Non-normal Distribution Selection where the system is constantly being corrected
Non-normal Distribution Selection where the system is constantly being corrected - Page 2
Some Related Topic Tags
non-normal data distribution
  Post Number #9
Old 13th March 2018, 01:08 PM
Miner's Avatar
Miner
Look! Re: Non-normal Distribution Selection where the system is constantly being corrected
I agree with Bob. That has been my experience also. The degree of sawtooth depends on a lot of variables (e.g., controller, inertia of the system being controlled, etc.).
Thanks to Miner for your informative Post and/or Attachment!
  Post Number #10
Old 13th March 2018, 01:19 PM
BradM's Avatar
BradM
It all depends....
I have seen perfectly tuned systems. If the rate is set properly, it should anticipate the rate of change and adjust accordingly. There are many applications ( aerospace, aluminum treating +/5 F) where there can be any overshoot.
Now... here's the thing... it also depends on the sensitivity of the probe and its location. They may have an MGO sheathed probe which is not as sensitive.
The recorder is connected to a wire thermocouple which is much more sensitive.
The controller can still be adjusted to remove a lot of the swing.
Also, i will get completely different results using 22awg versus 16awg thermocouples.
So the type of distraction can end up being a lot of things, depending on the type of control error.
  Post Number #11
Old 13th March 2018, 03:50 PM
bobdoering's Avatar
bobdoering
Exactly - the ability to see process change is based on the resolution of the measuring instrument (GAGE R%R to %PV). Enough resolution will definitely see sawtooth...its physics. You would have to have a perfectly synced load to energy application to have it on...constantly...and not have it rise in temperature. Pretty rare, I would say.
Also, remember, ramp up and ramp down are special causes, and do not relate to the capability, only the steady state.
  Post Number #12
Old 13th March 2018, 04:50 PM
BradM's Avatar
BradM
I have a temperature controller. If I have it tuned properly, it's not going to swing; or at least not to a significant level.
Ramp up is also affected by the PID settings. The point at which the controller starts throttling back (and at what percent) also depends on the PID.
Bottom line, if your process equipment is not a nice line, including a smooth ramp up with no overshoot, the controller needs adjustment.
Relationship between Normal Distribution and AQL siumks AQL - Acceptable Quality Level 13 1st March 2012 05:12 PM
Normal distribution - I have a series of data but I don?t know the distribution smedboy Statistical Analysis Tools, Techniques and SPC 24 23rd July 2009 09:28 AM
Calculating Cp and Cpk on a Non-Normal Distribution Andrews Capability, Accuracy and Stability - Processes, Machines, etc. 14 21st April 2009 07:50 PM
Non-Normal Distribution vs. Gamma Distribution fed-up Statistical Analysis Tools, Techniques and SPC 18 19th January 2009 02:41 PM
Determining whether a process yeild is normal or non-normal distribution? niotusen Six Sigma 10 4th November 2005 09:54 AM
The time now is 07:27 PM. All times are GMT -4.
Misc. Internal Links
