Q
qq130230416
Dear all
i have a question, why RPN is S*O*D? what is meaning for this formula?
Dear all
A flaw with this rankin is that a failure mode with very low severity, high occurrence with low detection would score an RPN of 100, while a failure mode with high severity, low occurence and low detection would also score a 100. I favor a rating system that weights the severity more than detection.
In the new version of the FMEA manual (AIAG) they present SOD as an alternative way to the RPN ranking
Example :
S= 9 / O= 4 / D= 5 would mean RPN= 180 but SOD = 945
S= 4 / O= 9 / D= 5 would mean RPN= 180 but SOD = 495
S= 5 / O= 4 / d= 9 would mean RPN =180 but SOD = 549
In the past years SOD is applied in several companies I visit, just for the reason you stated.
best regards,
Antoine
thank you so much.S = Severity
O = Occurrence
D = Detection
Severity is how "bad" the injury would be if there was a failure (1=none, 10=death)
Occurrence is how likely is the defect to occur (1=unlikely, 10=often)
Detection is how likely are your controls to detecting the defect should it occur (1 = Always detected, 10 = never detected)
You rate each failure mode on a rating of 1 to 10 for each of the three. Multiply the ratings to get the RPN for the failure mode.
thank you so much.
i am sorry. my english is poor so i can not express clearly.
i mean why not RPN≠S*O or S*D? what does S*O*D mean in statistics?
Bev:Jim has it correct - these are really arbitrary rankings. They have no statistical validity or mathematical validity or science validity. There was a desire to try to quantify ratings ratehr than use a qualitative ranking.
Some people do use S*O
I haven't heard of anyone using S*D
A 'recent' development is to use a matrix of Severity by Occurrence that isn't mathematically derived.
Another development primarily from the medical device community is to use just the severity and allow teh testign to determine the occurence rate.