21CFR820-820.50 Purchasing controls - Using Non-Approved Suppliers

S

suziwann

#1
HI, I have a question regarding approved suppliers. I hope that someone can help.

We have an approved supplier register and supplier assessment procedure.

There is a section within the procedure which allows us to use 'non-approved suppliers' when absolutely necessary as a 'one off.'

Is this practice acceptable by the FDA?

My thoughts are that this is not acceptable after reading the FDA guidelines on Purchasing Controls.

I thank you in anticipation.
 
Last edited by a moderator:
Elsmar Forum Sponsor

Al Rosen

Staff member
Super Moderator
#2
suziwann said:
HI, I have a question regarding approved suppliers. I hope that someone can help.

We have an approved supplier register and supplier assessment procedure.

There is a section within the procedure which allows us to use 'non-approved suppliers' when absolutely necessary as a 'one off.'

Is this practice acceptable by the FDA?

My thoughts are that this is not acceptable after reading the FDA guidelines on Purchasing Controls.

I thank you in anticipation.
I have a few questions since you may be restricting yourself unnecessarily.




  1. What does your assessment procedure require before a supplier is approved?
  2. What is are the FDA Guidelines you refer to?
This is the rule:
Sec. 820.50 Purchasing controls. Each manufacturer shall establish and maintain procedures to ensure that all purchased or otherwise received product and services conform to specified requirements.

(a) Evaluation of suppliers, contractors, and consultants. Each manufacturer shall establish and maintain the requirements, including quality requirements, that must be met by suppliers, contractors, and consultants. Each manufacturer shall:

(1) Evaluate and select potential suppliers, contractors, and consultants on the basis of their ability to meet specified requirements, including quality requirements. The evaluation shall be documented.
From the FDA's Medical Device Quality Systems Manual: A Small Entity Compliance Guide
Supplier Qualifications


A major factor in obtaining high quality components is the selection of suppliers. Although a manufacturer's knowledge of supplier operations may be limited and information about the operations difficult to obtain, the GMP requirement that a manufacturer is responsible for quality remains undiminished. To the maximum extent feasible, selection and qualification of suppliers by audits, performance analysis, etc., should be part of a quality system. If the manufacturer does not have the capability to test components for conformance to specifications, then supplier test data or outside lab results are acceptable provided that components are tested and inspected in a statistically valid manner to show their acceptability for use in the finished device. Any outside test results should be accompanied by relevant raw data used for the test so that judgments of authenticity may be made by the finished device manufacturer. Excluding a supplier whose components are unreliable from supplying components may help prevent problems with the final device and is certainly worthwhile as a cost reduction effort.

It is important to remember that raw components acquire cumulative value as they are processed through receiving, assembly, test, inspection, and as they ultimately become part of the finished device. If a component fails during assembly, or as part of the device, additional costs will be incurred for fault isolation, removal, replacement, inspection, testing, etc. When field failures occur, the ultimate cost of the component becomes even higher because its replacement requires travel, trouble-shooting, and retrofit. In addition, customer dissatisfaction, user injury, product liability action, medical device reporting, or regulatory action may result. Usually, the initial cost of a component is relatively insignificant compared to the later cost should the component prove to be defective or improper for the selected use. Many recalls occur because manufacturers fail to qualify components properly or to assure that a supplier's manufacturing methods and quality system are adequate.
Why not purchase material and perform an evaluation and based on the performance of the initial lot, the supplier may be added to your "approved supplier register". Your procedures can be revised to allow this.
 

Wes Bucey

Quite Involved in Discussions
#3
Nice expedient solution, Al!

Thanks again for reminding folks about the FDA manual available for download (a zip file containing all chapters or download individual chapters one-at-a-time.)
 
S

suziwann

#4
Thankyou for your reply.
In answer to your questions.

1 What does your assessment procedure require before a supplier is approved?Our procedure requires a self assessment questionaire for the supplier. we then decide on their suitability. Supplier audit is then performed, they are then either approved or rejected.

2 What is are the FDA Guidelines you refer to? The 'guidelines' I refer to are the regulations in the title of this post.


The FDA manual is very useful. I keep losing the link, thanks! :agree1:
 

Al Rosen

Staff member
Super Moderator
#5
suziwann said:
Thankyou for your reply.
In answer to your questions.

