21 CFR 820 Record Control Requirements
Mark75
Hello -
I'm looking for input about whether or not the following situation is considered a violation of 21 CFR 820.70, 820.80 and Good Documentation Practices in general. Here is the scenario -
The following information is being documented:
- Lab test results are recorded in non-controlled lab notebooks
- Some process parameter checks, such as oven temps and other equipment settings, and times for when these recordings were taken for each batch are recorded on non-controlled forms
- Quantities of product produced & scrapped are recorded on a blank sheet of paper
Then at a later time, perhaps at the end of each shift, or maybe even the next day (who really knows), this information is being entered into a validated Excel spreadsheet. The person entering the information isn't the same person that is recording the results. Then the above records are thrown away (not sure about the lab notebooks yet) after the data is inputed and not maintained in the DHR. The spreadsheet is the only thing that is printed out and maintained in the DHR.
This is not a closed system where data is immediately fed into the validated spreadsheet and the data is not automatically captured by any validated software. However, I was informed that the spreadsheet is validated as justification for recording data in this manner.
I'm not an expert when it comes to 21 CFR Part 11 compliance, but my opinion is that you still need to follow 21 CFR 820 record keeping req's and Good Documentation Practices. I definitely feel this is a compliance risk.
Is this even debatable? Whether or not the spreadsheet is validated when it's done this way my experience tells me "no".
Regards,
Mark
Stijloor
Hello -
Regards,
Mark
Anyone to help Mark?
Thank you!
Stijloor.
MIREGMGR
...you still need to follow 21 CFR 820 record keeping req's and Good Documentation Practices.
That statement of course is true.
As you've described the situation, I agree with your conclusion, assuming that this data is related to a process for which data retention is required.
Mark75
Thank you very much for some feedback. :thanx:
Yes, the recorded information all pertains to mfg process parameters and lab test results (also in place to verify product quality and conformance to spec req's). I figured this was pretty cut-and-dry, but 21 CFR Part 11 was thrown in as a potential justification, so I just wanted to gain feedback from others.
If anyone else happens to have any feedback please feel free to share.
Thanks,
Mark
Ajit Basrur
Hello -
Regards,
Mark
I agree Mark. 21 CFR part 11 is a distant non compliance but the example cited seems to be non compliance from 21 CFR Part 820.
Al Rosen
Although I agree the practices are not the best, for the sake of discussion, what specific requirement is being violated? I can't seem to find one.
MIREGMGR
Assuming as above that this data is related to a process for which data retention is required, the recordkeeping requirement would be inferred from 820.70 Production and Process Controls.
It's a reasonably standard inference, I think, that for processes that affect the conformance of products to specifications, recording of process parameters is an essential element of controlling that process and preventing deviations.
MIREGMGR
Aren't they meeting that requirement by noting the data on a "scratch pad" and then entering it into excel?
I think again that that depends on "assuming as above that this data is related to a process for which data retention is required".
For instance, a validated process, per Section "Documentation" of the "Medical Device Quality Systems Manual", Part 4 "Process Validation".
The rule actually states that a signature and date is required from the individual(s) approving the validation, but that's been interpreted in the past to mean that when multiple individuals collect data on which the validation depends, and no single individual can verify the collection of all the data, all such individuals must sign and date.
We've been gigged before for not having such an auditable retention of the original data collection records, with dated physical signatures, in an instance where we had transcribed too-informally collected data into a single record.
G
gar4guv
As I see it, this goes against the requirement 820.40. The noncontrolled lab books are being used to collect information that goes into the DHR. Since we know that a DHR is required, and 820.40 says to control all documents that are required, I would conclude that the lab notes should become part of the DHR. Therefore they should be controlled...or not used at all.
These are the 483s I've seen written regarding noncontrolled docs being used on the manufacturing floor or in a qc lab,
Documents that were not approved were observed at a location where they are being used. Specifically, (insert specific evidence)
and that one is almost always followed by,
Document control procedures were not [complete] [implemented]. Specifically, (insert specific evidence)
Sec. 820.40 Document controls. Each manufacturer shall establish and maintain procedures to control all documents that are required by this part. The procedures shall provide for the following:
(a)Document approval and distribution. Each manufacturer shall designate an individual(s) to review for adequacy and approve prior to issuance all documents established to meet the requirements of this part. The approval, including the date and signature of the individual(s) approving the document, shall be documented. Documents established to meet the requirements of this part shall be available at all locations for which they are designated, used, or otherwise necessary, and all obsolete documents shall be promptly removed from all points of use or otherwise prevented from unintended use.
