other properties such as Linkedin, We don't buy, sell, or "share" visitor information. Never have, Never will...
GDPR Compliant since 1996, keeping Marc's vision alive!