17025 and PAR 4.2.2

Douglas E. Purdy

Quite Involved in Discussions
#1
I am in the process of re-familiarizing myself with 17025 and ran into a question regarding paragraph 4.2.2 under Quality System. The first sentence establishes the requirement for a Quality Manual, but the next two sentences center on a quality policy statement. The lead-in sentence to the a) thru e) sub-requirements states "It shall include at least the following:." My question is ... What is 'It' refering to, the Manual or the Policy Statement?

Per the note it would seem 'the Quality Policy Statement,' but how can you cover the a) thru e) in a statement and still be useable as a Quality Policy?

Thanks!
 
Last edited:
Elsmar Forum Sponsor
J

Jimmy Olson

#2
You are correct in that the 'IT' does apply to the policy. As far as covering everything, it doesn't have to be real specific. As long as everything is basically implied it should be fine.
 

howste

Thaumaturge
Trusted Information Resource
#3
The way I read it, it must all be in the policy statement. Then of course there's a note that it should be concise. :rolleyes:
 

Douglas E. Purdy

Quite Involved in Discussions
#4
Any Examples?

Would someone send me an example that would comply? In the case of the laboratory being a part of an organization that is compliant with ISO-9001:2000, does one have a Quality Policy that complies with 9001 and another for 17025? Any suggestions or recommendations to comply with both standards would be welcome.

Thanks,
Doug
 

howste

Thaumaturge
Trusted Information Resource
#5
Sorry, I don't have an example policy for 17025. I wouldn't recommend having a combined 9001 and laboratory policy though unless all of the people in your organization work in/for the lab. There's way too much required in there that wouldn't apply to most of the organization. I used to have a laboratory scope where I worked for calibrations we performed, and it was completely separate from the quality policy.
 
J

Jimmy Olson

#6
Unfortunately I don't have an example either.

I do agree with Howste about keeping the policies seperate. However, as the note in the standard indicates, if the lab is part of a larger organization some elements of the policy may be located elsewhere. You could review the organizations quality policy established for 9001 and then develope the policy for the lab to fill in the gaps. Or you could just do an entirely seperate one if you want.
 
P

pancho-2007

#7
Out Quality Policy is seperate from the manual. We made and distribute wallet cards for all employee's as well as have copies of the policy on every line on the floor of the plant. Our policy states as follows:

We the management of (Company Name) are commited to provide our customers with quality product, at a competitive price, and to achieve Total Customer Satisfaction.

Managements policy will be realized through participation of ( Company Name) human resources, and implimented by a quality oriented manufacturing system.

I do Hope this helps
 

howste

Thaumaturge
Trusted Information Resource
#9
I emailed someone I know that has a 17025 certified lab for a copy of their policy, and this is what they sent back (I suspect they pulled the whole section of the quality manual):
5.3 QUALITY POLICY

MISSION STATEMENT

Provide the highest standard of laboratory
testing and customer service.
And have fun in the process.

VALUES (Quality Objectives)

1. Integrity - We act with integrity in all that we do.

2. Quality - Each employee is responsible for quality, and we aggressively strive for perfection.

3. Service - We work with our customers internally and around the world to learn their needs and exceed their expectations. We recognize that great customer service and quality are our primary marketing tools.

4. Productivity - We deliver a superior, affordable product by controlling expenses and maximizing efficiency with state-of-the-art technology and facilities.

5. Training - We invest in training each employee to assure that our customers will communicate with industry experts.

6. Employees - We believe that our employees are the best in the industry and our most important asset. We empower them to grow professionally, innovate, and share in the success and growth of the company.

7. Communication - We emphasize rapid and frequent communication. We treat each other and our customers professionally and with respect to facilitate teamwork.

8. Fun - We have fun in the process.

? ? ? ? ? ? ? ? ? ?

XXXX Laboratories' goal is to provide the highest quality laboratory results at a fair price, and to make every effort to provide the results when the customer needs them. We will treat all of our customers as we would like to be treated. We will endeavor not to make unnecessary rules for our customers. We do whatever it takes, without compromising quality, to complete the job on time. We always strive to provide 'knock their socks off' service. Every XXXX employee is a customer service representative.

