13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please

#1
Hi There, I have a quick question regarding the validation of Software used under scope, in relation to
13485:2016, Particularly sections 4.1.6, 7.5.6 and 7.6


Does Software used to create the Software that is used under scope (e.g Visual Studio, used to create a Application used in Production) need to be validated?

Seem to be having a lot of confusion on what precisely needs to be validated, The Actual app used in production, or the software used to make the App

Any help here would be appreciated!
 
Elsmar Forum Sponsor

Tidge

Trusted Information Resource
#2
Does Software used to create the Software that is used under scope (e.g Visual Studio, used to create a Application used in Production) need to be validated?
I believe you are describing NPS - non-product software.

4.1.6 is about software used to support the Quality System. It is unlikely that NPS used on the production floor supports the quality system. It is typical that all software used in a medical device company is subjected to a formal assessment which can serve as the record for how the software is used. (see the note in 4.1.6 about records)

7.5.6 is the general "validate your production processes" section. If an app (let's say "compiler studio") is used to make an executable used in production (not medical device software!), ultimately you only have to validate that the executable meets its intended use. If the "compiler studio" is not being used in production, there is no need to subject it to NPS validation(1). If libraries from the studio end up as part of the production executable, they will end up being "validated by inclusion."

7.6 has specific details if the executable used in manufacturing is applying some 'test method'. It is unlikely that a compiler studio itself is subject to the requirements of this section.

(1) It is typical that the "compiler studio" will be documented at some level specifically so that a manufacturer can support an NPS executable "through the lifetime" of that executable. This can include storing the "studio" such that executable code can be touched up and/or recompiled. Detailing and archiving an off-the shelf studio is a very narrow form of (NPS) validation.

As long as you are subjecting the output of the studio package to 7.5.6, you are under no circumstances are you obligated to 'prove' that the visual studio works.
 
#4
Not sure if the policy is to start a new thread or tag on my questions here: I'll move this at your urging

First: For 4.1.6: Software used to support the QMS. What is the standard approach taken with widely used enterprise tools like Google Drive and it's relatives, Google Docs, Sheets, E-signature software etc... Google and Microsoft products will already have better IQ, OQ, PQ done than just about anyone can manage. (just talking about 13485 here, not 21CFR requirements)

Secondly: with reference to 4.1.6, 7.5.6 and 7.6, for a SaMD company/product what is the expectation and sensible approach for externally hosted and widely used production environments, services and microservices (thinking GoogleCloud, AWS etc..). What are the expected articles of evidence here?
 
Thread starter Similar threads Forum Replies Date
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 1
Louddogsbark When your 13485:2016 certificate has been pulled ISO 13485:2016 - Medical Device Quality Management Systems 2
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
P ISO 13485:2016 MDSAP Certification Fee Survey ISO 13485:2016 - Medical Device Quality Management Systems 6
M MDSAP and 13485:2016 gap analysis Quality Management System (QMS) Manuals 1
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
H ISO 13485:2016 Gap Analysis by NB ISO 13485:2016 - Medical Device Quality Management Systems 7
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9
JoCam Difference between Approval and Registration - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
T Document control ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
Sam Lazzara Medical Device File (MDF per 13485:2016 4.2.3) versus FDA Device Master Record (DMR) ISO 13485:2016 - Medical Device Quality Management Systems 3
L ISO 13485:2016 Clause 8.4 - Analysis of Audit Observations ISO 13485:2016 - Medical Device Quality Management Systems 8
S When is ISO 13485:2016 6.4.2 Contamination Control appropriate? ISO 13485:2016 - Medical Device Quality Management Systems 11
L Templates for three ISO 13485:2016 SOPs ISO 13485:2016 - Medical Device Quality Management Systems 8
C What falls under the 'Customer Property' according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 10
M Is it possible to get iso 13485:2016 certified as a one man band ISO 13485:2016 - Medical Device Quality Management Systems 1
F ISO 13485:2016 Quality Policy Requirements Other ISO and International Standards and European Regulations 18
M Contract Manufacturers and MDF Responsibilities, ISO 13485:2016, Clause 4.2.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
J ISO 13485:2016 sample exam/test ISO 13485:2016 - Medical Device Quality Management Systems 3
M Informational Questionário – Análise crítica sistemática – ISO 13485:2016 (Portuguese-only) Medical Device and FDA Regulations and Standards News 0
M Informational ISO 13485:2016 under systematic review Medical Device and FDA Regulations and Standards News 5
C Updates on Documentation for outsourced OEM from ISO 13485:2003 to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 4
GStough Informational SN EN ISO 9001:2015 and SN EN ISO 13485:2016 on Same Certificate? Registrars and Notified Bodies 7
E MDSAP Audit - Our QMS conforms to ISO 13485:2016 and FDA GMP Canada Medical Device Regulations 9
Ronen E Informational ISO 13485:2016 Transition Period End - 1 March 2019 ISO 13485:2016 - Medical Device Quality Management Systems 0
B ISO 9001:2015 vs ISO 13485:2016 for MDR Compliance EU Medical Device Regulations 4
T ISO 13485:2016 - Processes exempt from process validation ISO 13485:2016 - Medical Device Quality Management Systems 12
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
J EU ISO 13485:2016 Recertification Audit - Effect of 10 Minor Nonconformances EU Medical Device Regulations 2
E Template of a Management Review Agenda or Report in compliance with ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 6
J ISO 13485:2016 Section 6.2 - Documenting the process for establishing competence ISO 13485:2016 - Medical Device Quality Management Systems 6
Q Any good Checklists for ensuring SOPs cover ISO 13485:2016 and 21CFR 820? ISO 13485:2016 - Medical Device Quality Management Systems 3
H Transition to ISO 13485:2016 together with ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 12
B Classes/ Online Training on ISO 13485:2016 and FDA QSR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 5
T Software Validation Certificate (ISO 13485:2016) ISO 13485:2016 - Medical Device Quality Management Systems 19
D ISO: 13485:2016 Sec. 7.5.2 (C) - Requirements for cleanliness of product or contamination control ISO 13485:2016 - Medical Device Quality Management Systems 2

Similar threads

Top Bottom