The Cove Business Standards Discussion Forums More Free Files Forum Discussion Thread Post Attachments Listing Cove Discussion Forums Main Page
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > Common Quality Assurance Processes and Tools > Quality Tools, Improvement and Analysis
Forum Username
Wooden Line
Automated and Manual Measurement System Comparison (Correlation)
Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed
Monitor the Elsmar Forum
Sponsor Links
Courtesy Quick Links
Links that Cove visitors will find useful in your quest for knowledge:
International Standards Bodies - World Wide Standards Bodies
International Standards Organization - ISO Standards and Information
Howard's
Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum
Bob Doering
Bob Doering's Blogs and,
Ajit Basrur
Claritas Consulting, LLC
Quality Digest
Some Related Topic Tags (Not all threads are Tagged)
automated measurement gages, correlation study, measurement systems
Reply
  Post Number #1
Old 31st October 2016, 11:10 AM
tahirawan11
Please Help! Automated and Manual Measurement System Comparison (Correlation)
I have a production process where 100% of machines are tested for ‘airflow’ using an automated measurement system. And a small number of machines 5% are also re-checked manually to see if the 'automatic testing' and 'manual testing' yield the same results. In order to check the consistency of both the results, I plan to do a ‘simple correlation’ analysis. And if the ‘Coefficient of correlation’ is higher than 0.5 then I would conclude that both the ‘automatic’ and the ‘manual’ measurement systems are consistent. I do not want to compare one machine against the other but just want to compare the results of the two measurements systems.
The raw data is attached below and the coefficient of correlation is – 0.241.
Can I conclude that there is no correlation between the results of the two measurement systems? Are there any suggestions to use other tools than simple correlation analysis or other ways to conduct the same study?
automated_testing manual_testing
18,4 18,9
18,1 19,4
18,3 17,8
18,1 18,7
18,2 17,2
18 17,1
18,2 16,3
18,1 17,8
18,1 18,1
18,2 18,2
18,2 17,6
18,2 18,5
18,4 16,1
18,3 18,4
18,2 14,8
18,2 18,8
18,1 17,6
18,1 18,4
18 18,8
18,4 18,5
18,2 19
18,2 18,8
18,3 19,3
18,3 19
18,1 18,2
18,2 17,2
18,1 18,9
18,1 19,8
18,3 20
thanks
Sponsored Links
  Post Number #2
Old 31st October 2016, 12:52 PM
Bev D's Avatar
Re: Automated and Manual Measurement System Comparison (Correlation)
I understand that you don't want to compare instruments to each just the measurement methods. This is a standard Measurement System Analysis.
I caution you that 'simple regression' and correlation coefficients are not appropriate for MSA's because this approach is used for two data series, one of which is dependent on the other. When comparing two measurements - or two measurement systems as in this case - the two measurements are independent of each other. Therefore the graphical display and the 'correlation coefficient are different. The graphical display is a chart with a 1:1 (45 degree) line intersecting the origin and not the best fit line. The coefficient of correlation is the "Intraclass correlation coefficient".
In your case, the agreement between the two methods is so bad that the coefficient is - regardless of the formula - is irrelevant. but the GRAPH is.
I have plotted the values you provided on a Youden plot. (see attached file) I'm not sure which method is represented in your two columns. A is the first column and B is the second column. The two methods are not equivalent.
Attached Files: 1. Scan for viruses before using, 2. Please report any 'bad' files by Reporting this post, 3. Use at your Own Risk.
Thanks to Bev D for your informative Post and/or Attachment!
Sponsored Links
  Post Number #3
Old 1st November 2016, 06:41 AM
tahirawan11
Thanks Bev D. I have one question how should i interpret the graphical display. How does the chart look like if there is a PERFECT match between the two measurement systems? Does it mean that all points will lie on the 45 degrees line and any deviation from this diagonal line indicates a mismatch between the two measurement system?
  Post Number #4
Old 1st November 2016, 08:42 AM
Bev D's Avatar
A perfect match would have all of the dots on the 1:1 line. variation centered around the 1:1 line is simple measurement error. if the cloud of values is not centered on the 1:1 line there is a bias between the two instruments or measurements.
I have two papers posted at www.qualityforumonline.com in the Resources tab under "Practical Quality Engineering Resources"
MSA Tools is a spreadsheet with various MSA graphs and calculations.
MSA Verification & Validation is a paper describing how the various tools work and when to use them
These two are at the bottom of the list...
Near the top of the list is "The Statistical Cracks in the Foundation of the Popular Gauge R&R Method".
Reply
Bookmarks
Search this Thread:
Advanced Forum Search
Rate Thread Content:
Forum Posting Settings
BB code is On
Emoticons are On
[IMG] code is On
Similar Discussion Threads
MSA (Measurement System Analysis) Study on an Automated Process Edward_Nell Quality Tools, Improvement and Analysis 4 19th August 2011 08:08 AM
MSA (Measurement System Analysis) on automated machine Caster Capability, Accuracy and Stability - Processes, Machines, etc. 1 12th December 2008 01:40 PM
Automated In-Line Inspection MSA (Measurement System Analysis) jorge2175 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5 2nd January 2008 05:04 PM