The Cove Business Standards Discussion Forums
Sample Size for Design Verification (Prototype Testing) for Medical Devices
UL - Underwriters Laboratories - Health Sciences
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Sample Size for Design Verification (Prototype Testing) for Medical Devices
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >
Forum Username

Elsmar Cove Forum Visitor Notice(s)

Wooden Line

Sample Size for Design Verification (Prototype Testing) for Medical Devices


Monitor the Elsmar Forum
Sponsor Links




Courtesy Quick Links


Links Elsmar Cove visitors will find useful in the quest for knowledge and support:

Jennifer Kirley's
Conway Business Services


Howard's
International Quality Services


Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum


Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining


Ajit Basrur
Claritas Consulting, LLC



International Standards Bodies - World Wide Standards Bodies

AIAG - Automotive Industry Action Group

ASQ - American Society for Quality

International Organization for Standardization - ISO Standards and Information

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Some Related Topic Tags
design verification, sample size, samples and sampling plans, prototypes and prototype testing, medical devices
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 4.00 average. Display Modes
  Post Number #1  
Old 17th November 2010, 02:59 AM
amadeuscurse

 
 
Total Posts: 5
Please Help! Sample Size for Design Verification (Prototype Testing) for Medical Devices

I have looked through various threads @ this forum and found some information as to what I need.
I need to figure out the number of samples during design verification (prototype testing) for medical devices.
We manufacture capital equipment (read pretty expensive prototypes) and I am trying to figure out justification for using 3, 5 or 7 samples (as I have read in other threads) as having sufficient confidence level and reliability to proceed w/ manufacturing.
I did see some posts by Bev D about the same. I could not find any statistically sound argument to justify the small sample size.
Any help would be appreciated.
Thank you in advance.
-Amadeus

Sponsored Links
  Post Number #2  
Old 17th November 2010, 04:36 AM
SteveK's Avatar
SteveK

 
 
Total Posts: 788
Lurker Re: Sample size for prototype testing

Hi Amadeus,

Welcome!

We had external testing for Class IIa portable electrical device based on one item. It was supposed to be the final design (rather than prototype) but needed modifications to meet EMC (EN 60601-1-2) requirements. Donít know if this helps. It has not been questioned by our NB.

Steve
  Post Number #3  
Old 17th November 2010, 12:17 PM
amadeuscurse

 
 
Total Posts: 5
Unhappy Re: Sample Size for Design Verification (Prototype Testing) for Medical Devices

Thank you Steve.
I have been telling the developers to come up with the sample size based on the confidence interval and repeatability of testing. I am comfortable with that approach. During a recent FDA audit, we were told that we need to have some sort of statistical rationale for the sample size.
I am struggling to come up with one since there does not seem to be a whole lot of sample size calculations for capital equipment (or maybe there is and I am not aware of it)
-Amadeus
  Post Number #4  
Old 17th November 2010, 01:02 PM
Statistical Steven's Avatar
Statistical Steven

 
 
Total Posts: 1,123
Re: Sample Size for Design Verification (Prototype Testing) for Medical Devices

This topic has come up over and over again throughout the medical device industry. A sampling unit does not have to be a unique prototype but might be a multiple tests on the same unit.
  Post Number #5  
Old 17th November 2010, 03:12 PM
Bev D's Avatar
Bev D

 
 
Total Posts: 3,600
Re: Sample Size for Design Verification (Prototype Testing) for Medical Devices

exactly so Steven. for medical devices or any instrument really that isn't single use, it's the number of runs or uses that constitute sample size.

this is particularly true in DESIGN verification as opposed to process or manufacturign validation, which woudl involve the manufcture of multiple devices to ensure that the manufacturign process is capable of meetign the requirements over the allowable range of inputs. (but even in these cases the number of actual devices need not be large)
  Post Number #6  
Old 17th November 2010, 04:52 PM
amadeuscurse

 
 
Total Posts: 5
Angry Re: Sample Size for Design Verification (Prototype Testing) for Medical Devices

Thank you Bev and Steven.
Thats excatly the rationale I provided to the FDA auditor, as long as the testing meets the confidence level and reliability criteria, we need to test 'x' samples based on the complexity of the instrument.
The FDA auditor did not like that and was looking for statistically 'sound rationale'. All my effort to explain the rationale fell to deaf ears. This is how we do it now
We determine the confidence level (say 95%) and reliability (say 90%) to verify design variability. We look at the mean, st dev of the test results to justify the number of samples we test.
I was asked to put in some statistically sound rationale around this.
I looked around and could not come up with one...any help or thoughts on how to proceed will be helpful.
-Amadeus
  Post Number #7  
Old 17th November 2010, 05:03 PM
Statistical Steven's Avatar
Statistical Steven

 
 
Total Posts: 1,123
Re: Sample Size for Design Verification (Prototype Testing) for Medical Devices

You can use statistical sample size rationale to justify your sample size. How you divide those samples in the design verification is your decision. Look at Dimitri Kececioglu, Reliability and Life Testing Handbook Page 47 for a sample size equation based on confidence and reliability.
  Post Number #8  
Old 17th November 2010, 09:24 PM
amadeuscurse

 
 
Total Posts: 5
Re: Sample Size for Design Verification (Prototype Testing) for Medical Devices

Thank you Steven.
I looked through page 47 of the book as you suggested. It talks about 'exact binomial confidence limits'.
Are you suggesting that I calculate the sample size based on the known values of confidence limits and reliability values?
Thanks in advance for your help.
-Amadeus
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > > >

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Emoticons are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Design Verification Sample Size - What philosophies are out there MNPIKE Design and Development of Products and Processes 5 16th February 2013 12:12 PM
Determining Sample Size for Design Verification and Design Validation edisoar Other Medical Device and Orthopedic Related Topics 20 3rd December 2012 07:32 PM
Sample Size for Design Verification and Validation gxzn008 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6 9th November 2011 09:17 AM
Sample Sizes for Design Verification of Medical Devices tomnmiami Inspection, Prints (Drawings), Testing, Sampling and Related Topics 16 27th July 2011 04:49 AM
Sample Size in Design Verification & Process Validation MasterBB Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2 16th November 2009 09:00 PM