We're in the process of reviewing our Company's quality manual and want to get thoughts on our Non-Applicability section:
The scope of our company's ISO 13485:2016 certification is basically Design and Development of IVDs, i.e., no product manufacturing.
I thought we could potentially exclude 7.5 altogether, but because the standard states " When the term “product” is used, it can also mean “service”. Product applies to output that is intended for, or required by, a customer, or any intended output resulting from a product realization process." we could not list it as non-applicable.
Would it make sense to then list the following clauses as non-applicable?
7.5.2 - Cleanliness of product
7.5.3 - Installation activities
7.5.4 - Servicing activities
7.5.5 - Particular requirements for sterile medical devices
7.5.7 - Particular requirements for validation of processes for sterilization and sterile barrier systems
7.5.9.2 - Particular requirements for implantable medical devices
7.5.11 - Preservation of product
Every time we undergo a notified body audit, they have differing opinions on which of our non-applicable clauses should be included, prompting constant revisions to our Quality Manual so I'd love to get thoughts on this and the justification.
If any clarifications are needed, please don't hesitate to ask.
The scope of our company's ISO 13485:2016 certification is basically Design and Development of IVDs, i.e., no product manufacturing.
I thought we could potentially exclude 7.5 altogether, but because the standard states " When the term “product” is used, it can also mean “service”. Product applies to output that is intended for, or required by, a customer, or any intended output resulting from a product realization process." we could not list it as non-applicable.
Would it make sense to then list the following clauses as non-applicable?
7.5.2 - Cleanliness of product
7.5.3 - Installation activities
7.5.4 - Servicing activities
7.5.5 - Particular requirements for sterile medical devices
7.5.7 - Particular requirements for validation of processes for sterilization and sterile barrier systems
7.5.9.2 - Particular requirements for implantable medical devices
7.5.11 - Preservation of product
Every time we undergo a notified body audit, they have differing opinions on which of our non-applicable clauses should be included, prompting constant revisions to our Quality Manual so I'd love to get thoughts on this and the justification.
If any clarifications are needed, please don't hesitate to ask.