Does the cable assembly supporting a medical device gets classified accordingly?

SGquality

Quite Involved in Discussions
I need advise on medical device classification -

if there is a cable assembly that supports a Class 2 medical device, would the cable assembly be labeled as a Class 2 medical device too? This question is specifically from a device listing perspective for one of the cable supplier who just manufactures cables for our class 2 medical device equipment.

Thank you!
 

somashekar

Leader
Admin
I need advise on medical device classification -

if there is a cable assembly that supports a Class 2 medical device, would the cable assembly be labeled as a Class 2 medical device too? This question is specifically from a device listing perspective for one of the cable supplier who just manufactures cables for our class 2 medical device equipment.

Thank you!
Yes, we have done as you have described with our NB having given the green light for this.
 