1 What does your assessment procedure require before a supplier is approved?Our procedure requires a self assessment questionaire for the supplier. we then decide on their suitability. Supplier audit is then performed, they are then either approved or rejected.

2 What is are the FDA Guidelines you refer to? The 'guidelines' I refer to are the regulations in the title of this post.


The FDA manual is very useful. I keep losing the link, thanks! :agree1:
  1. Have the survey submitted from the supplier with the first lot or in lieu of it evaluate the first lot as I stated in post 2.
  2. To be precise, that is the regulation and not a guideline. In the business we are in language and it's interpretation are very important.
  3. Bookmark that link.
 
S

suziwann

#6
That wasn't very clear at all was it!:eek:

Firstly the regulations that that I was concerned with are in the title of the post.

I then read the guidelines to get some advice but was still not absolutely sure after reading them.

I still haven't managed to pin down my manager to get this procedure changed. It states that all suppliers must be audited every twelve months. They haven't done an external audit in the last five years or more:eek: :eek:
 

Al Rosen

Staff member
Super Moderator
#7
suziwann said:
That wasn't very clear at all was it!:eek:

Firstly the regulations that that I was concerned with are in the title of the post.

I then read the guidelines to get some advice but was still not absolutely sure after reading them.

I still haven't managed to pin down my manager to get this procedure changed. It states that all suppliers must be audited every twelve months. They haven't done an external audit in the last five years or more:eek: :eek:
Incoming inspection is a means of controlling a supplier. By whatever means you plan on controlling suppliers, your procedure must state it. You must say what you do and do what you say.
 
D

Dosumi

#8
supplier approvals

Hello everyone,

This wasn't what I was looking for but would like an opinion on what we have for supplier approvals. I changed this some time ago to meet what was going on. We make a FDA CL 1 device.


5.0 APPROVED VENDOR / SUPPLIER
Vendor / Supplier performance is the main criteria used to maintain approved status. A vendor audit or survey may be used for the initial or continued approval of new or existing Vendor/s or Supplier/s. In lieu of an audit or survey a Vendor / Supplier approval is based on performance as tracked through the receiving inspection history process. Management has the ultimate authority to approve or disapprove a specific Vendor / Supplier, regardless of past performance.
Note: Approval status of a Vendor / Supplier in no way constitutes acceptance of a supplied product or component.


Thanks
Don
 

Al Rosen

Staff member
Super Moderator
#9
Dosumi said:
Hello everyone,

This wasn't what I was looking for but would like an opinion on what we have for supplier approvals. I changed this some time ago to meet what was going on. We make a FDA CL 1 device.


5.0 APPROVED VENDOR / SUPPLIER
Vendor / Supplier performance is the main criteria used to maintain approved status. A vendor audit or survey may be used for the initial or continued approval of new or existing Vendor/s or Supplier/s. In lieu of an audit or survey a Vendor / Supplier approval is based on performance as tracked through the receiving inspection history process. Management has the ultimate authority to approve or disapprove a specific Vendor / Supplier, regardless of past performance.
Note: Approval status of a Vendor / Supplier in no way constitutes acceptance of a supplied product or component.


Thanks
Don
What is your initial criteria for selecting a supplier in lieu of the survey?
 
D

Dosumi

#10
5.0 APPROVED VENDOR
Vendor performance is the main criteria used to maintain approved vendors. A vendor survey may be used for the initial approval of new vendors. Management may in lieu of a survey give conditional approval. Vendor performance is tracked through the receiving inspection history process and Quality Assurance Management. Management has the ultimate authority to approve or disapprove a specific vendor, regardless of past performance.

This is the current approval statement.
The previous post was of an in-work SOP.
Is the underlined bold statement acceptable or would I need to expand on it.

