Is anyone able to provide me with some guidance regarding the level of detail for items in the 510k such as providing a summary of configuration management or maintenance activities?

For example, do we have to explain that maintenance activities are requested by internal/external customers, request is documented, goes for quality assessment, etc.

Any advise would be most helpful.

Cheers, Tonia
 
Elsmar Forum Sponsor
I

isoalchemist

#2
At the level I believe you are asking the question the answer is NO. The 510k not to show control of your processes (21CFR820) but to show that the device you are proposing is substantially equivalent to the predicate devices chosen. You must show that it gives the same results and uses similar technology or other technolgy that is proven (usually a second predicate device).

Basically you are proving the device is safe and effective, not telling your processes.

With that said if the User need to perform maintenance, that should be included as part of the Use Manual (i.e. labeling)
 

tebusse

Involved In Discussions
#4
Now I'm a little confused.

The guidance documentation for software states that we have to discuss configuration and change management along with maintenance. How is one to describe this as it relates to the predicate device?

Please note, this information is to be included in the section pertaining to the device's level of concern.

Regards, Tonia
 

tebusse

Involved In Discussions
#6
Thank you. This guidance document is what I've been following and it seems to me that discussing the design process, configuration/change management, maintenance, V&V are all part of demonstrating that the device is safe and effective.

Correct me if I'm wrong, but this section of the 510(k) is not necessarily designed to demonstrate substantial equivalence to the predicate device.

Regards, Tonia
 

Marcelo

Inactive Registered Visitor
#7
This is related to substantial equivalence, but in a weird way. It's more related to safety in general.

The problem is the way safety is ensured for software.

You can assess safety and equivalence in hardware by means of information and testing and related stuff.

For software, you can only assess safety by controlling the development (full) lifecycle. And you really cannot directly compare two different softwares the way you can do to hardware.

So, in a way, the information you have to provide does show that your software is "safe".

Also, you could take a look into IEC 62304 for more help.
 
Last edited:
#8
A Major Level of Concern device requires enough detail to allow for a thorough understanding of the configuration management and maintenance plan.

For a Moderate Level of Concern Device a summary of the configuration management and maintenance plans is sufficient.

Correct me if I'm wrong, but this section of the 510(k) is not necessarily designed to demonstrate substantial equivalence to the predicate device.
You are correct. This section is designed to determine the Level of Concern and to supply the documentation according to that determination of the Level of Concern
 
Thread starter Similar threads Forum Replies Date
A Sterility Assurance Level? Special 510k US Food and Drug Administration (FDA) 5
N Traditional 510K Submission-Financial Certification or Disclosure Statement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
P Letter to file or special 510k or traditional 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Y 510k requirement for bringing in gloves Medical Device and FDA Regulations and Standards News 0
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
H 510k with Contract Manufacturer Other US Medical Device Regulations 2
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
K 510K new manufacturer but same components Other US Medical Device Regulations 6
S Two or more predicates suitable? 510K submission US Food and Drug Administration (FDA) 5
S 510k: What to include if your own device is the predicate US Food and Drug Administration (FDA) 5
L 510K - Trade Name Change 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Ed Panek Adverse Event Clinical Trial using a 510K approved Device Other US Medical Device Regulations 6
R Countries that accept 510k approvals - Class II nonsterile medical device Other Medical Device Regulations World-Wide 2
M Timing of 510k submission Other US Medical Device Regulations 9
S Examples of software changes that required a 510k US Food and Drug Administration (FDA) 2
D 510K and Changes to Verification and Validation US Food and Drug Administration (FDA) 2
Ronen E Can anyone please explain this? (510k submission requirement) Other US Medical Device Regulations 14
H New 510K VS Documents about design change. Other US Medical Device Regulations 1
P Medical Device not on commercial distribution but installed in the establishment. 510k? Other US Medical Device Regulations 10
V Presubmission FDA for 510k US Food and Drug Administration (FDA) 4
A How to deal with changed shared components in 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Can we take a part from 510k cleared medical device and use it in class I device? Other US Medical Device Regulations 3
D Contents of a 510k Q submission Other US Medical Device Regulations 1
L Class IIa Medical Device, CE Mark & 510k - Labeling Content CE Marking (Conformité Européene) / CB Scheme 4
Ed Panek Other company wants to use our FDA 510K bundled with their product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
L 510k exemption process - Do you still need to submit technical dossier? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S IEC 62133 - CB Scheme Required by FDA 510k Other Medical Device Related Standards 3
S The key differences in 510k types - Traditional vs. Abbreviated vs. Special US Food and Drug Administration (FDA) 0
P V&V for Special 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Clarity on requirements for a 510k exempt Class I device - Suture removal kit US Food and Drug Administration (FDA) 9
T Is 510k Anesthesia group particularly picky/difficult to work with during 510k? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J 510k - Product shelf life validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
K 510K statement Questions 21 CFR 807.93 - Discontinued or never released for sale US Food and Drug Administration (FDA) 0
J Applying for 510K - What is a significant process change and what is not US Food and Drug Administration (FDA) 5
K 510K Interactive Review 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
I Celsius and Fahrenheit in 510k US Food and Drug Administration (FDA) 4
R FDA 510K - Does the FDA accept the testing report from other countries? Other US Medical Device Regulations 1
I 510k Performance Testing Studies (Raw Data) - What does the FDA expect 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P Is a different 510k is required for each (stand alone) software configurations? Other US Medical Device Regulations 3
K ISO 10993 Requirements For 510k - Substantial Equivalence Other US Medical Device Regulations 11
P Two separate 510k submissions for one device from the same company Other US Medical Device Regulations 10
I Special 510k Imported Legacy Blogs 0
K 510k "Premarket" Submission for Existing Class II Software Medical Device US Food and Drug Administration (FDA) 3
N Is a new 510K required for change of manufacturing location for an existing device? US Food and Drug Administration (FDA) 1
Q 510k Submission for a New Medical Device Product Other US Medical Device Regulations 3
W New 510k required for a material change? Other US Medical Device Regulations 9
M What if information in predicate 510k is obviously wrong? Other US Medical Device Regulations 5
E 'Special' 510k as a Predicate Medical Device? Other US Medical Device Regulations 3

Similar threads

Top Bottom