Why two fuses are required - IEC 60601-1 cl. 8.11.5
Some Related Topic Tags
class i medical device, iec 60601 - medical electrical equipment
  Post Number #1
Old 27th June 2018, 07:34 AM
Kunal Deshpande
Question Why two fuses are required - IEC 60601-1 cl. 8.11.5
Hi all,
Why two fuses are required for Class I equipment, why not Class II.
Kunal
  Post Number #2
Old 27th June 2018, 08:11 AM
Pads38
re: Why two fuses are required - IEC 60601-1 cl. 8.11.5
Hi Kunal,
In a Class I device a fault could be short circuits -
1) L-PE
2) N-PE
3) L-N
To protect against all three faults fuses are required in both mains supply lines - a fuse in the L line would not necessarily protect a N-PE fault.
For a Class II device only the L-N fault needs protection - so a single fuse may suffice.
Welcome to the forums.
Thanks to Pads38 for your informative Post and/or Attachment!
  Post Number #3
Old 11th July 2018, 09:14 PM
Peter Selvey
Re: Why two fuses are required - IEC 60601-1 cl. 8.11.5
The reason might be a little different:
Although effort is made to have the correct polarity (L&N) in the supply, it's not assumed.
So, if the L & N are swapped, a fuse in the N can still clear the fault to earth in the equipment. Obviously this does not apply for Class II equipment.
In normal appliances this is not required, as it is assumed the wall supply can still clear a fault (at least, when I was working with the old IEC 335 series 25 years ago ... it could have changed).
But for medical devices, it is considered important that the device can clear its own faults as far as possible, and not take out the supply of the theatre or ward etc.
Note that typically a N-PE fault should not cause the fuse to open. They are similar potential.
What edition of the IEC 60601-1 standard is required? ganglai IEC 60601 - Medical Electrical Equipment Safety Standards Series 6 26th October 2017 12:03 AM
IEC/EN 60601-1 Certificate required before CE Marking? LievenDW IEC 60601 - Medical Electrical Equipment Safety Standards Series 3 14th June 2017 09:10 PM
When is IEC 60601 required? KatEspinoza IEC 60601 - Medical Electrical Equipment Safety Standards Series 4 8th May 2015 10:11 AM
Is compliance to IEC 60601-1 Required for my Product? kmbird33 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6 4th February 2014 04:41 PM
When is Compliance required for 4th edition IEC 60601-1-2 bassa IEC 60601 - Medical Electrical Equipment Safety Standards Series 5 5th August 2013 04:58 AM
The time now is 07:22 AM. All times are GMT -4.
Misc. Internal Links