Thanks
Don
 
Thread starter Similar threads Forum Replies Date
M 21 CFR 820 vs 21CFR820 vs 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
J 21CFR820 requirements for Foreign Exporter? US Food and Drug Administration (FDA) 4
F Distribution Records required by FDA - 21CFR820.160 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Q When does the FDA deem something "where appropriate"? 21CFR820.30(g) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
Q 21CFR820 as it was in 1993 or 1997? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Is a Dynamic Controlled Form acceptable per 21CFR820 Document Control Systems, Procedures, Forms and Templates 3
M Differing interpretations in ISO 17025 and 21CFR820.72(b)(2) ISO 17025 related Discussions 5
S Packaging and Label Change Requirements - ISO 13485/FDA 21CFR820 ISO 13485:2016 - Medical Device Quality Management Systems 3
M 21CFR820.120 Device Labeling Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
sagai Preventive Action vs. Corrective Action as defined by 21CFR820 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
J Correlation matrix between HSA TS-01 and 21CFR820 or ISO13485 Other Medical Device Regulations World-Wide 1
V Implementing Dock-to-Stock when regulated by ISO 13485 and 21CFR820 ISO 13485:2016 - Medical Device Quality Management Systems 12
M FDA 21CFR820, ISO13485 and AS9100 co-existence? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
J ISO 13485, Canadian MDR and FDA 21CFR820 Comparison Matrix ISO 13485:2016 - Medical Device Quality Management Systems 14
N Requirements to Comply With FDA 21CFR820 For Invitro Diagnostic Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A FDA Quality Audits 21cfr820.22 - Audits by the head of the company not allowed ISO 13485:2016 - Medical Device Quality Management Systems 1
R 21CFR820 vs. ISO 9000:2000 - Cross Reference Matrix Attached ISO 13485:2016 - Medical Device Quality Management Systems 5
A Document Review and Document Approval --- 2 Signatures needed acc. §820.40? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A TUV Audits - 21 CFR 820 General Auditing Discussions 6
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
T 21 CFR 820.20 - Quality Planning Requirements? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A 21 CFR 820 - Risk Management - Looking for some guidance US Food and Drug Administration (FDA) 3
P Equipment 21 CFR 820.70(g) - User Requirements Document for Off the shelf equipment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
D ISO 13485, FDA 21 CFR 820 and Auditing the Accounting Department ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek 21 CFR Part 820 - FDA Label Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B Per 21CFR 820.198 (b) - Justification for no investigation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 22
Watchcat Does 820.30 include the manufacturing process? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K 820.200 Servicing Procedure - Our system (because there is an API involved) is classified as Class 3 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1
R Addressing training requirements - 21 CFR Part 820.25 (1) & (2) Other US Medical Device Regulations 4
R How to improve a Validation program and procedures to FDA (21 CFR part 820) & ISO13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Marc Problem with 21 CFR Part 820 - US FDA Quality System Regulations (QSR) sub-forum link - 2 May 2019 Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 1
R 21 CFR Part 820 Contract Manufacturer of Medical Device Component 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Q Any good Checklists for ensuring SOPs cover ISO 13485:2016 and 21CFR 820? ISO 13485:2016 - Medical Device Quality Management Systems 3
B Classes/ Online Training on ISO 13485:2016 and FDA QSR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 5
C ISO 13485 certified as precursor to regulatory compliance to 21 CRF Part 820? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
N Change Control - Compliance with FDA 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
D 21 CFR Part 820 (Subpart A) - Question about "Authority" 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R RUO audits - CFR 820.22 and ISO 13485 8.2.4 Internal Auditing 3
A Training as well as certifications for ISO 13485:2016, 21 CFR 820... Training - Internal, External, Online and Distance Learning 5
K Production Unit Labeling - 21 CFR 820.184.e 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R 21 CFR Part 820.186 - Types of Quality System Records Document Control Systems, Procedures, Forms and Templates 1
M Does the Scope of 21 CFR Part 820.72 (Equipment) apply to Design? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V 21 CFR 820 Compliant way to use Kaizen to address Nonconformances Nonconformance and Corrective Action 9
C 21 CFR Part 820.184 - Label Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
shimonv 21 CFR 820.20(d) says: Each manufacturer shall establish a Quality Plan 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V Upgrading Systems from CFR 211 to CFR 820 (drug+device combination) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
S Looking for a checklist comparing ISO 13485:2016 and 21 CFR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 14
Q Internal Audit of Product Quality Complaint System (21 CFR Part 820) Customer Complaints 9
D 820.120 UDI (Unique Device Identification) Labeling Verification Requirements Other US Medical Device Regulations 11