2x MOOP for type B AP - Therapeutic device - Heater/cooler with pump

P

Publius

Hello,

Currently we’re designing a therapeutic device with a nonconductive type B applied part. The device consists of a heater/cooler where cooled or heated water is pumped through a non-conductive type B applied part. The water reservoir is conductive and bonded to PE. There are no external electrical connections to the device, only a GUI for the operator.

The pump and cooler are 24Vdc and are supplied by a 2x MOOP rated 24Vdc supply, the device is class I. The heater is mains supplied and bonded to PE with 1 MOPP insulation (1.5kV).
The 24Vdc pump and cooler are uncertified components so if these fail it isn’t considered as a SFC. Since the supply is only rated for 2x MOOP another MOP is required. The easiest solution is for the AP to provide the additional MOP but it is desirable for the AP to have minimal requirements, I’m not sure if this is possible though.

Although the AP is nonconductive if I assume it is conductive and bonded to PE than he NC/SFC condition AC leakage currents are met since these are the same for a type B AP as for the earth leakage currents. This also applies when the SFC is a loose PE connection. Knowing this I would like to use a conductive tube connector (for the additional MOP) bonded to the chassis (PE) for connecting the AP because this would result in the water potential being lowest in a SFC.

Because the pump is uncertified it failing isn’t considered a SFC. If I assume the pump has failed (24Vdc on water) AND the SFC is a loose PE, then the only option is to leave the secondary GND floating to comply with the allowed DC leakage current but this seems counterintuitive to me.

The worst-case scenario would be that the water becomes live when the pump and PSU fail simultaneously. The water resistance is assumed to be > 50Ohm (depending on salt level) towards the grounded tube connector. For this fault connecting the secondary ground to PE is surely better.

Alternatively, the tube connector could be insulating (1 MOP 500V).


Hopefully someone can make sense of the above as it is difficult to describe and I’m going in circles with this problem.

After re-reading the above I think the best option is:
- Tube connector bonded to PE.
- Secondary GND connected to PE.
- AP should provide 24V/10uA = 2.4 Meg DC resistance.

Does anyone see any flaws in my assessment of the system or have some more insights?

Thanks
 
M

MedMartin

Hello Publius,

an insulation diagram would be helpful to illustrate your thoughts and might also help you to find flaws in the considerations. With the diagram the requirements are obvious and the solution then follows.

Best regards,
Martin
 
Thread starter Similar threads Forum Replies Date
W External PSU Providing a MOOP -- Will This Necessitate Conducted Emissions Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Protective impedance as MOOP/MOPP - Requirements? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B Accessible parts per 8.4.2 b) - Touch currents in combination with MOOP or MOPP? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1 - MOPP and MOOP Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P MOOP and insulation coordination IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G Creepage and Clearance Distance for MOOP on Secondary Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
G Trend analysis according to Article 88, MDR (EU), type of data point. EU Medical Device Regulations 1
M Type of bodily contact Other Medical Device Related Standards 1
R MSA Type 3 vs Type 1 Capability, Accuracy and Stability - Processes, Machines, etc. 12
X Separation of battery powered FB type applied part IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Separation and PE Requirements of a Type B Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Gage type and data base maintainence Using GAGEpack Software 2
Judy Abbott Type of contact- 10993-1:2018 Medical Device and FDA Regulations and Standards News 2
P Type of MSA study according to measurement equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
CaliperJim Interval Type After First Use (AFU) General Measurement Device and Calibration Topics 6
R Repeatability Type 1 Gage and Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Z Why Control Limits are not the same depending on type of exclusion of data points Using Minitab Software 7
A IATF MSA study Type 1 General Measurement Device and Calibration Topics 7
M Type 1 Gage r study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Nihls MSA Study Type 2 (special conditions ) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
W What is the difference between TYPE B and TYPE BF? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D Separation of F-type applied part and remaining parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
M MSA Study Type 1 not capable. We are at the limit. And manufacturing wants to continue producing. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
cnbrosa Study Type 1 on a CMM using a measuring support Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
A Fabric roll inspection - What type of Control Chart to use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
A AMSQQP416E Class 2 type 2 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Are EC type examinations still being conducted under MDR? EU Medical Device Regulations 5
S Medical watch Class II (AP Type CF) with USB connection IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Johnnymo62 Non Aerospace topics - Anything for military trucks, trailers, Humvee type vehicles? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
N BF-type applied part MOPP vs secondary IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Marcel DS GOST-R (Type approval for Vehicles in Russia) Other ISO and International Standards and European Regulations 0
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
I Old Time Scatter diagrams for defect type and location- software Quality Tools, Improvement and Analysis 7
P Surgical masks ( type IIR) - Cleanroom Requirements EU Medical Device Regulations 3
W Insulation requirements of BF type infant skin temperature probe IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Jane's Like-for-like critical raw material change qualification - type of testing/ number of lots required ISO 13485:2016 - Medical Device Quality Management Systems 4
N Infant radiant warmer - Heater Quartz Element type Other Medical Device Related Standards 0
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
K Validation of new machine (second machine of the same type) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
F Operator Manual and Type B Uncertainty Measurement Uncertainty (MU) 3
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1
M SI #7 8.4.2.1 Type and extent of control_supplemental IATF 16949 - Automotive Quality Systems Standard 3
B F-type applied part - Separation from ALL(?) other parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
A IATF Sanctioned Interpretation No. 7 - Type and Extent of Control (supplemental) IATF 16949 - Automotive Quality Systems Standard 4
K F-TYPE APPLIED PART IEC 60601 - Medical Electrical Equipment Safety Standards Series 0

Similar threads

Top Bottom