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
B 21 CFR 820.180 Quality Record Retention Requirements Document Control Systems, Procedures, Forms and Templates 1
W Quality System Record as required by CFR 21 Part 820.186. ISO 13485:2016 - Medical Device Quality Management Systems 13
M Quality Record Retention Time - ISO 9001 vs. CFR 820.180(b) Records and Data - Quality, Legal and Other Evidence 25
A TUV Audits - 21 CFR 820 General Auditing Discussions 6
T 21 CFR 820.20 - Quality Planning Requirements? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M 21 CFR 820 vs 21CFR820 vs 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
A 21 CFR 820 - Risk Management - Looking for some guidance US Food and Drug Administration (FDA) 3
P Equipment 21 CFR 820.70(g) - User Requirements Document for Off the shelf equipment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
D ISO 13485, FDA 21 CFR 820 and Auditing the Accounting Department ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek 21 CFR Part 820 - FDA Label Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
R Addressing training requirements - 21 CFR Part 820.25 (1) & (2) Other US Medical Device Regulations 4
R How to improve a Validation program and procedures to FDA (21 CFR part 820) & ISO13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Marc Problem with 21 CFR Part 820 - US FDA Quality System Regulations (QSR) sub-forum link - 2 May 2019 Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 1
R 21 CFR Part 820 Contract Manufacturer of Medical Device Component 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
N Change Control - Compliance with FDA 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
D 21 CFR Part 820 (Subpart A) - Question about "Authority" 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R RUO audits - CFR 820.22 and ISO 13485 8.2.4 Internal Auditing 3
A Training as well as certifications for ISO 13485:2016, 21 CFR 820... Training - Internal, External, Online and Distance Learning 5
K Production Unit Labeling - 21 CFR 820.184.e 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R 21 CFR Part 820.186 - Types of Quality System Records Document Control Systems, Procedures, Forms and Templates 1
M Does the Scope of 21 CFR Part 820.72 (Equipment) apply to Design? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V 21 CFR 820 Compliant way to use Kaizen to address Nonconformances Nonconformance and Corrective Action 9
C 21 CFR Part 820.184 - Label Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
shimonv 21 CFR 820.20(d) says: Each manufacturer shall establish a Quality Plan 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V Upgrading Systems from CFR 211 to CFR 820 (drug+device combination) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
S Looking for a checklist comparing ISO 13485:2016 and 21 CFR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 14
Q Internal Audit of Product Quality Complaint System (21 CFR Part 820) Customer Complaints 9
R Managing Employee Training Files - 21 CFR Part 820.25 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M Release authorization before or after packaging per 21 CFR 820.80(d)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Need advice in streamlining antiquated 21 CFR 820 compliant QMS Quality Manager and Management Related Issues 2
J Minimum staff per 21 CFR Part 820 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Does a process map satisfy 21 CFR 820.20(d) (Quality Plan)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Calibration of processing equipment - 21 CFR 820.70 Production and process controls 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Issues with UDI additions to 21 CFR 820 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Interpreting Process Controls - 21 CFR Part 820.70(a) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How others have defined "reaudits"? 21 CFR Part 820.22 Internal Auditing 1
M FDA 21 CFR 820.250 - Does "valid statistical" always mean math? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
J Document Control Features (Document Content Requirements) and 21 CFR 820.40 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S 21 CFR Part 820.40(b) Clarification on Required Document Approvers 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Where to find 21 CFR Part 820 Translated into Czech, Spanish, and German 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
N Certificate of Compliance from a Sister Company for FDA 21 CFR 820 Exemption 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Does FDA expect suppliers to be independently assessed to 21 CFR Part 820 ? US Food and Drug Administration (FDA) 4
A Effective Date (of Documents) issue outlined in 21 CFR Part 820.4 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
L Mobile Medical App - Understanding 21 CFR Part 820 Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
T "Special Process" Validation Requirements (21 CFR Part 820) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
Y ISO 13485:2003 &FDA QSR (21 CFR 820) Quality Manual,34 Procedures and Form Book, Video, Blog and Web Site Reviews and Recommendations 8
R Policy Statement for FDA CFR 820 Acceptance Activities of Incoming Products. Other US Medical Device Regulations 1
shimonv Storage of finished medical devices per 21 CFR Part 820.150 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
M Does 21 CFR 820.80 (b) require a Supplier to have Receiving Inspection? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13