Pulse oximeters ISO 80601-2-61, is 0% okay for probe fault?

thirdplanet

Registered
Per 80601-2-61 section 201.13.101 , a visual indication of pulse oximeter probe faults is required. The example given is a "blank" display.

Is it acceptable to use 0% instead?

Anyone commenting on their experience getting a pulse ox through 510 k) would be greatly appreciated!
 

yodon

Leader
Super Moderator

thirdplanet

Registered
Thats a good point, I think blank would probably be best.

For arguments sake: if we risk-analyzed displaying 0% and determined the risk of misinterpretation was low (feasible for our device class/risk profile), do you think that would fly during 510k?

We are already in works to change it from 0% to blank, so this is an academic question to gain some insight.
 

yodon

Leader
Super Moderator
Thats a good point, I think blank would probably be best.

This is the whole idea behind usability engineering. What Draeger thought was clearly not what the users thought. Without getting user feedback, you're just guessing.
 