To accomplish turn-around-time goals and communicate with sponsors, XXXX management:

1. Monitors turn-around-time (TAT) performance

2. Sets turn-around-time goals with employees

3. Maintains lists of TATs on computer

4. Advises customers of estimated TAT upon receipt of samples

5. Flags samples which are needed by the sponsor sooner than normal processing times

To accomplish quality goals, XXXX management:

6. Has a high ratio of Quality Team staff to analysts

7. Requires the Quality Assurance Unit to normally review 100% of XXXX laboratory generated results. A reduced inspection level may be implemented for those tests which have a history of low defects.

8. Confirms the accuracy of each page at the time of final envelope sealing;

9. Performs a wide range of laboratory controls;

10. Maintains a Full-time Training Department to assure that new employees are trained by a master for that function in each area of expertise;

11. Maintains state-of-the-art laboratory facilities;

12. Maintains a sophisticated metrology laboratory with NIST-traceable standards;

13. Provides document accessibility via computer;

14. Analyzes trends of tests and critical systems to ensure continuous improvement

The Quality System is designed to provide a documented framework for providing world class quality service and comply with good laboratory practice (GLP), good manufacturing practice (GMP), good automated laboratory practices (GALP), ISO 9001:2000 quality standard requirements and requirements of ISO/IEC 17025 General Requirements for the Competence of Testing and Calibration Laboratories.

XXXX Laboratories is committed to good professional practice, providing the highest quality service, and striving for continuous improvement. The Quality Policy is reviewed at least annually; and quality goals are set in annual company-wide quality planning meetings, then reviewed approximately monthly in Quality Committee. The Quality Policy is given, in writing, to all new employees, is posted in the laboratory , and employees are trained regularly.
 
Thread starter Similar threads Forum Replies Date
V ISO 17025 Accreditation Consulting Needed ISO 17025 related Discussions 1
D Is Good Laboratory Practice (GLP) Required in ISO 17025? ISO 17025 related Discussions 4
W Looking for IATF 16949 (and ISO 17025) QMS software Suggestions Quality Tools, Improvement and Analysis 8
R Process Maps for ISO 17025 ISO 17025 related Discussions 2
K ISO 17025 Method Validation and Verification for Test Lab ISO 17025 related Discussions 4
R ISO 17025 Lead Assessor Training ISO 17025 related Discussions 0
J ISO 17025 Calibration Other Medical Device Related Standards 2
E ISO 17025 Key Performance Indicators (KPIs) ISO 17025 related Discussions 5
R Basic Lab Practices - related to ISO 17025 testing labs ISO 17025 related Discussions 1
D Training for ISO 17025 ISO 17025 related Discussions 2
D How to handle suppliers that are not ISO 17025 certified ISO 17025 related Discussions 5
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
Quality Specialist Quick question on the sequence of ISO 17025 external audits ISO 17025 related Discussions 14
E 17025 Accredited Microbiological Testing Laboratory ISO 17025 related Discussions 3
X Looking for 17025 auditor to perform internal audit on IT software testing laboratory ISO 17025 related Discussions 3
K ISO 17025 Facilities and environmental conditions SOP ISO 17025 related Discussions 6
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
G Why are 17025 scope angle measurements limited to 5 to 85 degrees? ISO 17025 related Discussions 2
B ISO 17025:2017 Clause 7.1.3 - Decision Rule ISO 17025 related Discussions 2
E Organisational Chart-ISO 17025 Laboratory ISO 17025 related Discussions 3
Casana ISO 17025 - Contractor certification? IATF 16949 - Automotive Quality Systems Standard 5
B Sanctioned interpretation #10 - ISO 17025 IATF 16949 - Automotive Quality Systems Standard 2
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 8
D Management reviews ISO 17025 ISO 17025 related Discussions 3
I ISO 17025:2017 / ANAB 3125 - Articulating / Communicating Risks vis-a-vis Audit Findings ISO 17025 related Discussions 2
W Do Reference materials for IATF need to be 17025? IATF 16949 - Automotive Quality Systems Standard 15
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 4
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 8
M Must a company be 17025 accredited to perform internal calibrations? ISO 17025 related Discussions 10
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
W 17025 and NIST handbook relationship (need advice) ISO 17025 related Discussions 8
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 4
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 14
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4

Similar threads

Top Bottom