#1
17025??????

Back to the old subject of 17025!!!!
Am I correct that our internal equipment such as load cells (for vessels up to 5 tonnes), balances etc used throughout the company (labs, manufacturing, dispensing) need to be calibrated by an extenal company certified to 17025?
What if the extenal calibration service concerned cannot calibrate equipment that weighs to say 5 tonnes (which is the position we currently face) because it is physically impossible to do so i.e. a calibrated 5 tonne weight on a load cell????
Have I got this issue completely wrong?
Can anyone clarify this for me?

Thanks

rvin28
 
Elsmar Forum Sponsor
T

Teri - 2011

#2
Can you get the originial manufacturer of the equipment to calibrate it. That is about the only way out of 17025. Other than customer waiver.
 

rvin59

Registered
#3
Unfortunately we can't. We have also submitted the customer waiver and you can imagine the response. Given the option, customers want calibration to 17025 - it seems to be a no win situation. Can you suggest any way around calibrating a 5 tonne vessel other than using calibrated weights to 5 tonnes?

Thanks

Rvin 28

:frust:
 
D

D.Scott

#4
Sorry for sticking my nose in here. I hope I'm not just adding useless information because I missed the point.

I don't think 16949 says you have to have the equipment calibrated by an external lab. If you do use one, it certainly has to be certified to 17025 or equivelent.

The requirement allows, for example, calibration by an internal lab. There is no requirement for the internal lab to be certified to 17025 although certification could be used to demonstrate conformity to the requirements of 7.6.3.1 Check out 7.6.3 and see if you meet the requirements under your internal system. If so, your "work around" may be as simple as bringing your lab up to snuff, writing your calibration procedures and doing the calibration yourself.

Dave
 

howste

Thaumaturge
Super Moderator
#5
What kind of equipment is this load cell on? I've got some ideas, but they assume you've got a lot of space under the load cell to work with...

Do know of anyone that has weights calibrated up to 5-tons? You may be able to work out an arrangement to rent the weights from them and calibrate the load cells yourself (with or without their assistance). I have a client that fabricates overhead bridge cranes that has weights calibrated up to 10 tons.

Another possibility is to use a huge water bag I've heard them talk about. They use it to load test cranes. They can fill it to known volumes and use the known weight of the water to calculate the weight. I don't know a lot of details about this one, but I thought I'd at least mention it.

Good luck!
 
Thread starter Similar threads Forum Replies Date
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
P ISO 9001 Requirements for ISO 17025 Accredited Calibration Laboratory ISO 17025 related Discussions 5
N ISO 17025 Requirements for Environmental Conditions for In-House Calibration ISO 17025 related Discussions 5
R ISO 17025 M&TE (Measuring & Test Equipment) Calibration Requirements ISO 17025 related Discussions 4
T ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements ISO 17025 related Discussions 6
D ISO 17025 and Calibration Laboratory Risk Management Requirements ISO 17025 related Discussions 1
N ISO 17025 External Calibration Laboratory Requirements within ISO/TS 16949 ISO 17025 related Discussions 28
S ISO 17025 Validation Requirements - "calibration using reference standards or..." ISO 17025 related Discussions 5
B Pin Gage Calibration Procedure - ISO 17025:2005 accreditation requirements? ISO 17025 related Discussions 8
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2
S Is it mandatory for a Calibration Lab to be ISO 17025 accredited? General Measurement Device and Calibration Topics 7
G ISO 17025 Calibration Laboratory Assessor's Common Questions and Procedures ISO 17025 related Discussions 11
J ISO 17025 Guide and Ideas for a New Calibration Business ISO 17025 related Discussions 18
pammesue Calibration in the UK to ISO 17025..cant find anyone ISO 17025 related Discussions 8
N Implementing ISO 17025 in a Calibration Laboratory ISO 17025 related Discussions 8
M Does Calibration to ISO/IEC 17025 conform to Z540.3? ISO 17025 related Discussions 1
Crusader Calibration Source not ISO 17025 accredited ISO 17025 related Discussions 6
E Should companies that perform calibration services be ISO 17025 accredited? ISO 17025 related Discussions 13
Sean Kelley What is the ISO 17025 Certified Laboratory Requirement for External Calibration? ISO 17025 related Discussions 8
AshleyE Calibration system must meet ISO 10012, ISO 17025 or ANSI/NCSL Z540? ISO 17025 related Discussions 38
L Calibration by a non-ISO 17025 Laboratory ISO 17025 related Discussions 4
Z Force & Strain Calibration Uncertainty for ISO/IEC 17025 ISO 17025 related Discussions 2
M Does 17025/NADCAP allow deferring Calibration Start Dates on Gages which are Stored? ISO 17025 related Discussions 10
D Calibration of Load Cells - ISO 17025 & MSA (Measurement Systems Analysis) ISO 17025 related Discussions 5
1 Pharma Industry stance on ISO 17025 Accreditation - In-House Calibration Laboratories ISO 17025 related Discussions 3
L Must Calibration Labs have IEC 17025 before doing Calibration Jobs for Customers ISO 17025 related Discussions 13
M ISO 17025 - Recommended or Required and other Calibration questions ISO 17025 related Discussions 5
M Specific Requirement for Calibration - ISO 17025 Clause 5.6.2 Interpretations ISO 17025 related Discussions 15
S Reference to original Calibration Certificate - ISO 17025 Section 5.10.9 ISO 17025 related Discussions 4
H Finding an ISO 17025 accredited laboratory to perform calibration ISO 17025 related Discussions 14
C Accredited Calibration Certificate Content - ISO 17025 Section 5.10.4.2 Question ISO 17025 related Discussions 7
suildur Assuring the Quality of Test and Calibration Results - ISO IEC 17025 Clause 5.9 ISO 17025 related Discussions 13
K Subcontracting of Tests and Calibration - ISO 17025 Clause 4.5 ISO 17025 related Discussions 1
A Compliance with ISO/ IEC 17025:2005 Clause 5.9 for Calibration Laboratories ISO 17025 related Discussions 6
howste Condition of Item Calibrated - ISO 17025 Clause 5.10.2f - Calibration Certificate ISO 17025 related Discussions 13
D Calibration Frequency of CMM as per ISO/IEC 17025 ISO 17025 related Discussions 11
U I need All ISO 17025 Documents for a calibration laboratory ISO 17025 related Discussions 7
H Can you tell me a suitable "tooling" for Caliper Calibration - ISO 17025 ISO 17025 related Discussions 16
Mohammed Gouda Old calibration standard methods - Question for ISO 17025 auditors..... ISO 17025 related Discussions 1
R ISO 17025 section 4.9 - Control of Nonconforming Testing and Calibration Work ISO 17025 related Discussions 6
T Calibration Procedures - ISO/IEC 17025, Clause 5.5.9 - Outsourced Calibration ISO 17025 related Discussions 10
J Compliant Calibration vs. Accredited 17025 Calibration - 1000ml Graduated Beaker ISO 17025 related Discussions 16
B ISO 17025 logo & ISO 9001 logo on calibration certificate/report ISO 17025 related Discussions 7

Similar threads

Top Bottom